Beholder ★ Russia, 2018-10-18 11:08 (1988 d 12:24 ago) Posting: # 19464 Views: 4,767 |
|
Hi to everybody! Recently (IMHO, too late for me), I have revealed that EEU BEQ GL has interesting note in validation chapter which is not presented in EU BEQ GL, namely: “Lower limit of quantification 14. The lower limit of quantification (LLOQ) is the lowest concentration of analyte in a sample that can be reliably quantified with acceptable accuracy and precision. The LLOQ is considered to be the smallest standard calibration sample (as indicated in the “Accuracy” and “Precision” subsections of this section). In this case, the signal of the analyte from the sample with the LLOQ must be not less than 5 times greater than the value of the blank sample. The LLOQ cannot be higher than 5% of Cmax (the minimum Cmax value from the entire sample of subjects) ).” EU GL states the same but without blue text. So, IMHO:
What do you do in case when LLOQ higher than 5% mean Cmax? I mean in EU countries. What is your opinion, where it could came from? Any old EU BEQ rule? Im quite surprised that the sentence about minimal Cmax was not excluded during public consultations. Edit: EEU GL linked. [Helmut] — Best regards Beholder |
Helmut ★★★ Vienna, Austria, 2018-10-18 14:28 (1988 d 09:04 ago) @ Beholder Posting: # 19468 Views: 4,366 |
|
Hi Beholder, first of all congratulations for mastering the forum’s tricky syntax of nested lists. ❝ […] I have revealed that EEU BEQ GL has interesting note in validation chapter which is not presented in EU BEQ GL, namely: ❝ “[…] The LLOQ cannot be higher than 5% of Cmax (the minimum Cmax value from the entire sample of subjects) ).” AFAIK, the Russian language leaves a lot of space for interpretation. Is the common meaning of »не должен« really equivalent to “cannot” or “must not”? Since I don’t speak Russian, I have to rely on A(un)I: , Yandex, and Bing translate it as “should not” – which is exactly the phrase used in the EMA’s GL. If it is mandatory (your #1), very stupid. #2 is a show-stopper. I can only tell you what is done in studies for the EMA. Use an estimate of the lowest expected Cmax and base the LLOQ on it. Depending on the expected variability CROs commonly opt for 2% or even 1% of the average Cmax. Nerds like me work with modeling, stuff like that (replace t½ with Cmax), or simulations. Of course, if individual data are available (pilot or other studies), better. Never saw revalidation. If you are cautious, exclude the subject. If that doesn’t occur too often, the loss in power is small. — Dif-tor heh smusma 🖖🏼 Довге життя Україна! Helmut Schütz The quality of responses received is directly proportional to the quality of the question asked. 🚮 Science Quotes |
Beholder ★ Russia, 2018-10-18 15:59 (1988 d 07:33 ago) (edited by Beholder on 2018-10-19 08:34) @ Helmut Posting: # 19469 Views: 4,266 |
|
Hi Helmut, ❝ first of all congratulations for mastering the forum’s tricky syntax of nested lists. THX. I reached it finally. Years passed. ❝ AFAIK, the Russian language leaves a lot of space for interpretation. Is the common meaning of »не должен« really equivalent to “cannot” or “must not”? Since I don’t speak Russian, I have to rely on A(un)I: , Yandex, and Bing translate it as “should not” – which is exactly the phrase used in the EMA’s GL. You are right regarding the meaning. Yes, it is rather "should not". But looks like I was not clear enough. I was talking about text which is marked with blue: "...(the minimum Cmax value from the entire sample of subjects)". It is one story when we take 5% from average Cmax, but it is another story when we take 5% from minimum value of Cmax in volunteers cohort. My question was caused by an article where such information was presented: "According to the literature, the minimum arithmetic average Cmax is 1,479.61 ng / ml (when taking atazanavir at a dose of 300 mg) [15]. Thus, the LLOQ should not exceed 74 ng / ml. However, it should be taken into account that, in accordance with the requirements of the EAEU, the LLOQ is calculated based on the minimum value of Cmax from the entire sample of subjects [4]. In the analyzed studies, the minimum value of Cmax was 60.8 ng / ml, while the calculated value of LLOQ is 3 ng / ml." Edited: link to the article updated. — Best regards Beholder |
Helmut ★★★ Vienna, Austria, 2018-10-18 20:20 (1988 d 03:13 ago) @ Beholder Posting: # 19472 Views: 4,232 |
|
Hi Beholder, ❝ You are right regarding the meaning. Yes, it is rather "should not". But looks like I was not clear enough. Oh, you were. ❝ My question was caused by an article […] The link is not working. Can you check? ❝ According to the literature, the minimum arithmetic average Cmax is 1,479.61 ng / ml. Arithmetic means of concentrations are stupid. Reporting concentrations to six significant digits even more. I assume that the “minimum arithmetic average Cmax” was calculated from different studies; each with a dose of 300 mg. ❝ Thus, the LLOQ should not exceed 74 ng / ml. OK, 5% of a doubtful value (likely the geometric mean is lower). Setting the LLOQ based on (any) average is stupid as well. ❝ However, it should be taken into account that, in accordance with the requirements of the EAEU, the LLOQ is calculated based on the minimum value of Cmax from the entire sample of subjects. Makes sense. (Lack of) carry-over has to be assessed on a subject-level. See what I wrote at the end of my previous post. We cannot get an unbiased treatment-effect in the presence of true carryover. Hence, we have to avoid CO by design. One method is to assess pre-dose concentrations in periods >1. And yes, make sure that we are able to measure 5% of every subject’s Cmax after each treatment. ❝ In the analyzed studies, the minimum value of Cmax was 60.8 ng / ml, Wow, that’s extreme! Not clear to me whether the authors are referring to the same dose of 300 mg in all studies. ❝ while the calculated value of LLOQ is 3 ng / ml." 5% again. If the same dose was used in all studies (I doubt it) it would be a nightmare:
PS: It can be even worse. Sometimes in studies in patients (mainly phase III) different LLOQs are used. Might be caused by limited sample volumes or interactions with co-medications. The world is not ideal. — Dif-tor heh smusma 🖖🏼 Довге життя Україна! Helmut Schütz The quality of responses received is directly proportional to the quality of the question asked. 🚮 Science Quotes |
mittyri ★★ Russia, 2018-10-18 21:50 (1988 d 01:43 ago) @ Helmut Posting: # 19473 Views: 4,126 |
|
Hi Helmut and Beholder, Let's take a hypothetical example Due to high intersubject variability we got some periods for some (not many) subjects less than 5% of LLOQ. 1. Would you ask analytical lab to revalidate the method and reanalyze those profiles again or since analytical part is over, there's no way to change the method when playing part starts (Cmax estimation is PK part, right?) 2. Do you think the reasons to exclude that subjects are feasible enough? Have you ever been asked to do that? For me this is mind-blowing since the applicants should invest money not only in dropouts, but in extreme metabolizes too! The power loss for replicate designs is more significant (double the subject data) PS: moreover, that criterion (Cmax< 20*LLOQ) is not listed in EMA GL as a reason for subject exclusion — Kind regards, Mittyri |
Beholder ★ Russia, 2018-10-19 11:43 (1987 d 11:50 ago) @ mittyri Posting: # 19476 Views: 4,116 |
|
Hi Helmut and mittyri! ❝ The link is not working. Can you check? Please check now. Should work. THX to Moderator. ❝ Reporting concentrations to six significant digits even more. I assume that the “minimum arithmetic average Cmax” was calculated from different studies; each with a dose of 300 mg. Looks like the value 1,479.61 ng / ml was taken from PAR of one big Indian pharmaceutical company. I’m not sure that it is average of different studies. BTW, what about seven and even eight significant digits (see PAR, pg 20)? ❝ In the analyzed studies, the minimum value of Cmax was 60.8 ng / ml, IMHO, that is actually the case. According to the article (Table 1), only 300 mg and 400 mg doses (not strengths) were included to the analysis. So looks like 60.8 ng / ml is an outlier due to extreme fast metabolizing? — Best regards Beholder |