 Setting up of Area Threshold Parameter in LC-MS/MS [Bioanalytics]

Dear Group Members,

We use API Sciex LC-MS/MS systems with Analyst 1.4.2 software. During study sample analysis, can we set the Area Threshold parameter based on the LLOQ response? If so, what % of LLOQ response should be the Area Threshold?

Thanks & regards,
Dr. Harish L. Rao
Helmut
★★★
avatar
Homepage
Vienna, Austria,
2007-11-02 18:45
(6809 d 11:28 ago)

@ Dr. Harish L. Rao
Posting: # 1267
Views: 6,536
 

 Area Threshold Parameter in LC-MS/MS

Dear Harish!

❝ During study sample analysis, can we set the Area Threshold

parameter based on the LLOQ response? If so, what % of

❝ LLOQ response should be the Area Threshold?


This is a tricky question - especially for LC-MS/MS...
Since I expect your method uses an internal standard (IS), your LLOQ response is based on the ratio of analyte/IS. Therefore it's difficult to set an area threshold for integration - the data system simly 'does not know' the suitable 'LLOQ response' - which in a strict sense can only be established based on the calibration curve for the particular batch.
What you can do is establish some kind of minimum area of the limit of detection (LOD), which generally is 1/3 - 1/10 of the LLOQ. By this you
  • avoid integration of artefacts (i.e., noise form the system), and
  • have quite a large safety margin not to miss integration of values close to the LLOQ.
In any case visual inspection of all chromatograms is a must.
Although the human brain is a perfect pattern recognition system (especially if you have seen many thousands of chromatograms), you should have an SOP in place guiding you through re-integration in a consistent and traceable way.

Dif-tor heh smusma 🖖🏼 Довге життя Україна! [image]
Helmut Schütz
[image]

The quality of responses received is directly proportional to the quality of the question asked. 🚮
Science Quotes
vishal vayeda
☆    

2008-01-01 09:53
(6749 d 20:20 ago)

@ Helmut
Posting: # 1446
Views: 6,392
 

 Area Threshold Parameter in LC-MS/MS

Dear Sir,

As per my knowldge area threshold require to neglet those area of the chromatogram which are no need to quantify so you can put whatever value as area thresold but it should be less then the area which you found at LLOQ level.
e.g. if the area of LLOQ is 12345 so you can put area thresold value less then the LLOQ area so whatever peak having the area less then area threshold will not reflect in your chromatogram. plese correct me if i am wrong.

Regards,

Vishal Vayeda

--
Edit: Full quote removed. Please see this post! [HS]
Helmut
★★★
avatar
Homepage
Vienna, Austria,
2008-01-02 02:57
(6749 d 03:16 ago)

@ vishal vayeda
Posting: # 1450
Views: 6,105
 

 Area Threshold Parameter (Example)

Dear Vishal!

❝ As per my knowldge area threshold require to neglet those area of the

❝ chromatogram which are no need to quantify ...


So far so good. :cool:

❝ ... so you can put whatever value as area thresold but it should be

❝ less then the area which you found at LLOQ level.


As said in my previous post, it's not that easy - at least for internal standards methods...

❝ e.g. if the area of LLOQ is 12345 so you can put area thresold value less

❝ then the LLOQ area so whatever peak having the area less then area

❝ threshold will not reflect in your chromatogram.


Simple example (sample at LLOQ):
Analyte area 12345
IS area 246900
Ratio analyte/IS 0.05

Now you are loosing 5% somewhere in the sample transfer process, getting
Analyte area 11723
IS area 234555
Ratio analyte/IS 0.05 (=unaffected!)

If you would have set the threshold to 12345, the analyte would not have been integrated (in other words, you would report the sample as BLQ, which is not correct).
Just remember that it's OK at LLOQ to have an inaccuracy of ±20% together with precision of 20%.
If you set the area threshold to the analyte's response calculated at LLOQ for the average internal standard area from the CC, you will miss roughly 1/3 of samples by pure chance!
I agree with Rajendra, that the threshold should be set somewhere between LOD and 50% of LLOQ.

Dif-tor heh smusma 🖖🏼 Довге життя Україна! [image]
Helmut Schütz
[image]

The quality of responses received is directly proportional to the quality of the question asked. 🚮
Science Quotes
rajendra
☆    

India,
2007-12-27 07:34
(6754 d 22:39 ago)

@ Dr. Harish L. Rao
Posting: # 1428
Views: 6,122
 

 Setting up of Area Threshold Parameter in LC-MS/MS

Dear Sir,

Area Threshold Parameter in LC-MS/MS, it should be less than 20% of your LLOQ response.
UA Flag
Activity
 Admin contact
23,655 posts in 4,993 threads, 1,570 registered users;
100 visitors (0 registered, 100 guests [including 10 identified bots]).
Forum time: 07:13 CEST (Europe/Vienna)

The great tragedy of Science – the slaying
of a beautiful hypothesis by an ugly fact.    Thomas Henry Huxley

The Bioequivalence and Bioavailability Forum is hosted by
BEBAC Ing. Helmut Schütz
HTML5