sambasivarao
☆    

2010-10-06 12:40
(5745 d 06:30 ago)

Posting: # 6006
Views: 4,128
 

 Recovery of drugs from biological fluids [Bioanalytics]

If I am able to Extract 30% of drug from plasma for specific molecule and it is reproducible & Same throught my Calibration curve range.
Now how to justify a auditor as i got less recovery?
Helmut
★★★
avatar
Homepage
Vienna, Austria,
2010-10-06 14:53
(5745 d 04:17 ago)

@ sambasivarao
Posting: # 6008
Views: 3,384
 

 Recovery of drugs from biological fluids

Dear Sambasiva,

first look here.

If the method is validated according to the original definition (“… demonstrated to be reliable and reproducible for the intended use”) any (!) recovery should be acceptable. If your method is sensitive enough even with a low recovery, fine. According to the FDA’s 2001 guidance (and countries following closely, like ANVISA) you have to report your findings – there’s no lower limit for acceptability:

Recovery of the analyte need not be 100%, but the extent of recovery of an analyte and of the internal standard should be consistent, precise, and reproducible.

(my emphasis)
The EMA in their 2009 draft GL went one step further; recovery is not mentioned at all. :ok:

❝ Now how to justify a auditor as i got less recovery?


Patient and politely.


Celebrating login #3000 today.

Dif-tor heh smusma 🖖🏼 Довге життя Україна! [image]
Helmut Schütz
[image]

The quality of responses received is directly proportional to the quality of the question asked. 🚮
Science Quotes
UA Flag
Activity
 Admin contact
23,656 posts in 4,994 threads, 1,570 registered users;
283 visitors (0 registered, 283 guests [including 20 identified bots]).
Forum time: 19:10 CEST (Europe/Vienna)

Most scientists today are devoid of ideas, full of fear, intent on
producing some paltry result so that they can add to the flood
of inane papers that now constitutes “scientific progress”
in many areas.    Paul Feyerabend

The Bioequivalence and Bioavailability Forum is hosted by
BEBAC Ing. Helmut Schütz
HTML5