 Causes of inconsistent recovery [Bioanalytics]

Hi everyone!

I want to know about the causes of inconsistent recovery specially in liquid extraction because I've been trying to eliminate it from my method but no success yet.

Thanks in advance.
Helmut
★★★
avatar
Homepage
Vienna, Austria,
2012-10-08 21:04
(4615 d 03:36 ago)

@ Ionsource
Posting: # 9343
Views: 3,221
 

 A manifold of potential causes & a story

Hi Ionsource!

❝ I want to know about the causes of inconsistent recovery specially in liquid extraction […]



Without knowing details of the method we have no clue. Sometimes the most weird things happen.
I give you an example (my personal translation):*

Practical example 14: Special case phenol in LLE
We had to analyze samples of several BE studies of silibinin within a certain time interval. The first study was a resounding success. In revalidation for the next study there were always outliers toward lower results in calibration and QC samples. Once a method works, one relieved can pass it to employees and finally want to enjoy the fruits of – in this case – two years of development. And now those results! Employees were closely monitored then in their operation and pH values of buffers critically examined (essential in this method) – and no discrepancies were found. Then you analyze a batch with all your own personal experience and find outliers as well. What’s to blame? To minimize suspense: A small detail had been changed by an employee without being noticed as important. For practical reasons we removed the stoppers from centrifuge vials before freezing out samples during extraction. In contrast to the first study, where all results had been fine, the employee did not use a small beaker (150 mL) with acetone / dry ice (space for about six samples and reaching up to half the height of centrifuge tubes), but a larger beaker (250 mL) – which could hold twice as many samples. The beaker was higher and had about 80% of the height of the centrifuge tubes. So where was the problem? While the samples cooled down, some of the dry ice evaporates from the acetone / dry ice mixture as CO2 gas, which is relatively dense and slowly trickles down over the beaker rim. With an occasional little air movement CO2 gas in the 250 mL beaker could flow into the centrifuge tube. CO2 was decanted along with the solvent and thereby changing the pH of the buffer (pH 10.5) slightly downwards. Since the pKa of phenols is about 9.5-9.8, one immediately sees that the back-extraction performed at pH values of 10 or just below was clearly impaired and gave worse recoveries.


Shit happens. Sometimes it is very difficult to find out why. My friend Hermann Mascher has close to 40 years of experience in bioanalytics. I can image his feelings.


  • H Mascher
    Klinische Analytik mit HPLC. Ein Ratgeber für die Praxis
    Wiley-VCH, Weinheim, pp 111–2 (2010)
    I can recommend the new English edition:
    HPLC Methods for Clinical Pharmaceutical Analysis
    Wiley-VCH, Weinheim (2012)
    ISBN: 978-3-527-33129-1

Dif-tor heh smusma 🖖🏼 Довге життя Україна! [image]
Helmut Schütz
[image]

The quality of responses received is directly proportional to the quality of the question asked. 🚮
Science Quotes
Ionsource
☆    

Pakistan,
2012-10-09 10:50
(4614 d 13:50 ago)

@ Helmut
Posting: # 9350
Views: 3,088
 

 A manifold of potential causes & a story

Hello HS

Really strange and interesting story.:cool:

❝ Without knowing details of the method we have no clue.


Details! Why not!
I'm trying to extract Amlodipine from plasma with following scheme.

1) 0.4ml plasma+ 50 ul IS+ 40ul 1N NaOH+ 1ml Tertiary Butyl Methyl Ether.
2) Vortex for 1min.
3) Centrifuge at 9000g for 10min.
4) Take 850ul supernatant. Dry under Nitrogen stream and reconstitute in 80ul mobile phase.

The analytical range is 0.5-12ng/ml on Fluorescence.

And thanks for the book reference.

Regards
UA Flag
Activity
 Admin contact
23,424 posts in 4,927 threads, 1,675 registered users;
76 visitors (0 registered, 76 guests [including 13 identified bots]).
Forum time: 00:40 CEST (Europe/Vienna)

Every man gets a narrower and narrower field of knowledge
in which he must be an expert in order to compete with other people.
The specialist knows more and more about less and less
and finally knows everything about nothing.    Konrad Lorenz