avinash
★    

India,
2014-07-03 08:49
(4370 d 09:29 ago)

Posting: # 13207
Views: 5,026
 

 Electrochemi­luminescence method [Bioanalytics]

Dear Members,

We have to perform “Thyroid hormone” assay for BA/BE studies, using “Electrochemiluminescence” method. The method was fully validated by manufacturer.

In this method the calibration curve is generated using the Master calibration curve provided by the kit, hence there is no spiking of calibration standards and quality controls.

Please suggest what would be the additional validation parameters that need to be performed to support the FDA method validation requirement?

Thanks & regards,
Avinash Jain
Ohlbe
★★★

France,
2014-07-03 14:08
(4370 d 04:10 ago)

@ avinash
Posting: # 13210
Views: 4,082
 

 Electrochemiluminescence method

Dear Avinash,

❝ The method was fully validated by manufacturer.


For patient follow-up, not for BE trials...

❝ Please suggest what would be the additional validation parameters that need to be performed to support the FDA method validation requirement?


Do your homework: read the draft FDA guidance on bioanalytical method validation, section C, particularly lines 853 to 883. This is just a draft, but I think it provides a good "suggestion".

Regards
Ohlbe
avinash
★    

India,
2014-07-04 13:43
(4369 d 04:34 ago)

@ Ohlbe
Posting: # 13215
Views: 3,981
 

 Electrochemiluminescence method

Dear Ohlbe,

Thanks for the response.

For qualitative clinical diagnostic assay like Electrochemiluminescence do not fully comply with the USFDA guidance, full calibration curve cannot be generated with each run, as Master calibration curve provided by the kit. In case if we prove the appropriateness of calibration by using QC at different concentrations spanning entire range of study sample concentration. Can this method be used for bioequivalence study?

Thanks & regards,
Avinash
UA Flag
Activity
 Admin contact
23,655 posts in 4,993 threads, 1,570 registered users;
163 visitors (0 registered, 163 guests [including 39 identified bots]).
Forum time: 18:18 CEST (Europe/Vienna)

I have finally come to the konklusion
that a good reliable set ov bowels
iz worth more to a man
than enny quantity of brains.    Josh Billings

The Bioequivalence and Bioavailability Forum is hosted by
BEBAC Ing. Helmut Schütz
HTML5