 increment in sample volume [Bioanalytics]

Dear All,

How the increment in the sample volume can increasing the method sensitivity. kindly explain me as i am unable to understand that during validation X concentration gives good results with Y sample volume and during study due to this Y sample volume sensitivity is affected.

kindly clear me on that.

Regards,

Auditor
Helmut
★★★
avatar
Homepage
Vienna, Austria,
2012-01-09 18:08
(4892 d 03:30 ago)

@ auditor
Posting: # 7884
Views: 1,967
 

 increment in sample volume

Dear Auditor!

❝ How the increment in the sample volume can increasing the method sensitivity.


Increasing the sample volume will increase sensitivity only if you have a quite good sample preparation / trace enrichment. For simple point-and-shoot methods (e.g. protein precipitation) most likely background and/or matrix effects will increase as well outweighing the enhanced signal of the analyte(s). Since separation of analyte(s) from matrix’ components is never perfect, even in a ‘good’ method doubling the sample volume likely will not half the LLOQ.

❝ […] i am unable to understand that during validation X concentration gives good results with Y sample volume and during study due to this Y sample volume sensitivity is affected.


Sorry, I don’t understand your point. Can you please reword it?

Dif-tor heh smusma 🖖🏼 Довге життя Україна! [image]
Helmut Schütz
[image]

The quality of responses received is directly proportional to the quality of the question asked. 🚮
Science Quotes
UA Flag
Activity
 Admin contact
23,424 posts in 4,927 threads, 1,677 registered users;
34 visitors (0 registered, 34 guests [including 25 identified bots]).
Forum time: 22:38 CEST (Europe/Vienna)

Everything is trivial, if you know the answer.    Thomas Jaki

The Bioequivalence and Bioavailability Forum is hosted by
BEBAC Ing. Helmut Schütz
HTML5