2008-09-30 23:15
(6475 d 11:39 ago)

Posting: # 2454
Views: 3,966
 

 Minimum required standards [Bioanalytics]

Hello,

As per the FDA guidance for method validation, A calibration curve should consist, among other things, of six to eight non-zero samples covering the expected range, including LLOQ. Furthermore, in the acceptance criteria for the run, they require 75%, or a minimum of six standards meeting acceptance criteria. (75% represents 6 out of 8 calibrants).
My question is: if there are 10 standards in the calibration curve, does the minimum of 6 apply or the 75% which is 8 out of 10 does?
I think that they specified a minimum of 6 because it's 75% of the example they gave which is 8 standards.

Regards.

Geokad

Great job you are doing here. :ok:
--
Edit: Category changed. [Ohlbe]
Ohlbe
★★★

France,
2008-10-01 00:02
(6475 d 10:52 ago)

@ Geokad
Posting: # 2455
Views: 3,119
 

 Minimum required standards

Dear Geokad,

My interpretation would be 75 % or 6, whichever is greater (or 75 %, with a minimum of 6).

If you have 6 to 8 calibration samples in your curve: at least 6 should pass. If you have 12 calibration samples: at least 9 should pass.

The problem comes with how to interpret the guidance when 75 % gives you decimals: 75 % of 10 is 7.5, should you go for at least 7, or at least 8 ? I guess the FDA would say at least 8...

❝ Great job you are doing here. :ok:


Thanks !

Regards
Ohlbe
Helmut
★★★
avatar
Homepage
Vienna, Austria,
2008-10-01 12:35
(6474 d 22:19 ago)

@ Ohlbe
Posting: # 2458
Views: 3,153
 

 Minimum required standards

Dear Ohlbe & Geokad,

❝ The problem comes with how to interpret the guidance when 75 % gives you

❝ decimals: 75 % of 10 is 7.5, should you go for at least 7, or at least 8 ?

❝ I guess the FDA would say at least 8...


Agree; just my 2 cents!

Dif-tor heh smusma 🖖🏼 Довге життя Україна! [image]
Helmut Schütz
[image]

The quality of responses received is directly proportional to the quality of the question asked. 🚮
Science Quotes
Geokad
☆    

Canada,
2008-10-01 17:24
(6474 d 17:31 ago)

@ Helmut
Posting: # 2462
Views: 3,149
 

 Minimum required standards

Hi Ohlbe & HS,

Thank you both for your prompt reply. I fully agree.

Regards.

Geokad
UA Flag
Activity
 Admin contact
23,655 posts in 4,993 threads, 1,571 registered users;
128 visitors (0 registered, 128 guests [including 18 identified bots]).
Forum time: 10:55 CEST (Europe/Vienna)

Scientists often have a naïve faith that
if only they could discover enough facts about a problem,
these facts would somehow arrange themselves
in a compelling and true solution.    Theodosius Dobzhansky

The Bioequivalence and Bioavailability Forum is hosted by
BEBAC Ing. Helmut Schütz
HTML5