haydonat
☆    

Turkey,
2011-01-13 11:27
(5645 d 14:33 ago)

Posting: # 6410
Views: 6,816
 

 variation in response of the analyte and internal standard [Bioanalytics]

Dear all;

I have response variation problem. When injections made from same vial responses of the analyte and IS changes but not with same proportionation.

extraction procedure: liquid-liquid sample preparation
column: x-terra c18
mobile phase: Acetonitrile: 10mM ammonium acetate:formic acid (50:50:0.530, v:v:v) pH: 4.6
reconstitue solution: water+25mM ammonia+ACN (50:50), pH: 9.9
injection volume: 15 µl
flow rate: 0.25 ml/min

Also after every six injections in a batch mobile phase is injected four times
and results get better with this procedure. But, I can not provide reproducibility between batches.Results of one batch could be good with MP cleaning,but the other ones results could be bad. :crying:

I have written my conditions above, without any changes in my conditions, what can you suggest me solve response difference problem? :ponder:

Hale Aydonat
Helmut
★★★
avatar
Homepage
Vienna, Austria,
2011-01-13 18:08
(5645 d 07:52 ago)

@ haydonat
Posting: # 6411
Views: 5,859
 

 variation in response of the analyte and internal standard

Dear Hale!

❝ mobile phase: Acetonitrile: 10mM ammonium acetate:formic acid

❝ (50:50:0.530, v:v:v) pH: 4.6

❝ reconstitue solution: water+25mM ammonia+ACN (50:50), pH: 9.9


One thing comes into my mind: A different pH of the analyte's solution and the mobile phase may lead to instable conditions on top of the column. Do you have problems getting the extracts into solution? Maybe ultrasonification helps. IMHO it would be better to dissolve extracts in mobile phase - or even better a solvent with slighlty less organic modifier (results in narrower peaks).

Dif-tor heh smusma 🖖🏼 Довге життя Україна! [image]
Helmut Schütz
[image]

The quality of responses received is directly proportional to the quality of the question asked. 🚮
Science Quotes
haydonat
☆    

Turkey,
2011-01-15 16:16
(5643 d 09:44 ago)

@ Helmut
Posting: # 6420
Views: 5,705
 

 variation in response of the analyte and internal standard

Dear Helmut,

Thanks for replying me :-)

I don't have any problem about getting the extracts into solution, I use diethylether:n-hexane (7:3) mixture during extraction, after evaporation step of my procedure; reconstitution was done mobile phase before but result were not satisfying so water+25mM ammonia+ACN (50:50), pH: 9.9 mixture is begin to use.

Also; validation was completed and this problem is occured in study so, I cannot change any of my conditions/solutions/pH.

By the way; I think this problem leads to ionization problem. What can you think about changing desolvation temperature and desolvation gas flow? And if I change desolvation temperature and/or desolvation gas flow, do I have to perform whole validation once more?


Edit: Full quote removed. Please delete anything from the text of the original poster which is not necessary in understanding your answer; see also this post! [Helmut]
Ohlbe
★★★

France,
2011-01-16 00:09
(5643 d 01:51 ago)

@ haydonat
Posting: # 6422
Views: 5,671
 

 variation in response of the analyte and internal standard

Dear Hale,

❝ Also; validation was completed and this problem is occured in study so, I

❝ cannot change any of my conditions/solutions/pH.


Well... If you have problems with the method during the study, you will have to change something in the method and this may require a re-validation !

❝ By the way; I think this problem leads to ionization problem. What can you

❝ think about changing desolvation temperature and desolvation gas flow?


Yes, you can try and see if it helps. Possibly also other ionisation parameters such as cone voltage / declustering potential or whatever it is called on your instrument.

❝ And if I change desolvation temperature and/or desolvation gas flow, do I

❝ have to perform whole validation once more?


Well, if it affects the ionisation significantly enough to solve your problem... Almost full validation, yes. You don't have to study stabilities again, but obviously you will have to go for matrix effects, and I would say you will also need to re-demonstrate LLOQ, precision and accuracy...

Regards
Ohlbe

Regards
Ohlbe
Ohlbe
★★★

France,
2011-01-13 20:14
(5645 d 05:46 ago)

@ haydonat
Posting: # 6413
Views: 5,779
 

 variation in response of the analyte and internal standard

Dear Hale,

❝ I have response variation problem. When injections made from same vial

❝ responses of the analyte and IS changes but not with same proportion.


❝ Also after every six injections in a batch mobile phase is injected four

❝ times and results get better with this procedure.


I can see two different reasons for this problem, in addition to the one mentioned by Helmut, but there may be some others:

- one common problem with LC/MS/MS can be the need to condition the ion source to obtain a stable signal. You may need a number of injections from plasma samples before your signal stabilises. Sometimes it is just half a dozen, but it may happen that you need 20 or 30 injections. Injecting mobile phase regularly doesn't really solve the issue, you rather need to inject some plasma blanks before your samples. If you are in this situation you probably see a trend towards a decrease or increase in the signal of the internal standard over time in the run. It may not be that clear if the IS is not affected much but the analyte is. You may be able to explore this by repeated injections of a QC sample in a run and monitoring the response of the analyte and IS.

- another possibility is the accumulation of phospholipids on the column, which can then "bleed" out of the column in an erratic way, leading to matrix effects issues. Variations in retention times can sometimes also be seen due to this accumulation, which modifies the polarity of the stationary phase. Possible solutions are a better sample clean-up, or use of a mobile phase gradient to clean the column between two injections (will not take longer than your injections of mobile phase).

Regards
Ohlbe

Regards
Ohlbe
haydonat
☆    

Turkey,
2011-01-15 16:22
(5643 d 09:38 ago)

@ Ohlbe
Posting: # 6421
Views: 5,666
 

 variation in response of the analyte and internal standard

Dear Ohlbe,

Thanks for replying me :-)

I am going to try your first suggestion about blank injection. Also I do not think that if accumulation happens in column because, retention times are constant for both analyte and internal standard. But I will keep this in my mind :-)


Edit: Full quote removed. Please delete anything from the text of the original poster which is not necessary in understanding your answer; see also this post! [Helmut]
NewInPK
☆    

2011-02-01 19:32
(5626 d 06:28 ago)

@ haydonat
Posting: # 6532
Views: 5,516
 

 variation in response of the analyte and internal standard

Dear Hale,

In case the problem has not been solved yet, I would suggest the following:

- you would benefit from injecting a pool of extracted samples, say low level QC, rather than blank to "warm-up" the instrument. In this way you can see how the ratio varies and how many samples it takes to obtain a consistent signal.
- If you decide to use a cleaning injection, for consistency that injection(s) should be made after each sample.
- for cleaning phospholipids, I would recommend adding a 1-2 minutes rinsing step with 100% methanol. You would need to determine the time based on the flow rate and all the conditions of your method, by monitoring the phospholipids. They have a common fragment of 184 and a precursor ion scan test will reveal their m/z in Q1. A few m/z that I remember are 496, 524 and 782. They elute fastest with about 80-85% acetonitrile, at higher organic % they start to retain more. Among other reasons, this is why the classic gradient that goes to 100% ACN is not recommended. However, 100% methanol will elute them.
- you have mentioned not having this issue during the validation, but with the samples. There could be a number of reasons:
- a metabolite with the same MRM as the IS (probably not the case, but it could happen)
- a metabolite co-eluting with IS and impacting the ionization
- the formulation components (this can be verified by looking if there is a trend of the IS response over the sampling time, corresponding to the absorption and elimination of the "fillers"). If you are allowed, you can also spike a low conc in a blank sample and a later time point of a placebo - if available.
- inter-subject variability.

All this stands true assuming that the variability of IS response is confirmed to occur in the mass spec, and not during extraction (LLE extractions are notorious for inconsistencies).

Last point: if a validation becomes necessary, and if it is possible, the reconstitution solution should be very close to the composition (including pH) of the mobile phase. Some compounds behave well, but other compounds are impacted by the difference in pH, especially if they have multiple pKa values, and don't have time to equilibrate from one form to another during the elution.
I hope this helps, if not for the current problem, maybe for potential future issues.
Ohlbe
★★★

France,
2011-02-02 11:22
(5625 d 14:38 ago)

@ NewInPK
Posting: # 6539
Views: 5,558
 

 variation in response of the analyte and internal standard

Dear Hale and NewInPK,

❝ [...] monitoring the phospholipids. They have a common fragment of 184 and a

❝ precursor ion scan test will reveal their m/z in Q1. A few m/z that I

❝ remember are 496, 524 and 782.


One possibility described in the literature is to use high energy in-source collision induced dissociation to generate in the source this common fragment at m/z 184, and follow it in Q1, have a low collision energy in Q2 to avoid further fragmentation, and follow m/z 184 in Q3 again. This avoids having to follow multiple transitions in addition to your analyte and IS. See for instance:

Liquid chromatography–mass spectrometry/mass spectrometry method development for drug metabolism studies: Examining lipid matrix ionization effects in plasma
James L. Little, Michael F. Wempe, Charles M. Buchanan
Journal of Chromatography B, 833 (2006) 219–230

Monitoring phospholipids for assessment of matrix effects in a liquid chromatography–tandem mass spectrometry method for hydrocodone and pseudoephedrine in human plasma
Omnia A. Ismaiel, Matthew S. Halquist, Magda Y. Elmamly,
Abdalla Shalaby, H. Thomas Karnes

Journal of Chromatography B, 859 (2007) 84–93

❝ - the formulation components


As far as I know this mostly applies to formulations for injection, where substances like PEG 400 or Tween 80 (polysorbate) have been described to create strong ion suppression. However this mostly affects the early samples in the PK profile (basically, first 1 or 2 hours).

regards
Ohlbe

Regards
Ohlbe
UA Flag
Activity
 Admin contact
23,655 posts in 4,993 threads, 1,570 registered users;
329 visitors (0 registered, 329 guests [including 21 identified bots]).
Forum time: 03:01 CEST (Europe/Vienna)