 PK profile outlier [Bioanalytics]

Dear Members,

The following query is with regard to concentration profile:

In subject concentration profile there is a sudden increase in concentration of one sample in elimination phase which is not in line with the rest of the profiles.

Same sample was reanalyzed as a part of investigation and the results were not matching with initial analysis (difference was found to be 175%).

Please suggest how to handle this situation.

Thanks & Regards,
Avinash Jain


Edit: Subject line changed. [Helmut]
Helmut
★★★
avatar
Homepage
Vienna, Austria,
2013-05-02 16:56
(4409 d 07:38 ago)

@ avinash
Posting: # 10526
Views: 4,850
 

 IS response?

Hi Avinash,

which regulatory agency is your “target”?
  • For the FDA and HPFB/TPD you can repeat a sample based on PK grounds and report the repeated value – if you have an SOP for the entire procedure.
  • Bad luck with EMA (and countries adapting its rules: Australia & New Zealand). :not really:
Did you compared the IS response of the original value with the repeat’s? If in the original analysis IS addition was incomplete, you will get a higher analyte/IS ratio → concentration.

Dif-tor heh smusma 🖖🏼 Довге життя Україна! [image]
Helmut Schütz
[image]

The quality of responses received is directly proportional to the quality of the question asked. 🚮
Science Quotes
avinash
★    

India,
2013-05-02 17:09
(4409 d 07:24 ago)

@ Helmut
Posting: # 10527
Views: 4,843
 

 IS response?

Hi

thanks for the reply.

its for FDA submission.

Best regards,
Avinash
Helmut
★★★
avatar
Homepage
Vienna, Austria,
2013-05-02 17:17
(4409 d 07:16 ago)

@ avinash
Posting: # 10528
Views: 4,939
 

 DB Summary Table 9

Hi Avinash!

❝ its for FDA submission.


OK, though FDA accepts PK repeats that doesn’t imply that they like it. I hope you have an SOP. Report the outcome in Table 9 (2013 version).

Other ideas:
  • Do you control the transfer volume(s) in your method? If yes, you could construct an external calibration curve (peak area instead of analyte/IR-ratio vs. concentrations). If the suspect value fits now into the PK profile you have some additional evidence that IS addition was not correct. Don’t forget that an IS only protects against errors in sample transfer – relying on the assumption of error-free IS addition. It’s always a good idea to transfer volumes as accurate as possible (like in an ES method). An IS is no excuse for sloppy work.
    See also this old (and lengthy) thread.
  • You said the suspect value is in the elimination phase. If you have to deal with an FDA assessor who likes EMA’s GLs (note: FDA’s guidance is currently under revision), keep the original value for the calculation of AUCt, but exclude it from the estimation of λz (→ AUC). In the worst case exclude the subject from the AUC-assessment but keep him/her for the comparison of Cmax.

Dif-tor heh smusma 🖖🏼 Довге життя Україна! [image]
Helmut Schütz
[image]

The quality of responses received is directly proportional to the quality of the question asked. 🚮
Science Quotes
AngusMcLean
★★  

USA,
2013-05-15 02:38
(4396 d 21:55 ago)

@ Helmut
Posting: # 10584
Views: 4,672
 

 DB Summary Table 9

Pharmacokinetic outlier

I think it is important to have an SOP in place and issued prior to running the study. It should be defined what the criteria are for identifying a "pharmacokinetic outlier" and repeat of the assay. The repeat data should be examined in accord with the lab. SOP for accepting and rejecting repeat assays.


The FDA will review above.

Angus
UA Flag
Activity
 Admin contact
23,424 posts in 4,927 threads, 1,675 registered users;
62 visitors (0 registered, 62 guests [including 8 identified bots]).
Forum time: 00:34 CEST (Europe/Vienna)

Every man gets a narrower and narrower field of knowledge
in which he must be an expert in order to compete with other people.
The specialist knows more and more about less and less
and finally knows everything about nothing.    Konrad Lorenz

The Bioequivalence and Bioavailability Forum is hosted by
BEBAC Ing. Helmut Schütz
HTML5