Phoenix WinNonlin: Variance estimate → CV, Welch-test [Software]
mystery solved. Phoenix WinNonlin gives in
Output Data → Final Variance Parameters the estimated variance and not the CV, which is for \(\small{\log_{e}\textsf{-}}\)transformed data:$$\small{CV=100\sqrt{\exp (\widehat{s}^{2})-1}}$$You have to setup a Custom Transformation in the Data Wizard:![[image]](img/uploaded/image876.png)
❝ I used average bioequivalence with parallel design and formulation R as the reference. The only fixed effect is the Treatment (TRT) with Ln(X) transformation. No variance structure (left all empty).
However, you should not assume equal variances in a parallel design. This was not stated in the User’s Guide of WinNonlin 8.1. See the online User’s Guide of v8.3 for the setup. Forget the Levene pretest (which inflates the type I error) – always use this setup in the future. Then instead of the t-test the Welch-Satterthwaite-test with approximate degrees of freedom is applied. For your complete data I got:
![[image]](img/uploaded/image877.png)
K015 had no concentrations after 48 h (which was also the Cmax) and the AUC was way smaller than the ones of the other subjects.This has consequences for the BE-calculation since the degrees of freedom will be different as well. Therefore, the point estimates in both models are the same but the confidence intervals are not:
![[image]](img/uploaded/image878.png)
SAS the Welch-test is the default.After excluding subject
K015 I could confirm your results:![[image]](img/uploaded/image879.png)
![[image]](img/uploaded/image880.png)
PowerTOST:library(PowerTOST)
metric <- c("Cmax", "AUC")
lower <- c( 74.75, 75.21)
upper <- c(136.56, 141.15)
df <- data.frame(metric = metric, CV = NA)
for (j in seq_along(metric)) {
df$CV[j] <- sprintf(" ~%.2f%%", 100 * CI2CV(lower = lower[j], upper = upper[j],
n = c(10, 9), design = "parallel"))
}
names(df) <- c("PK metric", "CV (total)")
print(df, row.names = FALSE, right = FALSE)
PK metric CV (total)
Cmax ~39.08%
AUC ~40.96%
CI2CV():The calculations are further based on a common variance of Test and Reference treatments in replicate crossover studies or parallel group study, respectively.
(my emphasis)
Dif-tor heh smusma 🖖🏼 Довге життя Україна!
![[image]](https://static.bebac.at/pics/Blue_and_yellow_ribbon_UA.png)
Helmut Schütz
![[image]](https://static.bebac.at/img/CC by.png)
The quality of responses received is directly proportional to the quality of the question asked. 🚮
Science Quotes
Complete thread:
- CV% and confidence interval Darborn 2024-05-29 08:26 [Software]
- Phoenix WinNonlin vs PowerTOST Helmut 2024-05-29 09:39
- Phoenix WinNonlin vs PowerTOST Darborn 2024-05-30 01:35
- Phoenix WinNonlin: Variance estimate → CV, Welch-testHelmut 2024-05-30 09:21
- Base 🇷 for comparison Helmut 2024-05-30 11:08
- Base 🇷 for comparison Darborn 2024-05-31 02:07
- Base 🇷 for comparison Helmut 2024-05-30 11:08
- Phoenix WinNonlin: Variance estimate → CV, Welch-testHelmut 2024-05-30 09:21
- Phoenix WinNonlin vs PowerTOST Darborn 2024-05-30 01:35
- Phoenix WinNonlin vs PowerTOST Helmut 2024-05-29 09:39