Dear All, Please let me know what's the reason or Consequence for difference in the retention time of Analyte and Internal Standard. As per regulatory requirment is there any specific criteria. Thanks in Advance ![]() |
|
Helmut ★★★ ![]() Vienna, Austria, 2009-07-21 15:48 (6186 d 19:40 ago) @ Hemant7613 Posting: # 3975 Views: 2,619 |
|
|
Dear Hemant! ❝ [...] reason or Consequence for difference in the retention time of Analyte ❝ and Internal Standard. As per regulatory requirment is there any specific ❝ criteria. Reason: lipophilicity, pk, degree of ionisation of both analyte and IS; pH, buffer capacity and percentage of organic modifier of the mobile phase, type of stationary phase, column length/diameter, temperature, flow rate,... Consequence: Retention times per se are irrelevant! You are interested in good separation. FDA recommends: Rs of >2 between the peak of interest and the closest potential interfering peak (impurity, excipient, degradation product, internal standard, etc.) is desirable. See this post for the reference.— Dif-tor heh smusma 🖖🏼 Довге життя Україна! ![]() Helmut Schütz ![]() The quality of responses received is directly proportional to the quality of the question asked. 🚮 Science Quotes |