Is it mandatiry to prepare differnt stock solution to prepare calibration curve and quality control sample ? if yes..... Why?

Regrards
pramod
Ohlbe
★★★

France,
2007-10-12 20:16
(6813 d 04:54 ago)

@ pb070681
Posting: # 1184
Views: 4,533
 

 Stock preparation

Dear Pramod,

According to the http://www.fda.gov/cder/guidance/4252fnl.pdf FDA guideline on bioanalytical method validation, page 15:

Standards and QC samples can be prepared from the same spiking stock solution, provided the solution stability and accuracy have been verified.

You can use the same stock solution for calibration and QC samples. But you have to show that the stock solution was prepared correctly and that its concentration is accurate. Which is usually done by comparing it to another, independently prepared stock solution...

Regards
Ohlbe

Edit: Link corrected for FDA’s new site. [Helmut]
chiragkhatri
★    

India,
2007-10-17 12:37
(6808 d 12:33 ago)

@ Ohlbe
Posting: # 1207
Views: 4,447
 

 Stock preparation

Dear Pramod,

I agree to Ohlbe, but if u go through FDAs bioresearch monitoring programme (link)

Access attachment A, Part III ( ANalytical data and operations), go to Section E and point 8 states that FDA inspectors do ask for, whether the control samples and calibrations standards were prepared from different stock or not

Regards

Chirag

--
Edit: Full quote removed. Please see this post! [HS]
Ohlbe
★★★

France,
2007-10-18 20:10
(6807 d 05:00 ago)

@ chiragkhatri
Posting: # 1214
Views: 4,358
 

 Stock preparation

Dear Chiragkhatri,

❝ Section E and point 8 states that FDA inspectors do ask for, whether the

❝ control samples and calibrations standards were prepared from different

❝ stock or not


This document is dated October 1999, whis is before the current FDA guideline of May 2001 on bioanalytical method validation. It was prepared from the draft guideline of 1998 and not updated since.

Regards
Ohlbe
UA Flag
Activity
 Admin contact
23,653 posts in 4,991 threads, 1,570 registered users;
325 visitors (0 registered, 325 guests [including 152 identified bots]).
Forum time: 01:11 CEST (Europe/Vienna)

Garbage in, garbage out.
It’s very very simple.    Anders Fuglsang

The Bioequivalence and Bioavailability Forum is hosted by
BEBAC Ing. Helmut Schütz
HTML5