j_kevin
☆    

Hong Kong,
2022-12-08 10:41
(885 d 06:22 ago)

Posting: # 23390
Views: 6,886
 

 FDA's HVD SAS Code / Stan­dard Error / CV [Regulatives / Guidelines]

Q1: May I ask why in 2*2*2 cross over design, SE = sqrt(MSE)*sqrt(2/N)?

Here is the SAS code for the analysis of ABE with fully replicate design data.

PROC MIXED;
CLASSES SEQ SUBJ PER TRT;
MODEL Y = SEQ PER TRT/ DDFM=SATTERTH;
RANDOM TRT/TYPE=FA0(2) SUB=SUBJ G;
REPEATED/GRP=TRT SUB=SUBJ;
ESTIMATE 'T vs. R' TRT 1 -1/CL ALPHA=0.1;


Q2: When should I write 1 -1 and when should I write -1 1 in estimate statements?

Here is the covariance output from SAS

Covariance Parameter Estimates
Cov Parm  Subject       Group         Estimate
FA(1,1)   subject                     0.2001
FA(2,1)   subject                     0.1263
FA(2,2)   subject                     0.1507
Residual  subject   formulation Ref   0.01495
Residual  subject   formulation Test  0.03265
G(1,1) = 0.04005, G(2,1)=G(1,2)=0.02527, G(2,2)=0.03866


In replicate design, the intra-subject variabilities are still calculated from the residuals as in a 2-way crossover, using:
IntraCV_T = 100%*sqrt(exp(sig_WT^2)-1)
IntraCV_R = 100%*sqrt(exp(sig_WR^2)-1)

The data include 10 subjects and 5 in seq 1 and 5 in seq2. This is a 2*2*4, 2 sequence 2 treatment and 4 period design, 90% CI of T/R is (0.8450984199, 1.104975448) and point estimate is 0.9663400049.
I tried to use CVfromCI function from R package PowerTOST to calculate CV, and here is my r code:
100*CVfromCI(alpha=0.05, lower=0.8450985339, upper=1.104975299, n=c(5,5), design="2x2x4")
and I get CV = 25.24456.

Q3: Is this CV a pooled CV? If I want to get CVwt and CVwr based on confidence interval, how can I do it using R package

Q4: I try to use the formula 100%*sqrt(exp(sig_W^2)-1) where sig_w^2 = (sig_wt^2+sig_wr^2)/2 to get the same CV(25.24456) based on the covariance output above, but the value is different. Is my calculation wrong? If so, can anyone please tell me the right way to get same CV?

Many thanks!


Edit: Code and results BBCoded; see also this post #10[Helmut]
Helmut
★★★
avatar
Homepage
Vienna, Austria,
2022-12-08 13:09
(885 d 03:53 ago)

@ j_kevin
Posting: # 23392
Views: 5,898
 

 FDA's HVD SAS Code / Stan­dard Error / CV

Hi j_kevin,

❝ Q1: May I ask why in 2*2*2 cross over design, SE = sqrt(MSE)*sqrt(2/N)?


Correct is $$SE=\sqrt{\frac{MSE}{2}\left(\frac{1}{n_1}+\frac{1}{n_2}\right),}\tag{1}$$ where \(\small{n_1}\) and \(\small{n_2}\) are the number of subjects in sequences \(\small{1}\) and \(\small{2}\), respectively.*
If and only if sequences are balanced (\(\small{n_1=n_2\to N=n_1+n_2}\)) you can use $$\eqalign{SE&=\sqrt{\frac{2\,MSE}{N}}\\
&=\sqrt{MSE}\,\cdot\,\sqrt{\frac{2}{N}}}\tag{2}$$ instead. In order to avoid falling into a trap (i.e., using \((2)\) if \(\small{n_1\neq n_2}\)), I strongly recommend to use always \((1)\).

❝ Here is the SAS code for the analysis of ABE with fully replicate design data.


PROC MIXED;

❝ CLASSES SEQ SUBJ PER TRT;

❝ MODEL Y = SEQ PER TRT/ DDFM=SATTERTH;

❝ RANDOM TRT/TYPE=FA0(2) SUB=SUBJ G;

❝ REPEATED/GRP=TRT SUB=SUBJ;

❝ ESTIMATE 'T vs. R' TRT 1 -1/CL ALPHA=0.1;


❝ Q2: When should I write 1 -1 and when should I write -1 1 in estimate statements?


All software use lexical order. Hence, it depends how you coded test and reference, i.e., in the FDA’s SAS-code as T and R. By default the comparison will be done R - T because in lexical order T > R and last one will act as the reference. If you have other ones, you may have to code TRT in ESTIMATE differently.

❝ Here is the covariance output from SAS


Covariance Parameter Estimates

❝ Cov Parm  Subject       Group         Estimate

❝ FA(1,1)   subject                     0.2001

❝ FA(2,1)   subject                     0.1263

❝ FA(2,2)   subject                     0.1507

❝ Residual  subject   formulation Ref   0.01495

❝ Residual  subject   formulation Test  0.03265

❝ G(1,1) = 0.04005, G(2,1)=G(1,2)=0.02527, G(2,2)=0.03866


❝ In replicate design, the intra-subject variabilities are still calculated from the residuals as in a 2-way crossover, using:

❝ IntraCV_T = 100%*sqrt(exp(sig_WT^2)-1)

❝ IntraCV_R = 100%*sqrt(exp(sig_WR^2)-1)


If you remove the %, correct.

❝ The data include 10 subjects and 5 in seq 1 and 5 in seq2. This is a 2*2*4, 2 sequence 2 treatment and 4 period design, 90% CI of T/R is (0.8450984199, 1.104975448) and point estimate is 0.9663400049.

❝ I tried to use CVfromCI function from R package PowerTOST to calculate CV, and here is my r code:

100*CVfromCI(alpha=0.05, lower=0.8450985339, upper=1.104975299, n=c(5,5), design="2x2x4")

❝ and I get CV = 25.24456.


❝ Q3: Is this CV a pooled CV?


Correct, since the 90% CI is based on it.

❝ If I want to get CVwt and CVwr based on confidence interval, how can I do it using R package


That’s not possible because there is an infinite number of combinations of CVwT and CVwR giving the same CVw. Example:

library(PowerTOST)
CV         <- 0.2524456 # CVw (pooled from the – unknown – CVwT and CVwR)
# (random) variance ratios
ratios     <- unique(sort(c(1, exp(runif(n = 10, min = log(0.5), max = log(2))))))
CVs        <- data.frame(var.ratio = ratios,
                         CVwT = NA_real_,
                         CVwR = NA_real_,
                         CVw = NA_real_)
CVs[, 2:3] <- CVp2CV(CV = CV, ratio = ratios)
# pool CVwT and CVwR
CVs[, 4]   <- mse2CV((CV2mse(CVs$CVwT) + CV2mse(CVs$CVwR)) / 2)
print(CVs, row.names = FALSE) # q.e.d.

 var.ratio      CVwT      CVwR       CVw
 0.5431852 0.2108360 0.2887240 0.2524456
 0.5917929 0.2168147 0.2841055 0.2524456
 0.5928429 0.2169383 0.2840082 0.2524456
 0.6628259 0.2246914 0.2777378 0.2524456
 0.8504489 0.2417246 0.2627799 0.2524456
 1.0000000 0.2524456 0.2524456 0.2524456
 1.1854322 0.2632860 0.2411789 0.2524456
 1.3868117 0.2728234 0.2305029 0.2524456
 1.6521422 0.2828633 0.2183837 0.2524456
 1.7311022 0.2854248 0.2151306 0.2524456
 1.8842596 0.2899458 0.2092142 0.2524456


❝ Q4: I try to use the formula 100%*sqrt(exp(sig_W^2)-1) where sig_w^2 = (sig_wt^2+sig_wr^2)/2 to get the same CV(25.24456) based on the covariance output above, but the value is different. Is my calculation wrong? If so, can anyone please tell me the right way to get same CV?


No idea, I don’s speak SAS. AFAIK, FA(2,2) is the subject-by-formulation interaction. 0.1507 is unusually large.



Dif-tor heh smusma 🖖🏼 Довге життя Україна! [image]
Helmut Schütz
[image]

The quality of responses received is directly proportional to the quality of the question asked. 🚮
Science Quotes
j_kevin
☆    

Hong Kong,
2022-12-09 08:28
(884 d 08:35 ago)

@ Helmut
Posting: # 23393
Views: 5,837
 

 FDA's HVD SAS Code / Stan­dard Error / CV

Dear Helmut,

Thanks for your detailed explanation. Based on your comments, I have other questions.Could you please take a took, thanks.

❝ ❝ Q1: May I ask why in 2*2*2 cross over design, SE = sqrt(MSE)*sqrt(2/N)?


❝ Correct is $$SE=\sqrt{\frac{MSE}{2}\left(\frac{1}{n_1}+\frac{1}{n_2}\right),}\tag{1}$$

❝ where \(\small{n_1}\) and \(\small{n_2}\) are the number of subjects in sequences \(\small{1}\) and \(\small{2}\), respectively.*

If and only if sequences are balanced (\(\small{n_1=n_2\to N=n_1+n_2}\)) you can use $$\eqalign{SE&=\sqrt{\frac{2\,MSE}{N}}\\&=\sqrt{MSE}\,\cdot\,\sqrt{\frac{2}{N}}}\tag{2}$$

❝ instead. In order to avoid falling into a trap (i.e., using \((2)\) if \(\small{n_1\neq n_2}\)), I strongly recommend to use always \((1)\).


New Q1: May I ask the formula to calculate SE in a 2*2*4 replicated design? I read your lection but I only find the formula to calculate in 2*2*2 design. Thanks.

❝ ❝ The data include 10 subjects and 5 in seq 1 and 5 in seq2. This is a 2*2*4, 2 sequence 2 treatment and 4 period design, 90% CI of T/R is (0.8450984199, 1.104975448) and point estimate is 0.9663400049.

❝ ❝ I tried to use CVfromCI function from R package PowerTOST to calculate CV, and here is my r code:

❝ ❝ 100*CVfromCI(alpha=0.05, lower=0.8450985339, upper=1.104975299, n=c(5,5), design="2x2x4")

❝ ❝ and I get CV = 25.24456.

❝ ❝

❝ ❝ Q3: Is this CV a pooled CV?


❝ Correct, since the 90% CI is based on it.


New Q3: (1) When we talk about CV, do we usually mean CVwr(within-subject variability of the reference treatment)? Because we need to use whether CVwr > 30% (Swr>0.294) or not to determine which bioequivalence method should use.
(2) If the answer of (1) is yes, then why we need to calculate pooled CV from 90% CI if we only care about CVwr? Or if pooled CV is important, when will we need to use pooled CV to make any decision?
(3) What's the relationship among CV(Sw), CVwr(Swr), CVwt(Swt)? For 2*2*2 design, is that S^2w = (S^2wr + S^2wt)/2, and CV=100*sqrt(exp(S^2w)-1)? Will this formula change in 2*2*4 design? If it will change, could you please write the formula in 2*2*4 design? Thanks.


Edit: Please don’t color your entire text. THX. [Helmut]
Helmut
★★★
avatar
Homepage
Vienna, Austria,
2022-12-09 14:36
(884 d 02:27 ago)

@ j_kevin
Posting: # 23394
Views: 5,714
 

 CV limbo

Hi j_kevin,

❝ New Q1: May I ask the formula to calculate SE in a 2*2*4 replicated design? I read your lection but I only find the formula to calculate in 2*2*2 design.


It’s the same one.

❝ New Q3: (1) When we talk about CV, do we usually mean CVwr(within-subject variability of the reference treatment)? Because we need to use whether CVwr > 30% (Swr>0.294) or not to determine which bioequivalence method should use.


When we talk about the CV in a crossover (simple 2×2×2, Higher-Order, or replicate) design, generally the intra-subject CV (CVw) is meant. In parallel designs it’s the total CV (pooled from CVw and CVb). Only in reference-scaling we are interested in the intra-subject CVs of treatments (CVwT and CVwR) as well. Of course, both are only accessible in full replicate designs. In the bloody partial replicate design we get only the latter. The former is nice to know (esp. in pilot studies when we want to estimate the sample size for a pivotal study; see there for an example).

❝ (2) If the answer of (1) is yes, then why we need to calculate pooled CV from 90% CI if we only care about CVwr? Or if pooled CV is important, when will we need to use pooled CV to make any decision?


The implemented methods of reference-scaling (RSABE for the FDA and China, ABEL in all other jurisdictions) are frameworks, where the decision whether RSABE/ABEL or ABE has to be applied hinges on the variability of the reference (see here and there).
If RSABE is applicable (\(\small{s_{\text{wR}}^2\ge0.294}\)), we are not interested in the confidence interval; we assess only whether the upper bound of the linearized criterion is ≤0 and the PE lies within 80.00 – 125.00%. Only in ABE (\(\small{s_{\text{wR}}^2<0.294}\)) we assess the 90% CI.
Contrary to that, in ABEL we assess always the 90% CI (if \(\small{CV_{\text{wR}}>30\%}\) within the expanded limits or within 80.00 – 125.00%, otherwise).

❝ (3) What's the relationship among CV(Sw), CVwr(Swr), CVwt(Swt)? For 2*2*2 design, is that S^2w = (S^2wr + S^2wt)/2, and CV=100*sqrt(exp(S^2w)-1)?


If the study is balanced, correct. Otherwise we would have to weigh the variances by the degrees of freedom. See also this post.

❝ Will this formula change in 2*2*4 design? If it will change, could you please write the formula in 2*2*4 design?


First of all, you don’t need CVw in replicate designs indended for reference-scaling. A clumsy attempt for the EMA’s ‘Data set I’ (imbalanced and incomplete):

library(replicateBE)
library(PowerTOST)
# Calculate ABE of the EMA’s reference data set 1
# and assign the relevant results to a data frame

x       <- method.A(print = FALSE, details = TRUE,
                    data = rds01)[, c(1, 3:6, 11:14, 19:20)]
# Calculate an ANOVA (all effects fixed)
y       <- anova(lm(log(PK) ~ sequence +
                              subject %in% sequence +
                              period +
                              treatment, data = rds01))["Residuals", 1:3]
# CVw from ANOVA
CVw1    <- mse2CV(y["Residuals", "Mean Sq"])
# Extract the number of subjects / sequence
nseq    <- as.integer(strsplit(x[["Sub/seq"]], "|", fixed = TRUE)[[1]])
# Extract the number of subjects with repeated administrations / treatment
ntrt    <- setNames(x[3:4], c("T", "R"))
df      <- setNames(ntrt - 2, c("T", "R"))
# Calculate variances weighted by degrees of freedom
s2w.df  <- setNames(c(x[["swT"]]^2 * df[["T"]], x[["swR"]]^2 * df[["R"]]),
                    c("s2wT.df", "s2wR.df"))
# Pool variances
s2w     <- sum(s2w.df) / sum(df) # That’s wrong - what is correct?
# CVw from pooled variances
CVw2    <- mse2CV(s2w)
# CVw from 90% CI
CVw3    <- CI2CV(lower = x[["CL.lo(%)"]],
                 upper = x[["CL.hi(%)"]],
                 design = "2x2x4", n = nseq, alpha = 0.05)
print(x, digits = 5, row.names = FALSE)
cat(paste0("Design: ", x[["Design"]],
    ", subjects / sequence: ", x[["Sub/seq"]], "\n"),
    sprintf("CVwR from Method A         = %.2f%%", x[["CVwR(%)"]]), "\n",
    sprintf("CVwT from Method A         = %.2f%%", x[["CVwT(%)"]]), "\n",
    sprintf("CVw  from ANOVA MSE        = %.2f%%", 100 * CVw1), "\n",
    sprintf("CVw  from pooled variances = %.2f%%", 100 * CVw2), "\n",
    sprintf("CVw  from CI               = %.2f%%", 100 * CVw3), "\n")

    Design  n nTT nRR Sub/seq CVwT(%) CVwR(%)     swT     swR CL.lo(%) CL.hi(%)
 TRTR|RTRT 77  71  73   39|38  35.157  46.964 0.34138 0.44645   107.11   124.89

Design: TRTR|RTRT, subjects / sequence: 39|38
 CVwR from Method A         = 46.96%
 CVwT from Method A         = 35.16%
 CVw  from ANOVA MSE        = 41.65%
 CVw  from pooled variances = 41.45%
 CVw  from CI               = 42.58%

As usual, I can be wrong. Such data are tricky. Definitely correct is only CVw from the ANOVA’s MSE.

With rds25 of package replicateBE, which is both balanced and complete:

    Design  n nTT nRR Sub/seq CVwT(%) CVwR(%)     swT     swR CL.lo(%) CL.hi(%)
 TRTR|RTRT 70  70  70   35|35  46.539  82.805 0.44277 0.72261   77.928   98.102

Design: TRTR|RTRT, subjects / sequence: 35|35
 CVwR from Method A         = 82.81%
 CVwT from Method A         = 46.54%
 CVw  from ANOVA MSE        = 63.61%
 CVw  from pooled variances = 65.73%
 CVw  from CI               = 63.61%

Now the CVw from the ANOVA’s MSE and the one from the CI by the function CI2CV() agree. Hence, my code for pooling variances is wrong.

To complicate things: CVwT and CVwR could be directly estimated from the FDA’s mixed-effects model (as you did in your OP). With the EMA’s ‘Data set I’ we get CVwR = 47.328% and CVwT = 35.293%. However, this is not what the FDA wants. Instead, it has to be estimat­ed by an ANOVA of the reference treatment’s data with subject and sequence as factors. Then we get CVwR = 47.627%. Different, right? The REML PE of the mixed-effects model is 115.658% (90% CI: 107.104 – 124.894%). However, if we apply the FDA’s method acc. to the guidances for RSABE, we get a PE of 115.461% (90% CI: 106.386 – 125.311%). Different, again…
As Mittyri once wrote: “Welcome to the hell of mixed-effects modeling.”

BTW, you are not the first asking for CVw in replicate designs. It’s on our todo-list for the next release of the package replicateBE.

Dif-tor heh smusma 🖖🏼 Довге життя Україна! [image]
Helmut Schütz
[image]

The quality of responses received is directly proportional to the quality of the question asked. 🚮
Science Quotes
ElMaestro
★★★

Denmark,
2022-12-09 16:49
(884 d 00:14 ago)

@ Helmut
Posting: # 23395
Views: 5,692
 

 CV limbo

Hi all,

this question boils down to "something" which is not so easy and straightforward.

First of all, when going full mixed there is to the best of my knowledge no particular relevance of a CI2CV kind of function. It provides an answer to something which is not really asked, but which may be identical to the answer truly sought in case of balance (not even sure about this last part, need to read up on things).

Now for the remainder, and this is everything based on the linear model as applied to replicated BE data, please bear in mind:

1. When we evaluate a subject, the guideline says the subject only contributes if she/he provides at least one T and one R measurement.
...but...

2. When we estimate swr, there is no equivalent wording. Look for example at subject 24 of EMA dataset I. The subject contributes with only one ref measurement. But in the grand scheme of things such a subject does not contribute to the within-ref (try and fit dataset I for swr with and without that subject). The subject's data is estimated with no residual because there is only one measurement. So, as someone wrote in another thread in another galaxy, it is all in the df's. The estimate of swr (if we go the anova way as per EMA) has 71 dfs, and 69 for swt.
But plugging the MSEs and df's into the equation for pooled variances still does not get you where you want to be (a pooled MSE reflecting the overall anova MSE if I got you right). And that's because of the fixed effect estimates that are allowed vary between the two anova's. Think about it. :-)

So, if you are looking for 'manual' and simple equations that allow you to swap back and forth between swr, swt and pooled variances, then I am afraid you are out of luck and you may be asking too much. You can easier get what you are looking for by working backwards from the matrix forms of the linear model.

Pass or fail!
ElMaestro
j_kevin
☆    

Hong Kong,
2022-12-14 14:12
(879 d 02:50 ago)

@ Helmut
Posting: # 23402
Views: 5,580
 

 CV limbo

Dear Helmut,

Thanks for your detailed explanation.

❝ When we talk about the CV in a crossover (simple 2×2×2, Higher-Order, or replicate) design, generally the intra-subject CV (CVw) is meant.


Question: Is CVw a pooled CV? So the CV we usually talk about is the pooled CV? The reason why I ask this question is when we do sample size calculation of bioequivalence study in clinical trails, I always see people write the CV of Cmax/AUC is XX%, I'm not sure whether the CV means CVw or CVwr.

❝ To complicate things: CVwT and CVwR could be directly estimated from the FDA’s mixed-effects model (as you did in your OP).

❝ If the study is balanced, correct. Otherwise we would have to weigh the variances by the degrees of freedom. See also this post.


PROC MIXED;
CLASSES SEQ SUBJ PER TRT;
MODEL Y = SEQ PER TRT/ DDFM=SATTERTH;
RANDOM TRT/TYPE=FA0(2) SUB=SUBJ G;
REPEATED/GRP=TRT SUB=SUBJ;
ESTIMATE 'T vs. R' TRT 1 -1/CL ALPHA=0.1;


Question: Based on the code provided by FDA, I can get two residuals which are Swt and Swr. If I need to calculate confidence interval, then I should first determine whether the design is balanced or not, and then find appropriate way to calculate Sw based on Swt and Swr. Then use Sw to calculate CVw, then use CVw to calculate confidence interval. I can follow the link you provided above. Are those steps correct? Thanks.


Edit: Standard quotes restored; see also this post #8[Helmut]
j_kevin
☆    

Hong Kong,
2023-01-02 04:06
(860 d 12:57 ago)

@ Helmut
Posting: # 23414
Views: 5,261
 

 CV limbo

Dear Helmut,

Could you please take a look at my questions above. Many thanks.
j_kevin
☆    

Hong Kong,
2023-01-02 06:07
(860 d 10:55 ago)

@ Helmut
Posting: # 23415
Views: 5,216
 

 CV limbo

Dear Helmut,

Please neglect my previous first question and I change it a little bit. Looking forwatd to your reply. Thanks.

❝ When we talk about the CV in a crossover (simple 2×2×2, Higher-Order, or replicate) design, generally the intra-subject CV (CVw) is meant. In parallel designs it’s the total CV (pooled from CVw and CVb). Only in reference-scaling we are interested in the intra-subject CVs of treatments (CVwT and CVwR) as well. Of course, both are only accessible in full replicate designs. In the bloody partial replicate design we get only the latter. The former is nice to know (esp. in pilot studies when we want to estimate the sample size for a pivotal study; see there for an example).


Question:
In simple 2*2*2 cross over design, Can we still get Swr or Swt? Or we can only get those two values in replicate design? Is RSABE method can only be applied in replicate design?
In parallel design, CV is the total(pooled) CV (pooled from CVw and CVb). Why we need to calculate pooled CV in parallel design? Will we use that pooled CV to calculate sample size? Or we only need to use CVw to calculate sample size.

❝ ❝ To complicate things: CVwT and CVwR could be directly estimated from the FDA’s mixed-effects model (as you did in your OP).

❝ ❝ If the study is balanced, correct. Otherwise we would have to weigh the variances by the degrees of freedom. See also this post.


PROC MIXED;
CLASSES SEQ SUBJ PER TRT;
MODEL Y = SEQ PER TRT/ DDFM=SATTERTH;
RANDOM TRT/TYPE=FA0(2) SUB=SUBJ G;
REPEATED/GRP=TRT SUB=SUBJ;
ESTIMATE 'T vs. R' TRT 1 -1/CL ALPHA=0.1;


Question: If the design is replicated design and based on the code provided by FDA, I can get two seperate residuals which are Swt and Swr. However, the calculation of confidence intervcal is based on Sw instead of Swt or Swr. Hence, I should first determine whether the design is balanced or not, and then find appropriate way to calculate Sw based on Swt and Swr. Then use Sw to calculate CVw, then use CVw to calculate confidence interval. Are those steps correct? Thanks.


Edit: Standard quotes restored; see also this post #8[Helmut]
Helmut
★★★
avatar
Homepage
Vienna, Austria,
2023-01-04 12:39
(858 d 04:24 ago)

@ j_kevin
Posting: # 23416
Views: 5,128
 

 CV limbo

Hi j_kevin,

for background about variances accessible in different designs see there.

❝ In simple 2*2*2 cross over design, Can we still get Swr or Swt?


No. We get only the within subject variance \(\small{s_\text{w}^2}\) (and, of course, the between subject variance \(\small{s_\text{b}^2}\) as well).
As an aside, the model for BE assumes homoscedasticity \((\small{s_\text{wT}^2\equiv s_\text{wR}^2})\). Bad luck, if \(\small{s_\text{wR}^2>s_\text{wT}^2}\). The ‘good’ test product will be punished by the ‘bad’ referrence.

❝ Or we can only get those two values in replicate design? Is RSABE method can only be applied in replicate design?


Yes and yes.

❝ In parallel design, CV is the total(pooled) CV (pooled from CVw and CVb). Why we need to calculate pooled CV in parallel design? Will we use that pooled CV to calculate sample size? Or we only need to use CVw to calculate sample size.


If you have data of a previous study in a parallel design, you get just the total CV anyway. Only if you have a crossover study (\(\small{CV_\text{w}}\) and \(\small{CV_\text{b}}\)) you have to pool variances in order to estimate the sample size of a study in a parallel design.

❝ If the design is replicated design and based on the code provided by FDA, I can get two seperate residuals which are Swt and Swr.


Correct.

❝ However, the calculation of confidence intervcal is based on Sw instead of Swt or Swr.


See the applicable decision scheme (i.e., whether to use ABE or RSABE). For obscure reasons the FDA does not use \(\small{s_\text{wR}}\) estimated from the mixed-effects model, but from intra-subject contrasts of the reference treatment with sequence as a factor. See this post for an example.

❝ Hence, I should first determine whether the design is balanced or not, and then find appropriate way to calculate Sw based on Swt and Swr. Then use Sw to calculate CVw, then use CVw to calculate confidence interval. Are those steps correct?


No. Follow the guidance.

Dif-tor heh smusma 🖖🏼 Довге життя Україна! [image]
Helmut Schütz
[image]

The quality of responses received is directly proportional to the quality of the question asked. 🚮
Science Quotes
UA Flag
Activity
 Admin contact
23,424 posts in 4,927 threads, 1,669 registered users;
113 visitors (0 registered, 113 guests [including 9 identified bots]).
Forum time: 18:03 CEST (Europe/Vienna)

Lack of clarity is always a sign of dishonesty.    Celia Green

The Bioequivalence and Bioavailability Forum is hosted by
BEBAC Ing. Helmut Schütz
HTML5