 Retention Times in LCMS [Bioanalytics]

Can any one clarify me regarding the importance of the Retention Time in the LCMS analysis.

Whether the interferences have to be considered based on the RT or based on the interested peak shape?

Regards,

nvs
vpardhasaradhi
☆    

2007-09-03 09:46
(6873 d 22:36 ago)

@ nvs
Posting: # 1052
Views: 3,896
 

 Retention Times in LCMS

Dear Mr. NVS,

Retention times in LC-MS/MS analysis are important with respect to separation of analyte(s) and the internal standard from the other impurities eluting out from the analytical column. Unfortunately, the elution of these impurities can not be seen or noticed and these may cause ion suppression or enhancement. While working at very lower concentrations of the analyte where the internal standard concentration is comparatively high, even the internal standard can influence (either suppress or enhance) the ionization of the analyte.

Bad peak shapes need not be because of interferences. Bad peak shapes can also be because of unsuitable chromatographgic conditions (analytical column or mobile phase).

Hence, it is important to check the matrix effect (ion suppression or enhancement) and improve the chromatographic conditions for better peak shapes and less matrix effects.

V.Pardhasaradhi

--
Edit: Full quote removed. Please see this post! [HS]
UA Flag
Activity
 Admin contact
23,655 posts in 4,993 threads, 1,570 registered users;
358 visitors (0 registered, 358 guests [including 14 identified bots]).
Forum time: 08:22 CEST (Europe/Vienna)

Most scientists today are devoid of ideas, full of fear, intent on
producing some paltry result so that they can add to the flood
of inane papers that now constitutes “scientific progress”
in many areas.    Paul Feyerabend

The Bioequivalence and Bioavailability Forum is hosted by
BEBAC Ing. Helmut Schütz
HTML5