ratnakar1811 ★ India, 2013-03-05 09:30 (4062 d 16:51 ago) Posting: # 10151 Views: 31,844 |
|
Dear All, As per EMA guideline AUC(0-t) should cover at least 80% of AUC(0-∞) but in one of my study I have got this percentage more than 20% and the % in the study is around 31%. Is there any way by which I can justify this issue? Anybody has any such experience with the European regulators’? Your views are highly appreciated. Best Regards, Ratnakar |
d_labes ★★★ Berlin, Germany, 2013-03-05 12:12 (4062 d 14:08 ago) @ ratnakar1811 Posting: # 10153 Views: 24,878 |
|
Dear ratnakar, the 80% coverage of AUC(0-tlast) in relation to AUC(0-inf) is specified further in the EMA guidance under the heading "Reasons for exclusion": "As stated in section 4.1.4, AUC(0-t) should cover at least 80% of AUC(0-∞). Subjects should not be excluded from the statistical analysis if AUC(0-t) covers less than 80% of AUC(0-∞), but if the percentage is less than 80% in more than 20% of the observations then the validity of the study may need to be discussed." I interpret the 20% of observations as 20% of AUC values i.e. 2xsubjects in a classical 2x2x2 crossover. Don't ask me how the validity of the study may be discussed. Duno. — Regards, Detlew |
ratnakar1811 ★ India, 2013-03-06 09:50 (4061 d 16:30 ago) @ d_labes Posting: # 10157 Views: 24,602 |
|
Dear d_labes/forum members, Yes, in more than 20% of observations AUCinf has been extrapolated for more than 20% and i have received a query from EMA to discuss the validity of the study. I am just thinking what all points should be considered for the justification and further is there any way by which it can be justified? Best Regards, Ratnakar |
ElMaestro ★★★ Denmark, 2013-03-06 12:10 (4061 d 14:10 ago) @ ratnakar1811 Posting: # 10158 Views: 24,602 |
|
Hello Ratnakar, ❝ Yes, in more than 20% of observations AUCinf has been extrapolated for more than 20% and i have received a query from EMA to discuss the validity of the study. From EMA? I guess you mean from one or more of the national agencies? ❝ I am just thinking what all points should be considered for the justification and further is there any way by which it can be justified? Try and tell why this happened in the first place. If it was not intended then I guess your study could be argued to be slightly badly planned for which there are no obvious and acceptable excuses. On the other hand if you deliberately shortened sampling time for some reason (PI opinion, ethics, ADME-tox properties) then please give a little further info and let's see. — Pass or fail! ElMaestro |
Helmut ★★★ Vienna, Austria, 2013-03-06 15:35 (4061 d 10:45 ago) @ ElMaestro Posting: # 10162 Views: 24,560 |
|
Hi ElMaestro! ❝ […] if you deliberately shortened sampling time for some reason (PI opinion, ethics, ADME-tox properties) then please give a little further info and let's see. No sure what you mean by ↑ that. Shortening the sampling wouldn’t help. — Dif-tor heh smusma 🖖🏼 Довге життя Україна! Helmut Schütz The quality of responses received is directly proportional to the quality of the question asked. 🚮 Science Quotes |
ElMaestro ★★★ Denmark, 2013-03-06 17:40 (4061 d 08:40 ago) @ Helmut Posting: # 10164 Views: 24,474 |
|
Hi HS, ❝ No sure what you mean by ↑ that. Shortening the sampling wouldn’t help. Sorry, this was a dumb formulation, you are absolutely right. If the tox properties are unfavourable then one might need to dose so low that the drug is adme'd away by the time one for some reason or other would like to measure. Erm... Heh... even this sounds dumb now. Does bebac offer language courses? PS: I don't mean to indicate this would be an excuse in itself, rather it could be a cause. — Pass or fail! ElMaestro |
Helmut ★★★ Vienna, Austria, 2013-03-06 18:19 (4061 d 08:01 ago) @ ElMaestro Posting: # 10165 Views: 24,450 |
|
Servus! ❝ Does bebac offer language courses? Depends. I’m far from being competent in standard German although I could offer counseling in its Austrian variant. The only language I really master is Viennese. Are you interested? Free lodging included. Visiting my favorite Heurigen mandatory. — Dif-tor heh smusma 🖖🏼 Довге життя Україна! Helmut Schütz The quality of responses received is directly proportional to the quality of the question asked. 🚮 Science Quotes |
ElMaestro ★★★ Denmark, 2013-03-07 18:19 (4060 d 08:01 ago) @ Helmut Posting: # 10171 Views: 24,409 |
|
Hi H., ❝ Depends. I’m far from being competent in standard German although I could offer counseling in its Austrian variant. The only language I really master is Viennese. Are you interested? Free lodging included. Visiting my favorite Heurigen mandatory. Austrian German is a funny kind of Mickey Mouse language. I like the sound of it. Let's go and visit that place of yours next time I am in the area. Oh and haven't you heard? I converted to the Vegetophysticiologists. We only drink cabbage juice and eat mung beans. — Pass or fail! ElMaestro |
Helmut ★★★ Vienna, Austria, 2013-03-07 18:26 (4060 d 07:54 ago) @ ElMaestro Posting: # 10172 Views: 24,164 |
|
Hi ElMaestro! ❝ Austrian German is a funny kind of Mickey Mouse language. I like the sound of it. Yep. For me it’s always amazing that native speakers of English love the Austrian’s accent whereas they hate the Kraut’s. Both sound awful to me. ❝ […] I converted to the Vegetophysticiologists. We only drink cabbage juice and eat mung beans. Go to Gujarat. Next step Breatharianism. — Dif-tor heh smusma 🖖🏼 Довге життя Україна! Helmut Schütz The quality of responses received is directly proportional to the quality of the question asked. 🚮 Science Quotes |
Helmut ★★★ Vienna, Austria, 2013-03-06 15:30 (4061 d 10:50 ago) @ ratnakar1811 Posting: # 10161 Views: 24,623 |
|
Dear Ratnakar, EMA’s opinion [sic] that AUCt is a reliable estimate of extent of absorption only if AUCt ≥ 80% of AUC∞ is scientifically questionable – if (!) IR formulations are concerned (absorption practically completed at 2–4× tmax). Therefore, we need more information.
— Dif-tor heh smusma 🖖🏼 Довге життя Україна! Helmut Schütz The quality of responses received is directly proportional to the quality of the question asked. 🚮 Science Quotes |
cakhatri ★ India, 2013-03-10 09:59 (4057 d 16:21 ago) @ Helmut Posting: # 10179 Views: 23,701 |
|
Dear All, This is interesting topic as I too have had similar situation where in the ratio of AUC0-t / AUC0-inf was less than 80% in several cases. Can anyone clear my doubt on this, does this apply to the untransformed or Ln transformed data The ratios were fine with the Ln transformed data. Regards Chirag |
Helmut ★★★ Vienna, Austria, 2013-03-11 02:17 (4057 d 00:04 ago) @ cakhatri Posting: # 10181 Views: 23,944 |
|
Dear Chirag, ❝ […] does this apply to the untransformed or Ln transformed data ❝ ❝ The ratios were fine with the Ln transformed data. Raw data! You need the ln-transformed data only in the BE-statistics. You would have to calculate the difference of logs and backtransform to get % covered. Example: AUCt 375, AUC∞ 500, covered: 100×375/500=75%. — Dif-tor heh smusma 🖖🏼 Довге життя Україна! Helmut Schütz The quality of responses received is directly proportional to the quality of the question asked. 🚮 Science Quotes |
Helmut ★★★ Vienna, Austria, 2013-03-11 04:17 (4056 d 22:03 ago) @ Helmut Posting: # 10182 Views: 24,253 |
|
Dear all, replying myself to clarify things. Starter: Scheerans et al. (2008).1 tmax in a monoexponential model is the root of the first derivative of the concentration-time curve and given by \(\small{t_{max}=\log(k_a / k_{el}) / (k_a-k_{el})}\) The inflection point tinpt where the sign of the curvature changes (or the tangent moves from above to below the curve) is the root of the second derivative and given by\(\small{t_{inpt}=2\cdot \log(k_a / k_{el}) / (k_a-k_{el})}\) If you want to play around yourself I suggest to convert the Bateman-function (with volume, fraction absorbed, and rate constants) to a sum of exponentials, i.e.,\(\small{f(x)=A(e^{-k_{el}\cdot x}-e^{-k_a\cdot x})}\). The first and second derivatives are\(\small{f'(x)=A(-k_{el}\cdot e^{-k_{el}\cdot x}+k_a\cdot e^{-k_a\cdot x})}\) and \(\small{\%\; abs = 100(1-e^{-k_a \cdot x})}\) AUC in the interval [0, a] is given as the definite integral of ƒ(x) by\(\small{A \left ( (e^{k_a \cdot a}-1)/k_a - (e^{k_{el}\cdot a}-1)/k_{el} \right )}\) At tmax 75% are absorbed (AUCt/AUC∞ 25%) and at tinpt already 93.75% (AUCt/AUC∞ 56.25%). When do we have a ratio of 80%? At 3.244 (where 98.89% are absorbed). That’s a very, very conservative value. Note that tinpt equals four absorption half-lives (alternative method for the calculation of % absorbed: 100 – 100/24·t½a = 93.75%). If one wants to be more conservative and follow the logic of the GL (wash-out, saturation to steady state) and use five half-lives one could stop at 2.5.
IMHO, the cut-off of 80% is scientifically not justified for monoexponential models (including ones with a lag-time: homework). Would only make sense if we would use AUCs corrected by kel,4 which seems not to be acceptable by EMA (at least). It’s strange to judge the validity of a study based on elimination rather than essentially complete absorption. Unfortunately it is not that easy for multicompartimental models – or even worse – profiles with multiple peaks, though absorption will follow the same time course. MR (CR in particular) is another cup of tea. We might be in flip-flop limbo and personally I would not use AUCt as the main metric at all. I’m curious what the draft MR GL will tell us… I will propose AUC∞ with an extrapolated area of well below 20%. @Ratnakar: If you want help it would be a good idea to answer all of the questions in my previous post (see the Forum’s policy).
— Dif-tor heh smusma 🖖🏼 Довге життя Україна! Helmut Schütz The quality of responses received is directly proportional to the quality of the question asked. 🚮 Science Quotes |
qualityassurance ☆ 2020-04-23 16:11 (1456 d 11:09 ago) @ Helmut Posting: # 21345 Views: 16,108 |
|
Dear Helmut and all, Hope everyone is good and safe. The same query raised by MoH and i want to answer. Could you help me? ❝ • IR or MR formulation? If MR, DR or CR? ❝ • Pilot studies performed? If yes, outcome? ❝ • Sampling planned based on average t½ from the literature? If yes, please give x±SD and sample size in your study. ❝ • Observed t½ (x±SD, range)? For test (9.2 ±2.96, 3.22-17.0) ❝ • tmax, tlast? For test: Tmax: 0.50, Tlast: 23.91 Forgive me for the formatting. I tried my best. Regards. |
Helmut ★★★ Vienna, Austria, 2020-04-23 19:32 (1456 d 07:48 ago) @ qualityassurance Posting: # 21347 Views: 15,984 |
|
Hi qualityassurance, that’s an interesting case. ❝ ❝ • Observed t½ (x±SD, range)? ❝ For reference (10.03 ±4.09, 3.5-19.0) ❝ For test (9.2 ±2.96, 3.22-17.0) ❝ ❝ ❝ • tmax, tlast? ❝ For reference: Tmax: 0.67, Tlast: 28.25 ❝ For test: Tmax: 0.50, Tlast: 23.91 I played around with your data. Such an early tmax means that this IR product acts practically like a solution (absorption t½ 10–15 minutes). However, there is an extreme between-subject variability in elimination. That’s why you missed the 80% rule in some subjects. In retrospect it would have been better to opt for truncated AUC0–72. Then you would not have to worry about extrapolation. However, think about the subject line of the post. At 2×tmax absorption is practically complete (~96.12%). You could even take a very conservative approach with 4×tmax (~99.85%). We are interested in comparing the performance of the formulations, not to assess its well-known PK, right? Hence, scientifically the large extrapolated AUC you see in some subjects is not relevant at all. Kamal Midha showed this nice example of partial AUCs. Once absorption is essentially complete, later T/R-ratios are extremely stable – only their variability increases. If the assessor prefers to “tick-the-box-in-the-form”, cards are stacked against you. Sorry. ❝ Forgive me for the formatting. I tried my best. No worries; I edited your post a bit. — Dif-tor heh smusma 🖖🏼 Довге життя Україна! Helmut Schütz The quality of responses received is directly proportional to the quality of the question asked. 🚮 Science Quotes |
Achievwin ★ US, 2020-05-08 00:59 (1442 d 02:21 ago) @ Helmut Posting: # 21404 Views: 14,540 |
|
Good discussions: ❝ ❝ For reference: Tmax: 0.67, Tlast: 28.25 ❝ ❝ For test: Tmax: 0.50, Tlast: 23.91 Aren't we lost in pure numbers and forgot to look at the individual curves? In order to have accurate AUC estimations you need to have at least two measurable concentrations (non-BQL) before Tmax and as many time points as possible on either side of Tmax. With 067 and 0.50 hours as Tmax how many blood draws you can practically take (assuming this is an oral formulation) may be we have to go to IV blood sampling schedule. I saw somewhere blood samples were taken 24 and 48 hrs (assuming these are ambulatory samples there will be lot of time variation.), Hardly we completed 1 or 2 half lives. and we expect to cover 80% AUC? Even assuming we covered AUC there is always some bad apples, in my experience due to some bad data points in the extremes you may tend to have all kinds of elimination patterns (positive slope, bad two many missed blood draws or BQL values… so it is not surprising that AUC% is less than 80% in more than 30% subjects what is that 30% in 12 or 18 subjects? If i am designing study I take following precautions (everyone knowns these things but I am listing)
Then a priori in the protocol SAP include limitations on reporting AUCinf and %Extrp.
Hope this helps. Edit: Please don’t shout! [Helmut] |
Helmut ★★★ Vienna, Austria, 2020-05-08 01:54 (1442 d 01:26 ago) @ Achievwin Posting: # 21405 Views: 14,650 |
|
Hi Achievwin, ❝ Then a priori in the protocol SAP include limitations on reporting AUCinf and %Extrp. ❝ 1. R2>0.800 Why this – arbitrary – number? Given, I have seen it in lots of SAPs. \(\small{R^2}\) (strongly!) depends on the number of time points \(n\) (see this rather old thread). \(\small{R_{\textrm{adj}}^{2}}\) is better though still not independent from \(n\) (that’s a misconception sold by software vendors). If clearance is variable and/or the analytical variability is high, a low correlation may be perfectly fine, whereas in the opposite case even 0.9 may indicate a poor fit. Unless you have data of a previous study with the same analytical method, IMHO, a pre-specified cut-off does not make sense. Furthermore, never trust in results of a silicon-based life-form. Visual inspection of the fit is mandatory. PS: 0.8 with three significant digits: 0.8005 is OK and 0.8004 not. Really? — Dif-tor heh smusma 🖖🏼 Довге життя Україна! Helmut Schütz The quality of responses received is directly proportional to the quality of the question asked. 🚮 Science Quotes |
Brus ★ Spain, 2018-11-20 14:01 (1976 d 12:19 ago) (edited by mittyri on 2018-11-20 14:25) @ ratnakar1811 Posting: # 19644 Views: 19,285 |
|
Dear Ratnakar, How did you justify the clarification? How can it be justified that the extrapolated AUC is greater than 20%? Best Regards, Brus Edit: Full quote removed. Please delete everything from the text of the original poster which is not necessary in understanding your answer; see also this post #5! [Mittyri] |
Astea ★★ Russia, 2020-06-05 21:05 (1413 d 06:15 ago) @ Brus Posting: # 21501 Views: 12,981 |
|
Dear Friends! I've got I question concerned to this issue. Until now I thought that the rule of 80% AUCinf was invented in order to demonstrate that the total duration of sampling is sufficient. But consider a following case: IR drug with T1/2=12-18 hours (according to literature data). The last sampling time in the protocol was choosen to be 48 hours with a large step before, like: ..8 hours, 12 hours, 24 hours and 48 hours. For several subjects the PK curves were like this: The concentration in the last point was below LLOQ, however AUClast didn't cover 80% of AUCinf. So I may conclude, although the rule was not followed, the duration of the sampling time was sufficient. In this case the rule may indicate: too large distance between sample time points (48-24=24) that is an error in study planning or too large LLOQ. Am I right in this conclusion? How can regulators interpret this issue? — "Being in minority, even a minority of one, did not make you mad" |
Helmut ★★★ Vienna, Austria, 2020-06-06 14:02 (1412 d 13:18 ago) @ Astea Posting: # 21502 Views: 12,826 |
|
Hi Nastia, ❝ Until now I thought that the rule of 80% AUCinf was invented in order to demonstrate that the total duration of sampling is sufficient. Read the entire thread again. I’m referring to the EMA’s guideline. Too lazy to Google-translate the ones in Russian. The phrase in the EMA’s guideline “The sampling schedule should also cover the plasma concentration time curve long enough to provide a reliable estimate of the extent of exposure…” is nonsense. Why it was invented [sic] is beyond my intellectual reach.
❝ But consider a following case: […] The tmax is ≤ 1 h and hence, t½,a ≤ 30 minutes. It means that at four hours ≥ 99.6% were already absorbed. That’s what we are interested in. What follows shows that we have a two-compartment model (and possibly additional enterohepatic recycling). Nice to know, but already described by the innovator, right? Any (‼) AUC after 2–4times tmax is a reliable estimate of the extent of exposure. Full stop. ❝ So I may conclude, although the rule was not followed, the duration of the sampling time was sufficient. In this case the rule may indicate: too large distance between sample time points (48-24=24) that is an error in study planning or too large LLOQ. Am I right in this conclusion? EMA: “Subjects should not be excluded from the statistical analysis if AUC0–t covers less than 80% of AUC0–∞, but if the percentage is less than 80% in more than 20% of the observations then the validity of the study may need to be discussed.” OK, discuss it. It’s quite possible that even with an additional sampling point at 36 h (≥ LLOQ) you would have ended with < 80% AUC0–∞ as well. For an IR formulation you could have sampled longer and with AUC0–72 this entire extrapolation business would disappear. Let’s play the devil’s advocate. IR formulation, drug has a fast distribution and long elimination, the LLOQ is sufficient to measure all samples at the last sampling time t; two studies:
❝ How can regulators interpret this issue? Can or will? Regulators have the whip hand and therefore, can do whatever they like. If the assessor is a “checkbox-bureaucrat”, cards are stacked against you. If the assessor is a scientist, IMHO, good chances. Science should always be the basis
— Dif-tor heh smusma 🖖🏼 Довге життя Україна! Helmut Schütz The quality of responses received is directly proportional to the quality of the question asked. 🚮 Science Quotes |
Astea ★★ Russia, 2020-06-06 23:11 (1412 d 04:09 ago) @ Helmut Posting: # 21503 Views: 12,624 |
|
Dear Helmut! I'm grateful for your reply! ❝ Read the entire thread again. Ok, my HW was not perfect No I reread it. Still don't understand what to argue to the assessors... By the way, there is a typo in f' (the graph is correct), first term should be with minus: \(\small{f'(x)=A(-k_{el}\cdot e^{-k_{el}\cdot x}+k_a\cdot e^{-k_a\cdot x})}\)} and for AUC as the integral from 0 to a, "a" should go instead of x:\(\small{A \left ( (e^{k_a \cdot a}-1)/k_a - (e^{k_{el}\cdot a}-1)/k_{el} \right )}\) ❝ I’m referring to the EMA’s guideline. Too lazy to Google-translate the ones in Russian. Do not bother yourself, EAEC experts had already translated it: "Количество отобранных образцов также должно быть достаточным, чтобы обеспечить надежную оценку длительности экспозиции. Это достигается, когда AUC(0–t) перекрывает не менее 80 процентов от AUC(0–∞)." Though the statement is not scientifically based it is still in the list of regulator's requirements. So in case if some problems with the "80% covering rule" we have to justify somehow that the validity of the study should not be called into question - that's what I've asked about. ❝ Therefore, the FDA does not have such a bizarre “AUC0–t ≥ 80% AUC0–∞ rule By the way, was this question ever discussed on the conferences on harmonization? ❝ ...71 h → AUC0–71... Usually the last sampling point divides completely by 12, so I may expect 48 to be the last sampling point or may be 60, but who is so crazy to leave 71 as the last point? But there arising another interesting situation: imagine there is a IR drug with T1/2 equals to 18 hours. 4*T1/2=72 so we can leave 72 hours as the last sampling point. Which of the two parameters AUClast or AUC72 should be used to choose the best strategy to confirm BE in this case? At first sight they should be equal but this is not true. The NCA software (like Phoenix) uses AUCpartial to calculate AUC72, so for 72 hours it would be equal to AUCall but not AUClast! For some subjects the last sample could be below LLOQ, for them AUCall>AUClast with an area of triangle. — "Being in minority, even a minority of one, did not make you mad" |
Helmut ★★★ Vienna, Austria, 2020-06-07 14:41 (1411 d 12:39 ago) @ Astea Posting: # 21504 Views: 12,289 |
|
Hi Nastia, ❝ […] Still don't understand what to argue to the assessors... You could refer to the papers of Scheerans et al. and Midha et al. (see this post). Again: Scientifically there is no point on measuring that long (IR only, of course) and the regulatory “≥80%-rule” is made out of thin air. As you rightly stated previously: It’s an invention (Who is to blame – the CHMP’s PKWP?) and lacks any \(\small{\frac{\textsf{evidence}}{\textsf{justification}}}\). If you have a lot of time and nothing better to do, perform simulations. On my to-do list for ages. Currently in the “dead dogs” folder of my machine. Concept: Simulate studies for various PK models (1-, 2-compartments [different ratios of distribution/elimination rate constants], with/without lag-times). Have a range of T/R-ratios (say 0.85–1). Power the studies for AUC0–t (or AUC0–72?). Contaminate the data with missings at the end (mimicking BLQs). Calculate serial partial AUCs starting from tmax to the end (say, pAUC0–2, pAUC0–3, …, pAUC0–72). Assess each one for BE. Record the PE, CV, and fraction of studies passing (post hoc power). Additionally calculate the %RE of the PE to the “true T/R-ratio”. My preliminary results (trivial and already shown by Kem Midha with real data sets): The CV of partial AUCs decreases with time and their PE stabilizes towards the true value pretty early (guess when). In Table 1 of Kem’s 1996 paper he reported results of nine studies (22 analytes – parent and metabolites with t½ 1 – 656 h); sampling up to 1,848 (‼) hours ( hydroxychloroquine) and concluded […] a sampling period of 24 h would have given the same bioequivalence decisions as AUClast […]. The last two sentences:In bioequivalence, the issue is do deal with formulation differences. Once absorption is over, formulation differences no longer apply. Sigh. Regulators, are you reading along?❝ By the way, there is a typo in f' and for AUC … Oh dear! I’m not sooo bad in basic calculus but in translating my original text to MathJax. Original post corrected. THX! ❝ "Количество отобранных образцов также должно быть достаточным, чтобы обеспечить надежную оценку длительности экспозиции. Это достигается, когда AUC(0–t) перекрывает не менее 80 процентов от AUC(0–∞)." ❝ Though the statement is not scientifically based it is still in the list of regulator's requirements. So in case if some problems with the "80% covering rule" we have to justify somehow that the validity of the study should not be called into question - that's what I've asked about. … должно быть ≈ should be or must be? In guidelines commonly “should be” is used. What’s your interpretation of the Russian original? Does it smell of a recommendation or rather a requirement? ❝ ❝ Therefore, the FDA does not have such a bizarre “AUC0–t ≥ 80% AUC0–∞ rule ❝ ❝ By the way, was this question ever discussed on the conferences on harmonization? Not that I recall. For IR products the main discussion (Amsterdam 2015, Rockville 2016) was about whether the FDA \(\small{\frac{\textsf{should}}{\textsf{will}}}\) drop the additional AUC0–∞. ❝ ❝ ...71 h → AUC0–71... ❝ ❝ Usually the last sampling point divides completely by 12, so I may expect 48 to be the last sampling point or may be 60, but who is so crazy to leave 71 as the last point? Yes but you go me wrong. For sure you know the correspondence between Einstein and Bohr about the Copenhagen interpretation of quantum mechanics? My example was a Gedankenexperiment to point out how absurd this requirement can be. Let’s heighten the absurdity. We want to submit to two agencies, A and B, where B accepts a study with a foreign reference product. Both agencies have GLs asking for AUC0–t. We write the SAPs accordingly. Before the study starts, agency A allows AUC0–72. We write an amendment for agency A. Study performed. We have one (!) data set and will have no problems with agency A but likely receive a deficiency letter from agency B. ❝ […] imagine there is a IR drug with T1/2 equals to 18 hours. 4*T1/2=72 so we can leave 72 hours as the last sampling point. Which of the two parameters AUClast or AUC72 should be used to choose the best strategy to confirm BE in this case? Since this is an IR formulation, AUC0–72. Stated in the protocol and no problems with extrapolation. ❝ At first sight they should be equal but this is not true. The NCA software (like Phoenix) uses AUCpartial to calculate AUC72, so for 72 hours it would be equal to AUCall but not AUClast! For some subjects the last sample could be below LLOQ, for them AUCall>AUClast with an area of triangle. AUCall is \(\small{\frac{\textsf{Pharsight’s}}{\textsf{Certara’s}}}\) invention. If you find it in a single textbook about PK, let me now. IMHO, it should go into the waste bin. An example in PHX8.1: Two compartments, half lives: absorption 30 minutes, distribution one hour, elimination 18 hours, lin-up/log-down trapezoidal.
AUCall 2376 if you set the last concentration to zero. IMHO, not a good idea.The value of the last triangle depends on the time interval between the time point of the last measured concentration and the next one with C=0. [Thoughtless comment removed by the author.] — Dif-tor heh smusma 🖖🏼 Довге життя Україна! Helmut Schütz The quality of responses received is directly proportional to the quality of the question asked. 🚮 Science Quotes |
mittyri ★★ Russia, 2020-06-07 22:32 (1411 d 04:48 ago) @ Helmut Posting: # 21505 Views: 11,910 |
|
Hi Helmut, ❝ If you have a lot of time and nothing better to do, perform simulations. On my to-do list for ages. Currently in the “dead dogs” folder of my machine. I had in mind more simplistic approach: shiny app to give points, parameters of 2cpt model and to see what parameters will give ratio less than 80%. I like 'dead dogs', will rename it ❝ … должно быть ≈ should be or must be? ❝ In guidelines commonly “should be” is used. What’s your interpretation of the Russian original? Does it smell of a recommendation or rather a requirement? That's a matter of expert's taste and mood see later: Однако если AUC(0–t) не перекрывает 80 процентов AUC(0–∞) в более чем 20 процентах случаев, следует усомниться в результатах такого исследования. Are you good enough to extinguish doubts from expert's mind? You never know! ❝ Which version of Phoenix are you using? That rule is not changed for decades in Winnonlin. ❝ The value of the last triangle depends on the time interval between the time point of the last measured concentration and the next one with C=0. You would only get a meaningful result if (1) you use the lin-up/log-down trapezoidal and (2) the interval equals t½. Since (2) is never fulfilled for all profiles, forget it. Regarding (1): lin-up/log-down is overriden by linear rule if the end point is non-positive. — Kind regards, Mittyri |
Astea ★★ Russia, 2020-06-08 02:08 (1411 d 01:12 ago) @ mittyri Posting: # 21506 Views: 12,093 |
|
Dear Helmut and Mittyri! ❝ Scientifically there is no point on measuring that long (IR only, of course) and the regulatory “≥80%-rule” is made out of thin air. Do you know a legend about Ehrenfest's parrot? I wish I have that parrot to say: "Aber meine Herren das ist keine Wissenschaft!" every time I get the request to use 80% rule ❝ If you have a lot of time and nothing better to do, perform simulations. "Spare time" are the two very funny words, I wish I knew their meaning. Nevertherless thank you for the point, may be I will investigate it one day. Let all dogs be alive! By the way is not it similar to one used in 1997 by Endrenyi L. et al1? ❝ In guidelines commonly “should be” is used. What’s your interpretation of the Russian original? Does it smell of a recommendation or rather a requirement? As far as I know the war almost starts when russian experts try to translate this word properly. For me "должно быть" sounds like an order. We use "следует" to slightly soften the meaning and "рекомендуется" for the recommendation. ("Marain was a synthetic language, designed to be phonetically and philosophically as expressive as the pan-human speech apparatus and the pan-human brain would allow") ❝ Let’s heighten the absurdity. We want to submit to two agencies, A and B, where B accepts a study with a foreign reference product. Both agencies have GLs asking for AUC0–t. We write the SAPs accordingly. Before the study starts, agency A allows AUC0–72. We write an amendment for agency A. Study performed. We have one (!) data set and will have no problems with agency A but likely receive a deficiency letter from agency B. This situation is not that absurd. I often deal with protocols where the definition of AUC is ambiguously: in the one part of the protocol they are talking about AUC0–72 and in the other - about AUC0–t (AUClast). Noone understands how it can differ if t=72 hours. About AUC0–t (AUCall) see also Oishi M. et al.2 ❝ Which version of Phoenix are you using? The last triangle is not added in the second data set (containing the 72 h time point and BQL). You would only get Yes, you guessed, that was a stupid case of zero as LLOQ. I use 8.2. As Mittyri pointed out upper, the calculation doesn't depend from the version. Anyway one may note that AUC0-72 and AUC0-t when t=72 are not always equal (especially considering the cases with deviations in sampling time).
— "Being in minority, even a minority of one, did not make you mad" |
Helmut ★★★ Vienna, Austria, 2020-06-08 03:04 (1411 d 00:16 ago) @ Astea Posting: # 21507 Views: 11,990 |
|
Hi Nastia, ❝ Do you know a legend about Ehrenfest's parrot? Yes. Do you know Wigner’s friend? ❝ I wish I have that parrot to say: "Aber meine Herren das ist keine Wissenschaft!" every time I get the request to use 80% rule ❝ By the way is not it similar to one used in 1997 by Endrenyi L. et al*? Yes and no. There are many papers about partial AUCs but most of them deal with early exposure. That’s not what I’m interested in. BTW, a plot of Mei-Ling’s paper1 shows also that the CV decreases with time. OK, little bit unfair. Not with time itself but with median tmax, where pAUC0–ref,tmax. A small case study2 shows also the high variability of pAUC0–ref,tmax. ❝ As far as I know the war almost starts when russian experts try to translate this word properly. For me "должно быть" sounds like an order. Shall I say I expected that? ❝ ("Marain was a synthetic language, designed to be phonetically and philosophically as expressive as the pan-human speech apparatus and the pan-human brain would allow") Great! Do you know Volapük? ❝ I often deal with protocols where the definition of AUC is ambiguously: in the one part of the protocol they are talking about AUC0–72 and in the other - about AUC0–t (AUClast). Shit. That’s why I give in my protocols even a plot to show how I deal with cut-off times… ❝ About AUC0–t (AUCall) see also Oishi M. et al. Funny abstract. Can you send me a copy?
— Dif-tor heh smusma 🖖🏼 Довге життя Україна! Helmut Schütz The quality of responses received is directly proportional to the quality of the question asked. 🚮 Science Quotes |
Helmut ★★★ Vienna, Austria, 2020-06-08 14:43 (1410 d 12:37 ago) @ Astea Posting: # 21510 Views: 12,446 |
|
Hi Nastia, THX for the paper! 1 Regrettably the authors got many things wrong. The FDA does not require AUC0–t or AUC0–∞ but both. Reading educates. The infamous AUCall is nowhere recommended in the Japanese Guideline 2 (AUCt as everywhere else). Given, the “80% rule” is mentioned as well. The Q&A 3 is more specific and states in A-28: AUC is calculated by a trapezoidal rule as the area under the straight lines connecting the measured points. (my emphasis)OK, the bloody linear trapezoidal but can we measure zero? Nope. Hence, this one as the base of all simulations is already wrong and IMHO, the article missed the point. BTW, this figure is extremely funny cause AUCall is >100% of AUCinf. Rule fulfilled → tick the box → mission accomplished. LLOQ 17% of Cmax, fantastic bioanalytics. This example demonstrates that setting the first value after tlast to zero is nonsense. I have seen splendid rules like “set first one after tlast to LLOQ/2 and keep subsequent ones as BQL”. Here also >100% of AUCinf. Forget it.
— Dif-tor heh smusma 🖖🏼 Довге життя Україна! Helmut Schütz The quality of responses received is directly proportional to the quality of the question asked. 🚮 Science Quotes |
ElMaestro ★★★ Denmark, 2020-06-08 17:33 (1410 d 09:48 ago) @ Helmut Posting: # 21511 Views: 11,729 |
|
Hi all, possibly a little off topic but even with AUCall in all its ugliness and over-estimation, it will not necessarily imply a bias on the BE conclusion unless someone can prove that E(ln(AUCall,T - ln(AUCall,R)) does not equal E(ln(AUCT) - ln(AUCR)), regardless of whether the latter is expressed as AUCinf or AUCt. This isn't about imagining scenarios where the equality does not hold (that would be easy enough) but a consideration of the general (expected) case. For the general case I think we're ok. — Pass or fail! ElMaestro |
Helmut ★★★ Vienna, Austria, 2020-06-08 17:53 (1410 d 09:27 ago) @ ElMaestro Posting: # 21512 Views: 11,884 |
|
Hi ElMaestro, ❝ […] it will not necessarily imply a bias on the BE conclusion unless someone can prove that E(ln(AUCall,T - ln(AUCall,R)) does not equal E(ln(AUCT) - ln(AUCR)), regardless of whether the latter is expressed as AUCinf or AUCt. There will always be a bias in any AUC-approach if BQLs come into play and the true T/R-ratio is not Since the bias will point away from unity, the T/R-ratio will get worse and regulators don’t care (wrong but conservative). Example: I used the C1-column of this post as the reference (complete data). Then I generate three others (T) as R/2. Lin-up/log-down trapezoidal AUCs. The T/R-ratio should be 0.5.
— Dif-tor heh smusma 🖖🏼 Довге життя Україна! Helmut Schütz The quality of responses received is directly proportional to the quality of the question asked. 🚮 Science Quotes |
ElMaestro ★★★ Denmark, 2020-06-09 10:45 (1409 d 16:35 ago) @ Helmut Posting: # 21514 Views: 11,619 |
|
Hi again Helmut, there is no doubt that AUC calculation can be biased. Whether the ln difference or ratio is biased, generally, is another matter and this is where I meant to draw the attention - this is the BE outcome. If one could make a case for saying that this bias affects T more often or to a larger extent than it affects R, generally, then that is where the use of this approach would be not just ugly but even worse also clearly biased. I am not aware this would or should be the case. — Pass or fail! ElMaestro |
Helmut ★★★ Vienna, Austria, 2020-06-09 16:07 (1409 d 11:13 ago) @ ElMaestro Posting: # 21519 Views: 12,090 |
|
Massa, I dunno understand wat ya sayin’. ❝ there is no doubt that AUC calculation can be biased. OK. ❝ Whether the ln difference or ratio is biased, generally, is another matter and this is where I meant to draw the attention - this is the BE outcome. If one could make a case for saying that this bias affects T more often or to a larger extent than it affects R, generally, then that is where the use of this approach would be not just ugly but even worse also clearly biased. If we can measure all concentrations, any of the methods is unbiased as is the T/R estimate (yep, even by the linear trapezoidal and the bloody AUCall). I’m only talking about cases where we deal with BQLs. If T = R (and for nitpickers: we have balanced sequences / equal group sizes) both treatments will be affected to the same extent, the bias of the AUC will mean out and the estimate will tell used T ≈ R. If T < R, T will be more often affected than R. If T > R, R will be more often affected than T. In the first case the T/R estimate will have a negative bias (since we have more cases where T is underestimated) and in the second a positive one (since we have more cases where R is underestimated). Example: One-compartment, D 100, V 1, t½a 1 h, t½e 4 h, FR 1, FT 0.8–1, t 0, 1, 2, 2.5, 3, 3.5, 4, 4.5, 6, 9, 12, 16, 24; λz estimated from the last two measurable concentrations. LLOQ 3% of theoretical Cref,max (62.996). Lin-up / log-down trapezoidal method for AUC0–t, AUC0–∞, and AUC0–t(common). As long as we can measure all concentrations of both T and R, the PE of AUC0–t is unbiased. Once we drop below the LLOQ, the PE will be negatively biased because we are comparing AUC0–16 of T with AUC0–24 of R. Here it happens below a true f of 0.9071 where Clast of T with 1.8898 will be below the LLOQ of 1.8899. The PEs of AUC0–∞ and AUC0–t(common) are always unbiased.*
— Dif-tor heh smusma 🖖🏼 Довге життя Україна! Helmut Schütz The quality of responses received is directly proportional to the quality of the question asked. 🚮 Science Quotes |
Astea ★★ Russia, 2020-06-12 16:15 (1406 d 11:06 ago) @ Helmut Posting: # 21532 Views: 11,125 |
|
Dear Helmut! ❝ Regrettably the authors got many things wrong. Oh, I suspected something like that. Thank you for your valuable comments! I shall start a new entry to discuss some features of PK parameters that were raised here. — "Being in minority, even a minority of one, did not make you mad" |
Helmut ★★★ Vienna, Austria, 2020-06-08 12:40 (1410 d 14:40 ago) @ mittyri Posting: # 21509 Views: 12,010 |
|
Hi mittyri, ❝ I had in mind more simplistic approach: shiny app to give points, parameters of 2cpt model and to see what parameters will give ratio less than 80%. Yep but that’s for a particular study where you have some ideas about the PK model. I had a more general approach in mind. ❝ ❝ … должно быть ≈ should be or must be? ❝ ❝ In guidelines commonly “should be” is used. […] Does it smell of a recommendation or rather a requirement? ❝ ❝ That's a matter of expert's taste and mood ❝ see later: ❝ ❝ Are you good enough to extinguish doubts from expert's mind? You never know! следует усомниться ≈ should doubt. As Nastia wrote: ❝ We use "следует" to slightly soften the meaning… ❝ ❝ The value of the last triangle depends on the time interval between the time point of the last measured concentration and the next one with C=0. You would only get a meaningful result if (1) you use the lin-up/log-down trapezoidal and (2) the interval equals t½. Since (2) is never fulfilled for all profiles, forget it. ❝ ❝ Regarding (1): lin-up/log-down is overriden by linear rule if the end point is non-positive. I should have RTFM. While you replied I deleted this part from my OP, since esp. (2) is nonsense. — Dif-tor heh smusma 🖖🏼 Довге життя Україна! Helmut Schütz The quality of responses received is directly proportional to the quality of the question asked. 🚮 Science Quotes |