 Incorporation of additional QC during subject sample analysis. [Bioanalytics]

Dear Members,

Seeking for better clarity ;-)

As stated in RDC27/2012 (ANVISA) "at least two QCs are within the range of measured concentrations." unlike other regulatory agencies where it is stated " At least two QCs should be below AVG Cmax".

How it should be evaluated during subject sample analysis for ANVISA submission studies?

Can we set pre defined criteria like....67% of total periods meet above mentioned criteria or it should be on the basis of average Cmax? :confused:

Kindly share your practice and inputs.

Regards,
Shreyas Goswami.
dshah
★★  

India,
2024-03-12 18:56
(813 d 08:35 ago)

@ shreyas goswami
Posting: # 23906
Views: 5,287
 

 Incorporation of additional QC during subject sample analysis.

Dear Shreyas Goswami

Even ICHM10 follows the same which is adopted by FDA and EMA.
[image]
Regards,
Divyen Shah
shreyas goswami
☆    

India,
2024-03-13 10:30
(812 d 17:01 ago)

@ dshah
Posting: # 23907
Views: 5,197
 

 Incorporation of additional QC during subject sample analysis.

Dear Divyen Shah,

Thanks for the response:-)

Is it acceptable to incorporate additional QC on the basis of AVG Cmax of analyzed subjects. or it should be evaluated by the concentration data of each period separately?????

Regards,
Shreyas Goswami
ElMaestro
★★★

Denmark,
2024-03-14 17:26
(811 d 10:05 ago)

@ shreyas goswami
Posting: # 23910
Views: 5,210
 

 Incorporation of additional QC during subject sample analysis.

Hi shreyas goswami,

❝ Is it acceptable to incorporate additional QC on the basis of AVG Cmax of analyzed subjects. or it should be evaluated by the concentration data of each period separately?????


Rather than thinking in terms of periods, I prefer to shift the focus to runs per se:
For every run you look at the QC response for the second set, and there should be unknowns (=PK-samples) falling above that QC level. Then the QC levels are well chosen.

If a run suddenly has all unknowns below the second QC level, then then right there is an issue you can nominate as a "risk", and thus it is time for a time-out and and update of the QC-levels. You might add a QC level somewhere between first and second level, depending on the actual observations in the run in question but with an eye also towards concentrations in the previous runs.

Pass or fail!
ElMaestro
UA Flag
Activity
 Admin contact
23,653 posts in 4,991 threads, 1,571 registered users;
489 visitors (0 registered, 489 guests [including 30 identified bots]).
Forum time: 04:32 CEST (Europe/Vienna)

I’m all in favor of the democratic principle
that one idiot is as good as one genius, but I draw the line
when someone takes the next step and concludes
that two idiots are better than one genius.    Leo Szilard

The Bioequivalence and Bioavailability Forum is hosted by
BEBAC Ing. Helmut Schütz
HTML5