 sample centrifugation [Study Per­for­mance]

what are the parameters of sample centrifugation to be followed unless the sponser specifies it.
Jaime_R
★★  

Barcelona,
2006-10-12 14:04
(6399 d 18:35 ago)

@ ssk
Posting: # 310
Views: 6,932
 

 sample centrifugation

Hi ssk!

❝ what are the parameters of sample centrifugation to be followed unless the

❝ sponser specifies it.


The ones which give you good separation of plasma in the shortest possible time ;-)

In a cooled centrifuge 1500×g for 15 minutes is a rule of thumb (you have to calculate rev/min basing on the rotor's diameter).

PS: be polite!

Regards, Jaime
ssk
●    

2006-10-12 14:29
(6399 d 18:11 ago)

@ Jaime_R
Posting: # 312
Views: 6,965
 

 sample centrifugation

Sir/Madam,
What does the g stand for. What is the standard temperature at which the centrifugation usually done. How are these parameters calculated by the sponsor (rpm, temperature and the duration).
Thanking you,
ssk
Jaime_R
★★  

Barcelona,
2006-10-12 15:15
(6399 d 17:24 ago)

@ ssk
Posting: # 315
Views: 7,037
 

 sample centrifugation

Hi ssk!

❝ What does the g stand for.


g: approximately 9.81 m/s².
It's good practice to state something like '1500 × g' (1500times acceleration due to earth's gravity) in order to be able to compare different centrifuges.
Stating rpm alone means nothing (since the diameter of the rotor is not given).

For a conversion have a look at this document or this nomogram.

❝ What is the standard temperature at which the centrifugation usually done.


Cooled ;-)
Actually you don't know the temperature of the sample, but it's quite common to set the temperature of the centrifuge to +4 °C.

❝ How are these parameters calculated by the sponsor (rpm,temperature and the duration).


As already stated it should allow for easy separation of plasma.
I don't believe that sponsors do calculate anything - mainly they are skillful in 'copy-and-paste'... ;-)

Regards, Jaime
UA Flag
Activity
 Admin contact
22,988 posts in 4,825 threads, 1,654 registered users;
96 visitors (0 registered, 96 guests [including 5 identified bots]).
Forum time: 08:40 CEST (Europe/Vienna)

The whole purpose of education is
to turn mirrors into windows.    Sydney J. Harris

The Bioequivalence and Bioavailability Forum is hosted by
BEBAC Ing. Helmut Schütz
HTML5