pmgjoshi ☆ India, 2009-06-28 13:17 (6206 d 18:18 ago) Posting: # 3912 Views: 4,909 |
|
|
Dear sir, I want to know the basis for the following Calibration curve acceptance criteria in the guideline: "Calibration Curve/Standard Curve/Concentration-Response The following conditions should be met in developing a calibration curve: #20% deviation of the LLOQ from nominal concentration #15% deviation of standards other than LLOQ from nominal concentration" Specifically regarding the %RSD, is there any Statistical basis for the limits ? Why 20% at LLOQ level? and why 15% at other than LLOQ level? Best regards, PMGJOSHI |
|
Helmut ★★★ ![]() Vienna, Austria, 2009-06-28 15:30 (6206 d 16:04 ago) @ pmgjoshi Posting: # 3913 Views: 4,406 |
|
|
Dear pmgjoshi! ❝ Dear sir, ^^ Not interested in opinions of madams in the forum? ![]() ❝ Specifically regarding the %RSD, is there any Statistical basis for the limits ? No! Only the trivial observation that both inaccuracy and imprecision increase towards lower concentrations. The limits are arbitrary. ❝ Why 20% at LLOQ level? and why 15% at other than LLOQ level? History and convenience. Shah’s paper (1987) [1] was the basis of the first Arlington Conference in 1990. Two papers dealing with the topic[2,3] were published before the conference report.[4a-d] Arlington I (and Arlington II in 2000[5]) lead to FDA’s guidance (2001).[6] The most recent conference in the series (Arlington III) was held in 2006.[7] For a historical overview see Shah (2007).[8]There was a lively discussion at Arlington I about the arbitrary cut-offs (20%/15%). Actually at the conference wider limits at LLOQ were proposed for bioassays only (I still have the conference notes somewhere); limits were widened for chromatographic methods by the editors of the report – contrary to the consensus reached at the conference. I know some of them personally – they exchanged arguments for quite a long time (before the age of the ’net – faxes going to and fro). If you want to know where the numbers 20%/15% come from, please ask Vinod Shah – and share his answer with us. ![]() Remember: 20%/15% are not carved in stone. In Ligand Binding Assays wider limits are not uncommon. Even for ‘conventional methods’ the guidance allows for wider limits, if justified. But to be honest, you have to have strong arguments from the method development phase to justify such a widening. First I would recommend replicate analyses (duplicates, triplicates) in order to bring variability down. I have seen only one case in almost 30 years where 30%/20% were used in an LC-method – but it was accepted. ![]() References:
Bioanalytical Method Validation—A Revisit with a Decade of Progress Pharm Res 17(12), 1551-7 (2000) online resource Center for Drug Evaluation and Research (CDER), Center for Veterinary Medicine (CVM) Guidance for Industry: Bioanalytical Method Validation Rockville, MD, USA (2001) online resource Workshop/Conference Report—Quantitative Bioanalytical Methods Validation and Implementation: Best Practices for Chromatographic and Ligand Binding Assays The AAPS Journal 9(1), E30-E41 (2007) online resource The History of Bioanalytical Method Validation and Regulation: Evolution of a Guidance Document on Bioanalytical Methods Validation The AAPS Journal 9(1), E43-E47 (2007) online resource — Dif-tor heh smusma 🖖🏼 Довге життя Україна! ![]() Helmut Schütz ![]() The quality of responses received is directly proportional to the quality of the question asked. 🚮 Science Quotes |




![[image]](https://static.bebac.at/pics/Blue_and_yellow_ribbon_UA.png)
![[image]](https://static.bebac.at/img/CC by.png)