 change in Calibration Curve range [Bioanalytics]

Dear group members,

we are doing subject sample analysis based on validation range. And we have completed one study dosage (500mg). But dosage for next study became half (250mg). We reduced Calibration Curve STD1 to half of original. But changed Calibration Curve range within the Validation range.
Shall I continue my study analysis or is there any need to perform partial validation?

Thanks in advance

Regards,
gunta
Helmut
★★★
avatar
Homepage
Vienna, Austria,
2009-08-25 22:31
(6148 d 20:46 ago)

@ gunta
Posting: # 4113
Views: 3,298
 

 Change of CC range within validation range

Dear Gunta!

If your new lowest CC-standard is within the validated range of the method IMHO there is no need for a revalidation. I assume that you have also data at the new LLOQ?

Dif-tor heh smusma 🖖🏼 Довге життя Україна! [image]
Helmut Schütz
[image]

The quality of responses received is directly proportional to the quality of the question asked. 🚮
Science Quotes
gunta
☆    

2009-08-27 07:44
(6147 d 11:33 ago)

@ Helmut
Posting: # 4123
Views: 3,284
 

 Change of CC range

Dear HS,

Thank you for your reply. LLOQ has not changed for new calibration curve range. But change has been at ULOQ. first range is 20ng to 8000ng (validation range). New range is 20ng to 4000ng without change in LLOQ.
Hoping to get your reply,

Thanks and Regards,
gunta.


Edit: Full quote removed. Please delete anything from the text of the original poster which is not necessary in understanding your answer; see also this post! [Helmut]
Helmut
★★★
avatar
Homepage
Vienna, Austria,
2009-08-27 18:01
(6147 d 01:16 ago)

@ gunta
Posting: # 4126
Views: 3,393
 

 Arlington III WP

Dear Gunta!

❝ […] first range is 20ng to 8000ng (validation range). New range is 20ng to 4000ng without change in LLOQ.


Nothing in FDA’s guidance, but in the Arlington III Whitepaper:
Viswanathan CT, Bansal S, Booth B, DeStefano AJ, Rose MJ, Sailstad J, Shah VP, Skelly JP, Swann PG, and R Weiner
Workshop/Conference Report – Quantitative Bioanalytical Methods Validation and Implementation: Best Practices for Chromatographic and Ligand Binding Assays
The AAPS Journal 9(1), E30-E41 (2007)
online resource

If a narrow range of analysis values is known or anticipated before the start of sample analysis, it is recommended that either the standard curve be narrowed and new QC concentrations used as appropriate, or if the original curve is used, existing QC concentrations be revised or sufficient QC samples at additional concentration(s) added to adequately reflect the concentrations of the study samples. Narrowing of the standard curve and preparation of new QC samples requires only a partial validation to ensure adequate performance of the new curve and QCs. A full validation is not required.


Dif-tor heh smusma 🖖🏼 Довге життя Україна! [image]
Helmut Schütz
[image]

The quality of responses received is directly proportional to the quality of the question asked. 🚮
Science Quotes
gunta
☆    

2009-08-30 20:44
(6143 d 22:33 ago)

@ Helmut
Posting: # 4135
Views: 3,166
 

 Arlington III WP

Dear HS,
Thank you very much for your valuable information,

Best Regards,
gunta


Edit: Full quote removed. Please delete anything from the text of the original poster which is not necessary in understanding your answer; see also this post! [Helmut]
UA Flag
Activity
 Admin contact
23,655 posts in 4,993 threads, 1,570 registered users;
415 visitors (0 registered, 415 guests [including 21 identified bots]).
Forum time: 19:18 CEST (Europe/Vienna)

Ignorance more frequently begets confidence
than does knowledge.    Charles Darwin

The Bioequivalence and Bioavailability Forum is hosted by
BEBAC Ing. Helmut Schütz
HTML5