 system suitability / Preassay calibration check [Bioanalytics]

Dear All,

System suitability: Usually Extracted (or neat) Mid-QC level concentrations injections are given prior to submission of a batch. How many such injections needs to be given to evaluate - system performance (or) % CV of response ratio. Please cite for reference if any.

Best Regards
Indrakumar
moblak
☆    

2010-04-22 17:30
(5911 d 17:02 ago)

@ IK
Posting: # 5187
Views: 3,226
 

 system suitability / Preassay calibration check

Hi Indrakumar,

in the reference cited below you can find a very nice example of how an SST can be performed:
"System suitability in bioanalytical LC/MS/MS" Briscoe CJ, Stiles MR, Hage DS. J Pharm Biomed Anal. 2007 Jun 28;44(2):484-491.


Regards
Marko
IK
☆    

India,
2010-04-26 07:37
(5908 d 02:55 ago)

@ moblak
Posting: # 5222
Views: 2,999
 

 system suitability / Preassay calibration check

Hi Marko,

Thanks a lot for the reference cited in your response. It was really helpful.