 Recovery [Bioanalytics]

Hi..

We are writing a SOP for Bioanalytical method validation.

As per US FDA's guideline on Bio analytical method validation "Recovery experiments should be performed by comparing the analytical results for extracted samples at three concentrations (low, medium, and high) with unextracted standards that represent 100% recovery"

During validation we demonstrate QCs at HQC, MQC1, MQC2 (lower conc. than MQC1), LQC levels.

Now my doubt is if an additional QC (say MQC3 (lower conc. than MQC2)) is introduced into the run, so that the subject samples meet at least two levels of QCs as per EMEA's guideline recommendation, is there a chance that the regulator may ask us to demonstrate the recovery at that additional QC level also? Is a bracketing approach valid in this case?

Regards
P.S.Srinivas


Edit: Category changed. [Helmut]
Ohlbe
★★★

France,
2014-03-20 12:19
(4084 d 16:06 ago)

@ sarada06884
Posting: # 12677
Views: 5,653
 

 Recovery

Dear Srinivas,

❝ As per US FDA's guideline on Bio analytical method validation "Recovery experiments should be performed by comparing the analytical results for extracted samples at three concentrations (low, medium, and high) with unextracted standards that represent 100% recovery"


Which is actually incorrect: you should compare with samples spiked post-extraction, not with unextracted standards. Otherwise in LC-MS/MS methods what you will calculate is a mix of recovery and matrix effects, not just recovery.

❝ Now my doubt is if an additional QC (say MQC3(lower conc. than MQC2)) is introduced into the run, so that the subject samples meet at least two levels of QCs as per EMEA's guideline recommendation, is there a chance that the regulator may ask us to demonstrate the recovery at that additional QC level also?


I think not. There were even discussions at some bioanalytical conferences lately, where the FDA was present, that checking recovery at 2 levels of concentration (low and high) should be enough. The draft guidance asks you to check at 3 levels of concentration, not at each QC level.

Regards
Ohlbe
Ken Peh
★    

Malaysia,
2014-03-23 08:23
(4081 d 20:02 ago)

@ Ohlbe
Posting: # 12688
Views: 5,529
 

 Recovery

Dear Ohlbe,

❝ Which is actually incorrect: you should compare with samples spiked post-extraction, not with unextracted standards. Otherwise in LC-MS/MS methods what you will calculate is a mix of recovery and matrix effects, not just recovery.


The unextractd standard is drug solution. Right ? Comparing the response of sample after extraction and unextracted standard which is drug solution of same concentration. Do you mean instead of using drug solution, we should compare with sample spiked post-extraction. Kindly elaborate.

What about samples that do not go through extraction step ? Plasma samples treated with just the addition of ACN or methanol before injecting into LCMSMS.
We have been doing recovery for sample treated in this way. Is it necessary to have recovery ? We also compare the response of treated sample and drug solution (untreated sample) having similar concentration.

Appreciate your comment.

Thank you.

Regards,
Ken
nobody
nothing

2014-03-23 12:01
(4081 d 16:24 ago)

@ Ken Peh
Posting: # 12689
Views: 5,502
 

 Recovery

❝ What about samples that do not go through extraction step ? Plasma samples treated with just the addition of ACN or methanol before injecting into LCMSMS.

❝ We have been doing recovery for sample treated in this way. Is it necessary to have recovery ?


You might precipitate in real samples some of the analyte with the protein in your sample when adding ACN or methanol. But in general, as long as you treat standards, QCs and verum samples identically (and have representative empty matrix), in most cases recovery is not very interesting. More a parameter to take care for while debugging, in my opinion...

Kind regards

Kindest regards, nobody
Ohlbe
★★★

France,
2014-03-23 17:28
(4081 d 10:57 ago)

@ nobody
Posting: # 12690
Views: 5,507
 

 Recovery

Dear Nobody,

❝ [...] in most cases recovery is not very interesting. More a parameter to take care for while debugging, in my opinion...


I'd say recovery is more something to check during method development, to optimise your extraction method. But indeed it is of limited interest in method validation. That is, unless indeed the FDA requests it to be done on 6 different sources of matrix, as planned in their draft BMV guideline. I've seen a poster recently mentioning a variable recovery from one lot to the next.

Regards
Ohlbe
Ohlbe
★★★

France,
2014-03-23 17:37
(4081 d 10:48 ago)

@ Ken Peh
Posting: # 12691
Views: 5,562
 

 Recovery

Dear Ken,

❝ Do you mean instead of using drug solution, we should compare with sample spiked post-extraction. Kindly elaborate.


Yes. Actually you can easily combine matrix effects and recovery experiments:
- extract spiked samples in 6 different lots of matrix
- spike the same 6 lots of matrix post-extraction, at the same level of concentration
- prepare neat solutions, at the same level of concentration
- inject all these samples
- compare the response in spiked samples to samples spiked post extraction: recovery
- compare the response to samples spiked post-extraction to neat solutions: matrix factor
- if you compare the response in spiked samples to neat solution: what you see is a kind of overall efficiency, which takes into consideration recovery + matrix effects.

❝ What about samples that do not go through extraction step ? Plasma samples treated with just the addition of ACN or methanol before injecting into LCMSMS.


I agree with Nobody: part of your analyte and IS can get trapped in the precipitate and the recovery may not be 100 %. It depends on the precipitation reagent, the way it is added, how you mix the sample afterwards, etc.

Regards
Ohlbe
Ken Peh
★    

Malaysia,
2014-03-27 20:16
(4077 d 08:09 ago)

@ Ohlbe
Posting: # 12732
Views: 5,427
 

 Recovery

Dear Ohlbe,

Thank you very much for your elaboration.

Regards,
Ken
UA Flag
Activity
 Admin contact
23,424 posts in 4,927 threads, 1,670 registered users;
28 visitors (0 registered, 28 guests [including 10 identified bots]).
Forum time: 05:26 CEST (Europe/Vienna)

Far better an approximate answer to the right question,
which is often vague, than an exact answer to the wrong question,
which can always be made precise.    John W. Tukey

The Bioequivalence and Bioavailability Forum is hosted by
BEBAC Ing. Helmut Schütz
HTML5