 Change in ion source [Bioanalytics]

Dear All,

Kindly let me have some idea about how the change in ion source affect the matrix effect?

Kindly give your valuable comment on this.

Regards,

Compliance
Ohlbe
★★★

France,
2012-07-20 12:07
(4696 d 19:36 ago)

@ Compliance
Posting: # 8962
Views: 3,441
 

 Change in ion source

Dear Compliance,

If you change between two different principles of ion source (e.g. ESI and APCI): matrix effects will be totally different. The principle of ionisation is totally different between ESI (ionisation in the liquid phase, then evaporation of ions into the gas phase) and APCI (evaporation, then ionisation in the gas phase). APCI is far less sensitive to matrix effects. In APCI you sometimes have ion enhancement, not just ion suppression (enhancement is not frequent in ESI).

If you change between two different types, with the same principle (between 2 ESI sources, or 2 APCI sources, for instance between an API 3000 and an API 4000, or between Sciex and Micromass): you can still get differences in matrix effects, due to differences in the geometry of the source, in the capillary, shape of the spray, design of the cone... For one molecule you can have a strong ion suppression with one source, and less problems with the other; for the next molecule it may be the other way ! It is very difficult to predict.

So to summarise:

❝ Kindly let me have some idea about how the change in ion source affect the matrix effect?


You have to try and see !

Regards
Ohlbe

Regards
Ohlbe
Compliance
★    

India,
2012-07-20 13:36
(4696 d 18:07 ago)

@ Ohlbe
Posting: # 8963
Views: 3,351
 

 Change in ion source

Dear Ohlbe,

Thank you for the response.

In this case it is always better to evaluate Matrix effect during instrument change from API to waters or vice versa right? Are you agree with me?

Regards,

Compliance
Ohlbe
★★★

France,
2012-07-20 16:24
(4696 d 15:19 ago)

@ Compliance
Posting: # 8964
Views: 3,346
 

 Change in ion source

Dear Compliance,

❝ In this case it is always better to evaluate Matrix effect during instrument change from API to waters or vice versa right?


Not just better: essential !

Regards
Ohlbe

Regards
Ohlbe
UA Flag
Activity
 Admin contact
23,424 posts in 4,927 threads, 1,678 registered users;
110 visitors (0 registered, 110 guests [including 9 identified bots]).
Forum time: 07:44 CEST (Europe/Vienna)

If you want to get people to believe
something really, really stupid,
just stick a number on it.    Charles Seife

The Bioequivalence and Bioavailability Forum is hosted by
BEBAC Ing. Helmut Schütz
HTML5