 ICH guideline M10 on BA method validation [Regulatives / Guidelines]

Dear all experts,

Referring to ICH guideline M10 on BA method validation step2b (line no.413-414, p.11), CC and QCs should be prepared from separate stock solution in order to avoid biased estimations which are not related to the analytical performance of the method.

Can you explain how separate stock solution can avoid biased estimations? and what are the bias estimations that happen regarding this sentence, if we use the same stock solution ?

Thank you so much

PSU
Helmut
★★★
avatar
Homepage
Vienna, Austria,
2022-05-04 16:48
(694 d 03:46 ago)

@ PSU
Posting: # 22960
Views: 1,237
 

 Stock solutions for CCs and QCs

Hi PSU,

❝ Referring to ICH guideline M10 on BA method validation step2b …


Not new. Similar in the EMA’s guideline of 2011 and the FDA’s guidance of 2018.

❝ Can you explain how separate stock solution can avoid biased estimations?


If you screw up in preparing either the stock solution for calibrators or QCs, likely many batches will fail.

❝ and what are the bias estimations that happen regarding this sentence, if we use the same stock solution ?


If you use different ones, you demonstrate that you can utilize the analytical balance, volumetric flasks, and pipettes properly. :-D

Dif-tor heh smusma 🖖🏼 Довге життя Україна! [image]
Helmut Schütz
[image]

The quality of responses received is directly proportional to the quality of the question asked. 🚮
Science Quotes
paulhurleyuk
☆    

NE UK,
2022-05-05 13:47
(693 d 06:48 ago)

@ PSU
Posting: # 22962
Views: 1,183
 

 ICH guideline M10 on BA method validation

❝ Can you explain how separate stock solution can avoid biased estimations? and what are the bias estimations that happen regarding this sentence, if we use the same stock solution ?


If you just used one weighing, and weighed too little material you would decrease the signal to noise, and potentially make the assay appear less precise/accurate. Conversely weighing too much material would increase signal to noise and make the assay appear more precise. Having two weighing's makes it more likely to spot that issue.

Paul.
UA Flag
Activity
 Admin contact
22,957 posts in 4,819 threads, 1,636 registered users;
88 visitors (0 registered, 88 guests [including 7 identified bots]).
Forum time: 19:35 CET (Europe/Vienna)

Nothing shows a lack of mathematical education more
than an overly precise calculation.    Carl Friedrich Gauß

The Bioequivalence and Bioavailability Forum is hosted by
BEBAC Ing. Helmut Schütz
HTML5