 Balance calibration [Bioanalytics]

Dear Members,

Is there any guideline specifying "Which class weights (Class E1, E2 or F1) need to be used for calibration of micro balance used in bioanalytical studies". If not, can any one give some recommendation from experience? What can be the tolerances in the balance qualification test?

Thanking you in advance,

V.Pardhasaradhi
Helmut
★★★
avatar
Homepage
Vienna, Austria,
2007-12-18 18:42
(6764 d 11:00 ago)

@ vpardhasaradhi
Posting: # 1394
Views: 4,336
 

 Balance calibration

Dear V. Pardhasaradhi!

❝ Is there any guideline specifying "Which class weights (Class E1, E2 or

❝ F1) need to be used for calibration of micro balance used in bioanalytical

❝ studies".


At least I don't know any (this one from the FDA refers to cGMP)...

❝ If not, can any one give some recommendation from experience?

❝ What can be the tolerances in the balance qualification test?


We obtained our first GLP-certificate in 1992, and never got any problems ever since by the following procedure:
  1. The balance is daily calibrated by the internal 100 g weight (after >15 min of warm-up).
  2. If the displayed result is 100.0000 g, proceed with Step 3. If not, Step 2 may be repeated once - if the result is 100.0000 g, proceed with Step 3 - or Step 6 otherwise.
  3. External weights are also used daily; at least one of two certified weights (10 g and 50 g - class E2) is used.
  4. If the displayed weight differs <0.0001 g from theoretical, the result is documented in a form (date, weight used, displayed weight, signature of analyst) and the balance is released for use.
  5. If the displayed weight differs >0.0001 g from theoretical, both internal (Step 1) and external calibration (Step 2 and followings) have to be repeated once.
  6. If the balance fails internal calibration (Step 2) or external calibration (Step 5), results are documented as in Step 4, but the balance is not released for use; service has to be initiated.
Furthermore both balance and weights are officially verfied by the national standards institute on a yearly basis.

For a controversial discussion have a look at this thread at the Chromatography Forum, and a nice article by C. Burgess and RD. McDowall in LCGC-Europe (2004).

Dif-tor heh smusma 🖖🏼 Довге життя Україна! [image]
Helmut Schütz
[image]

The quality of responses received is directly proportional to the quality of the question asked. 🚮
Science Quotes
chidambarajoshi
★    

Hyderabad-INDIA,
2008-11-08 12:28
(6438 d 17:14 ago)

@ Helmut
Posting: # 2641
Views: 3,705
 

 Balance calibration

Dear V. Pardhasaradhi!
I admit with HS and I have additional information. For Microbalance calibration in Bioanalytical lab E1 or E2 weights can be used. Minimum weight cloud be anywhere between 0.5µg-2mg. That's why you can calibrate internally with 2mg and max 1g. Additionally you can also get the balance calibrated with external agencies and calculate measurement uncertainty at defined frequency.
Regards

Best Regards
Joshi
Ohlbe
★★★

France,
2008-11-08 14:50
(6438 d 14:52 ago)

@ Helmut
Posting: # 2642
Views: 3,686
 

 Balance calibration

Dear Helmut and V. Pardhasaradh,

❝ 3. External weights are also used daily; at least one of two certified

❝ weights (10 g and 50 g - class E2) is used.


If the balance is to be used to prepare stock solutions, where you usually weigh a few mg of reference substance, I would suggest to use control weights in the same range (e.g. 10 mg).

Regards
Ohlbe
UA Flag
Activity
 Admin contact
23,655 posts in 4,993 threads, 1,571 registered users;
147 visitors (0 registered, 147 guests [including 13 identified bots]).
Forum time: 06:42 CEST (Europe/Vienna)

Ignorance more frequently begets confidence
than does knowledge.    Charles Darwin

The Bioequivalence and Bioavailability Forum is hosted by
BEBAC Ing. Helmut Schütz
HTML5