Hello all members
For a bio analytical method how relevant is the evaluation of dead volume or void volume
For a method having very short retention time within void volume what should be the approach
Is it ok and should the evaluation of dead / void volume be ignored as it is estimated by LCMSMS technique
I am new to this area and hence request opinion of on this.
Helmut
★★★
avatar
Homepage
Vienna, Austria,
2021-10-03 23:14
(927 d 13:26 ago)

@ Bioanalytical
Posting: # 22617
Views: 1,530
 

 dead / void volume

Hi Bioanalytical,

❝ For a bio analytical method how relevant is the evaluation of dead volume or void volume


Interesting1 but not relevant as long the method passes all criteria in method validation.

❝ For a method having very short retention time within void volume what should be the approach


All the usual stuff: Pumps and mixing systems with low volumes, low injection volume, narrow bore tubings as short as possible, correct column nuts and ferrules, clean frits, no dead volume on top of aged column(s)…

❝ Is it ok and should the evaluation of dead / void volume be ignored as it is estimated by LCMSMS technique


See above and this goody.2


  1. Hint: Thiourea SC(NH2)2 has practically no retention on reverse phase and can be used to estimate t0 of the entire system (including the column).
  2. Dolan JW. Extracolumn Effects. LCGC North America. 2008; 26(11): 1092–8. [image] Open access.

Dif-tor heh smusma 🖖🏼 Довге життя Україна! [image]
Helmut Schütz
[image]

The quality of responses received is directly proportional to the quality of the question asked. 🚮
Science Quotes
UA Flag
Activity
 Admin contact
22,986 posts in 4,823 threads, 1,663 registered users;
62 visitors (0 registered, 62 guests [including 4 identified bots]).
Forum time: 12:41 CEST (Europe/Vienna)

Art is “I”; science is “we”.    Claude Bernard

The Bioequivalence and Bioavailability Forum is hosted by
BEBAC Ing. Helmut Schütz
HTML5