babu ☆ India, 2008-08-05 14:47 (6531 d 16:47 ago) Posting: # 2148 Views: 4,622 |
|
|
Dear members, I want to suggestion in the below case....... In method validation dilution integrity experiment was established with 1/5 and 1/10 dilution. But in the study sample analysis samples were diluted with 1/2 dilution only. Is it acceptable? kindly explain. thank you, With regards, Bab. |
|
Helmut ★★★ ![]() Vienna, Austria, 2008-08-05 15:39 (6531 d 15:54 ago) @ babu Posting: # 2149 Views: 3,909 |
|
|
Dear Bab. ❝ In method validation dilution integrity experiment was established with ❝ 1/5 and 1/10 dilution. But in the study sample analysis samples were ❝ diluted with 1/2 dilution only. According to common sense, if the method was validated for no dilution, 1:5 and 1:10 dilution it should be possible to go with 1:2 also. But: regulators want to see data, not common sense. ❝ Is it acceptable? It casts a damning light on both the analytical performance and the QAU (why was the 1:2 dilution performed ouside the validated method?). I would go with a partial validation and
— Dif-tor heh smusma 🖖🏼 Довге життя Україна! ![]() Helmut Schütz ![]() The quality of responses received is directly proportional to the quality of the question asked. 🚮 Science Quotes |
|
ElMaestro ★★★ Denmark, 2008-08-06 11:03 (6530 d 20:31 ago) @ Helmut Posting: # 2151 Views: 3,704 |
|
|
❝ But: regulators want to see data, not common sense. Oh how I wish I could understand what goes on in the minds of the regulators.... ![]() EM. |