blood centrifugation [Study Per­for­mance]

posted by Helmut Homepage – Vienna, Austria, 2008-04-22 16:41 (6230 d 18:42 ago) – Posting: # 1790
Views: 6,233

Dear Sudeshna!

❝ Generally parameters set for centrifuge is 3000 to 4000 rpm ...


Since the separation effect is dependent on the force applied – not on “revolutions per minute” – such a description is meaningless unless the diameter of the centrifuge’s rotor is stated as well.
See this post and the linked document.

We routinely applied 2040×g at +4 ℃ for 20 minutes.

Dif-tor heh smusma 🖖🏼 Довге життя Україна! [image]
Helmut Schütz
[image]

The quality of responses received is directly proportional to the quality of the question asked. 🚮
Science Quotes

Complete thread:

UA Flag
Activity
 Admin contact
23,424 posts in 4,927 threads, 1,671 registered users;
141 visitors (0 registered, 141 guests [including 7 identified bots]).
Forum time: 11:23 CEST (Europe/Vienna)

Only dead fish go with the current.    Scuba divers' proverb

The Bioequivalence and Bioavailability Forum is hosted by
BEBAC Ing. Helmut Schütz
HTML5