avinash
Junior

India,
2018-11-13 12:30

Posting: # 19595
Views: 882
 

 Selectivity [Bioanalytics]

Dear Members,

In Bioanalytical Method Validation - Guidance for Industry - May 2018, For Selectivity parameter one of the acceptance criteria is mentioned as “Blank and zero calibrators should be free of interference at the retention times of the analyte(s) and the IS”.

Which means? Should we run Selectivity samples with calibrators?

Practice is after Specificity and Selectivity, pooling of matrix lots and spiking of calibrators and quality control samples.

Please opine.

Thanks & regards,
Avinash Jain M V


Edit: Category changed; see also this post #1[Helmut]
Ohlbe
Hero

France,
2018-11-13 13:56

@ avinash
Posting: # 19596
Views: 752
 

 Selectivity

Dear Avinash,

» Which means? Should we run Selectivity samples with calibrators?

Not necessarily a full calibration curve, but at least samples spiked at the LLOQ, so that you can compare the signal in the blanks to the signal at the LLOQ.

Regards
Ohlbe
avinash
Junior

India,
2018-11-14 04:35

@ Ohlbe
Posting: # 19599
Views: 716
 

 Selectivity

Dear Ohlbe,

Thank you for the response.

Regards,
Avinash
chandru
Junior

India,
2019-01-01 12:16

@ Ohlbe
Posting: # 19720
Views: 378
 

 Selectivity

Dear Ohlbe,

» Not necessarily a full calibration curve, but at least samples spiked at the LLOQ, so that you can compare the signal in the blanks to the signal at the LLOQ.

For Selectivity experiment whether only 8 blanks and 8 LLOQ is sufficient.

Because even in our company we have changed the procedure of performing Selectivity by running this samples against freshly spiked Calibration and quality control samples.

Regards

Chandru
Activity
 Thread view
Bioequivalence and Bioavailability Forum |  Admin contact
19,271 posts in 4,099 threads, 1,315 registered users;
online 19 (1 registered, 18 guests [including 13 identified bots]).

It’s easy to lie with statistics;
it is easier to lie without them.    Frederick Mosteller

The BIOEQUIVALENCE / BIOAVAILABILITY FORUM is hosted by
BEBAC Ing. Helmut Schütz
HTML5