Dear All,

Please let me know what's the reason or Consequence for difference in the retention time of Analyte and Internal Standard. As per regulatory requirment is there any specific criteria.

Thanks in Advance
:-)
Helmut
★★★
avatar
Homepage
Vienna, Austria,
2009-07-21 15:48
(6186 d 19:40 ago)

@ Hemant7613
Posting: # 3975
Views: 2,619
 

 Resolution (analyte - IS)

Dear Hemant!

❝ [...] reason or Consequence for difference in the retention time of Analyte

❝ and Internal Standard. As per regulatory requirment is there any specific

❝ criteria.


Reason: lipophilicity, pk, degree of ionisation of both analyte and IS; pH, buffer capacity and percentage of organic modifier of the mobile phase, type of stationary phase, column length/diameter, temperature, flow rate,...
Consequence: Retention times per se are irrelevant! You are interested in good separation.
FDA recommends:

Rs of >2 between the peak of interest and the closest potential interfering peak (impurity, excipient, degradation product, internal standard, etc.) is desirable.

See this post for the reference.

Dif-tor heh smusma 🖖🏼 Довге життя Україна! [image]
Helmut Schütz
[image]

The quality of responses received is directly proportional to the quality of the question asked. 🚮
Science Quotes
UA Flag
Activity
 Admin contact
23,655 posts in 4,993 threads, 1,570 registered users;
279 visitors (0 registered, 279 guests [including 21 identified bots]).
Forum time: 11:28 CEST (Europe/Vienna)

Most scientists today are devoid of ideas, full of fear, intent on
producing some paltry result so that they can add to the flood
of inane papers that now constitutes “scientific progress”
in many areas.    Paul Feyerabend

The Bioequivalence and Bioavailability Forum is hosted by
BEBAC Ing. Helmut Schütz
HTML5