Dear All,

We have validated method for FDC (2 analytes) by lcmsms. But when we initiated subject sample analysis one of the analyte is showing variation in response due to which 4 subjects out of 12 analysed until now need to be repeated due to results for this analyte is not within batch acceptance criteria. There are total 64 subjects to be analysed. We have checked all possibilities for this response variation but there is no change.
I think there is some change in the lcmsms parameter for this molecule; therefore I want to know can I tune or optimize the lcmsms parameters for this molecule in-between the analysis and continue further analysis of remaining subject after carrying out a PA batch in partial method validation or how to go about it.

Thanks & regards
Dr. Ketan Nerurkar
Ohlbe
★★★

France,
2010-05-17 12:25
(5884 d 18:08 ago)

@ ketan
Posting: # 5338
Views: 3,806
 

 Change in analyte response

Dear Ketan,

1. What internal standard are you using: stable-labelled, structural analogue, other ?
2. What type of sample preparation are you using: protein precipitation, liquid-liquid extraction, solid phase extraction ?
3. Any "strange result" in the validation ?
4. What do you mean by "variation in response" of the analyte: do you mean that for your QC samples, for a same concentration you get a very different response for the analyte, while the response for the internal standard is stable ? And how much "variation" do you see ?
5. What LC/MS/MS parameter do you plan to change ?

Regards
Ohlbe

Regards
Ohlbe
ketan
☆    

India,
2010-05-18 09:51
(5883 d 20:42 ago)

@ Ohlbe
Posting: # 5341
Views: 3,822
 

 Change in analyte response

Dear Ohlbe,

Thank you for going through my query, please find below answers to your questions you have raised for my query

❝ 1. What internal standard are you using: stable-labelled, structural analogue, other ?


No, internal standard is not stable-labelled, structural analogue, but the internal standard response throughout analysis is reproducilbe with excellent % CV.

❝ 2. What type of sample preparation are you using: protein precipitation, liquid-liquid extraction, solid phase extraction ?


sample preparation is by liquid-liquid extraction method.

❝ 3. Any "strange result" in the validation ?


No, method validation was extremely smooth and satisfactory with hardly 3 to 4 QCs out of acceptance criteria throughout validation.

❝ 4. What do you mean by "variation in response" of the analyte: do you mean that for your QC samples, for a same concentration you get a very different response for the analyte, while the response for the internal standard is stable ? And how much "variation" do you see ?


variation in response means in system suitability test (prepared in matrix) first five injection shows reproducible response but the sixth injection shows different response for only that analyte while another analyte and IS shows good reproducibility. Similar results are observed during subject analysis after 4 to 5 injections there is change in response for that particular analyte and due to which we have to reject the batch (e.g. in cc and QC sample shows accuracy of 70 or 140 % for any of the in-between standard/QC sample).

❝ What LC/MS/MS parameter do you plan to change ?


I planning to change slightly Source parameters of LC/MS/MS and not the compound parameters because I think gases need to be optimized.

Once again thank you,
Regards
Ketan


Edit: Standard quotes restored. [Helmut]
Ohlbe
★★★

France,
2010-05-18 12:48
(5883 d 17:45 ago)

@ ketan
Posting: # 5345
Views: 4,444
 

 Change in analyte response

Dear Ketan,

❝ No, internal standard is not stable-labelled, structural analogue, but the internal standard response throughout analysis is reproducilbe with excellent % CV.


Which is actually a problem: your IS noes not mimic the analyte and therefore does not compensate for the variability you observe.

❝ variation in response means in system suitability test (prepared in matrix) first five injection shows reproducible response but the sixth injection shows different response for only that analyte while another analyte and IS shows good reproducibility. Similar results are observed during subject analysis after 4 to 5 injections there is change in response for that particular analyte and due to which we have to reject the batch (e.g. in cc and QC sample shows accuracy of 70 or 140 % for any of the in-between standard/QC sample).


This looks like your ion source needs some injections to stabilise. May be due to matrix effects (endogenous components accumulating on the column and getting released gradually thereafter) or due to a sort of conditioning. Does the response still vary afterwards, or does it get stable (but at a different level) after these first 5-10 (or more) injections ? If it does get stable afterwards, you may need to start your run with some injections of blank samples, and pay attention if there is an interruption in your run and you resume injections after some time.

❝ I planning to change slightly Source parameters of LC/MS/MS and not the compound parameters because I think gases need to be optimized.


This could indeed improve the situation, and just running a P&A run could be acceptable. However if you saw no problem during your initial validation, there are chances the P&A run will pass too. How do you plan to make sure that your new parameters will improve the situation, before you re-start sample analysis ?

Regards
Ohlbe

Regards
Ohlbe
ketan
☆    

India,
2010-05-18 14:06
(5883 d 16:26 ago)

@ Ohlbe
Posting: # 5347
Views: 4,264
 

 Change in analyte response

Dear Olhbe,

Thanks for your views and valuable suggestions on my query, please find my observation and view for your perusal.

❝ This looks like your ion source needs some injections to stabilise. May be due to matrix effects (endogenous components accumulating on the column and❝ getting released gradually thereafter) or due to a sort of conditioning.


We have carried out matrix effect during validation and also for bulk cc and QC sample preparation for subject analysis which was satisfactory for all analytes and IS.

❝ Does the response still vary afterwards, or does it get stable (but at a different level) after these first 5-10 (or more) injections ?


We had to inject at least 25 SST injections to stabilize the system before injecting six injection for the analysis batch which was not the case during method validation.

❝ If it does get stable afterwards, you may need to start your run with some injections of blank samples, and pay attention if there is an interruption in your run and you resume injections after some time.


If there is interruption in run it should have effected the response for another analyte also which is not seen here.

❝ This could indeed improve the situation, and just running a P&A run could be acceptable. However if you saw no problem during your initial validation, there are chances the P&A run will pass too. How do you plan to make sure that your new parameters will improve the situation, before you re-start sample analysis ?


I am planning to optimize instrument parameters (Gases) and observe the response for both analyte and IS by injecting SST solution, if the system gates stable in minimum number of injections, I will analyze one PA batch in PMV and then proceed for subject analysis.

I want to know can I continue with subject analysis from no 13 as I have already analysed 12 out of which 4 needs to be repeated or I will have to start from first subject as instrument parameters will change from 13th subject.


Regards
Ketan


Edit: standard quotes restored [Ohlbe]
Ohlbe
★★★

France,
2010-05-18 16:29
(5883 d 14:04 ago)

@ ketan
Posting: # 5348
Views: 3,753
 

 Change in analyte response

Dear Ketan

❝ We have carried out matrix effect during validation and also for bulk cc and QC sample preparation for subject analysis which was satisfactory for all analytes and IS.


True. But matrix effects are tested on a limited number of injections, and accumulating effects may not be visible during the validation.

❝ If there is interruption in run it should have effected the response for another analyte also which is not seen here.


No. Whatever the cause for your problem, ionisation is an analyte-specific issue. You may have problems with one analyte and not with the other. This is actually your case: you have problems with one analyte, but not with the internal standard.

❝ I want to know can I continue with subject analysis from no 13 as I have already analysed 12 out of which 4 needs to be repeated or I will have to start from first subject as instrument parameters will change from 13th subject.


As you have a clearly identified analytical issue I would re-analyse all samples after optimisation, if the situation improves. Will you be doing incurred sample reanalysis ? If some subject samples are affected also in the runs you have validated so far, you may end up failing ISR. Better be on the safe side, if you still have enough samples.

Regards
Ohlbe

Regards
Ohlbe
jchapdelaine
☆    

2010-05-18 17:01
(5883 d 13:32 ago)

@ Ohlbe
Posting: # 5349
Views: 3,763
 

 Change in analyte response

Dear Ketan

I recommend to do an infusion test to determine the extent of the problem. Apparently, you have an ionization effect, therefore by infusing your drugs of interest along with the internal standards and injecting a QC sample versus a clinical sample, you will be able to see any ionization differences with time (maybe 30 minutes is sufficient). This will allow you to demonstrate the method is functioning properly (or not) without consuming a high volume of the clinical samples.

From past experience, I expect that modifying the gases will not solve your problem.
UA Flag
Activity
 Admin contact
23,655 posts in 4,993 threads, 1,570 registered users;
364 visitors (0 registered, 364 guests [including 21 identified bots]).
Forum time: 06:33 CEST (Europe/Vienna)

Don’t compromise yourself.
You are all you’ve got.    Janis Joplin

The Bioequivalence and Bioavailability Forum is hosted by
BEBAC Ing. Helmut Schütz
HTML5