jag009
★★★

NJ,
2013-12-19 18:41
(4575 d 13:26 ago)

(edited on 2013-12-19 20:13)
Posting: # 12081
Views: 11,646
 

 tada: Another issue with FDA's RSABE (please confirm?) [RSABE / ABEL]

Hi guys,

It seems like I encountered another problem with FDA's SAS code for calculating 90% CI using the PROC Mixed procedure. The parameters did not qualify for RSABE assessment. This time I don't have any missing data. Sorry Helmut if the data is too long. dataset is space limited. SAS and WinNonlin gurus, please confirm?

My SAS output showed:

For AUCt: WARNING: Did not converge.

For AUCi (I don't think this is a problem(?)):
NOTE: Convergence criteria met but final hessian is not positive definite.
NOTE: Asymptotic variance matrix of covariance parameter estimates has been found to be singular and a generalized inverse was used. Covariance parameters with zero variance do not contribute to degrees of freedom computed by DDFM=SATTERTH.


The fields are: Subject, Sequence, Period, Treatment, AUCt, AUCi
Sequence: 1 = AB1B2, 2=B1AB2, 3=B1B2A

1 2 2 1 1986.987 2028.539
2 3 3 1 2411.950 2526.036
3 1 1 1 1690.350 1770.166
4 3 3 1 5814.296 5922.447
5 2 2 1 4414.033 4882.854
6 1 1 1 7464.493 8706.046
7 1 1 1 6134.403 6354.843
8 2 2 1 1613.515 1691.363
9 3 3 1 3448.915 3688.104
10 3 3 1 1290.754 1389.463
11 1 1 1 6401.609 6489.934
12 2 2 1 4115.371 4179.525
13 2 2 1 2440.672 2711.280
14 3 3 1 940.629 1326.512
15 1 1 1 2390.001 2507.927
16 2 2 1 1011.974 1084.656
17 3 3 1 1425.998 2071.054
18 1 1 1 3756.342 3818.330
19 2 2 1 2788.777 2876.920
20 3 3 1 4482.295 4569.567
21 1 1 1 4765.237 5033.406
22 1 1 1 2818.664 2963.035
23 2 2 1 3721.059 5315.704
24 3 3 1 1651.403 2359.010
25 3 3 1 4550.919 4602.326
26 1 1 1 3249.760 3402.651
27 2 2 1 1359.468 1487.107
28 1 1 1 4084.693 4440.407
29 3 3 1 632.116 655.684
30 2 2 1 1996.837 2265.452
31 3 3 1 3786.507 4359.212
32 2 2 1 2292.445 2498.451
33 1 1 1 3688.778 3936.592
34 1 1 1 5123.381 5191.885
35 2 2 1 3075.198 3498.299
36 3 3 1 3700.028 3871.734
37 1 1 1 3927.375 4163.267
38 3 3 1 3888.209 4011.221
40 3 3 1 3067.678 3250.625
41 1 1 1 2324.201 2458.099
43 1 1 1 1569.884 2435.485
44 3 3 1 3508.294 4372.614
45 2 2 1 1854.899 2213.825
46 1 1 1 4608.492 4713.071
47 2 2 1 1571.500 1902.432
48 3 3 1 2675.384 2755.825
49 2 2 1 2078.712 4101.855
50 1 1 1 2638.387 3046.370
51 3 3 1 3938.841 4051.813
52 2 2 1 4639.706 5237.773
53 1 1 1 6743.505 7953.264
54 3 3 1 5228.621 5486.796
55 2 2 1 1523.521 1610.193
56 1 1 1 3919.978 3967.199
57 3 3 1 2525.702 2802.588
59 3 3 1 5293.810 5518.208
60 2 2 1 5762.699 7522.105
61 1 1 1 1791.163 1935.079
62 2 2 1 3560.518 3890.326
63 3 3 1 2018.500 2167.172
64 3 3 1 639.486 750.382
65 2 2 1 8435.241 8598.730
66 1 1 1 2335.596 2789.875
67 2 2 1 1854.219 4973.702
68 1 1 1 4767.972 4886.123
69 3 3 1 8523.859 8614.662
70 3 3 1 1992.374 2076.292
71 1 1 1 5891.358 6048.819
72 2 2 1 3509.377 3631.466
73 1 1 1 2300.501 2385.739
74 3 3 1 3014.357 3327.482
76 1 1 1 2190.961 3533.236
77 3 3 1 5108.458 5209.829
78 2 2 1 2454.949 2723.965
79 1 1 1 8335.814 8443.137
80 2 2 1 4809.817 4900.255
81 3 3 1 6255.335 6785.737
82 1 1 1 713.013 3248.165
83 3 3 1 1777.401 2419.373
84 2 2 1 2454.707 2554.941
85 1 1 1 2905.955 2959.859
86 2 2 1 950.031 1347.994
87 3 3 1 3072.612 3169.770
88 1 1 1 1404.276 1565.716
89 3 3 1 5634.534 5805.369
90 2 2 1 1972.698 3015.333
91 1 1 1 6044.275 6317.111
92 3 3 1 4140.777 4194.132
93 2 2 1 2587.609 3041.097
94 3 3 1 2981.478 3308.986
95 2 2 1 3161.578 3361.797
96 1 1 1 4359.266 4481.890
97 2 2 1 6445.984 6557.993
98 3 3 1 3313.374 3419.453
99 1 1 1 1752.319 2279.186
100 2 2 1 3157.205 3336.087
103 2 2 1 1646.301 1700.174
104 1 1 1 4166.170 4357.331
106 1 1 1 2845.445 3068.462
107 3 3 1 2098.512 3342.156
109 1 1 1 1968.416 2025.039
110 3 3 1 2142.577 3182.653
111 2 2 1 5322.128 5591.891
112 3 3 1 1570.012 1740.816
113 2 2 1 2602.400 2654.715
114 1 1 1 2509.354 2671.323
115 1 1 1 4391.295 4540.127
117 2 2 1 2947.856 3175.737
118 3 3 1 2847.161 2930.837
119 1 1 1 4477.920 4525.512
120 2 2 1 1252.486 1319.169
121 3 3 1 3919.694 3979.781
122 1 1 1 3131.665 3495.320
123 2 2 1 3966.638 4562.475
124 1 1 1 9507.404 13363.943
125 3 3 1 5163.216 5931.813
126 2 2 1 9142.358 9341.179
1 2 1 2 1393.955 1696.903
1 2 3 2 3505.870 3608.707
2 3 1 2 3736.474 4020.712
2 3 2 2 2515.933 2701.962
3 1 2 2 4269.521 5560.004
3 1 3 2 1944.560 2193.368
4 3 1 2 10714.492 10827.848
4 3 2 2 3366.038 3503.346
5 2 1 2 4425.526 4747.656
5 2 3 2 3802.169 4404.318
6 1 2 2 4814.234 5175.735
6 1 3 2 6650.650 7587.377
7 1 2 2 10340.443 10534.260
7 1 3 2 4363.923 4516.547
8 2 1 2 1541.656 1593.710
8 2 3 2 2096.269 2423.027
9 3 1 2 3883.217 3951.200
9 3 2 2 3469.208 3679.107
10 3 1 2 1158.412 1598.101
10 3 2 2 1684.590 1771.260
11 1 2 2 8529.765 8652.602
11 1 3 2 8589.721 8724.174
12 2 1 2 2838.978 2955.794
12 2 3 2 2558.276 2662.546
13 2 1 2 2825.109 3045.681
13 2 3 2 1832.087 1997.048
14 3 1 2 1692.009 1754.715
14 3 2 2 2179.266 2322.747
15 1 2 2 3954.950 4017.519
15 1 3 2 2406.415 2448.923
16 2 1 2 1864.904 2008.329
16 2 3 2 1010.471 1197.877
17 3 1 2 1581.669 2339.947
17 3 2 2 1703.455 1782.926
18 1 2 2 2641.073 2822.153
18 1 3 2 2077.247 2141.970
19 2 1 2 3443.842 3512.693
19 2 3 2 2523.653 2742.111
20 3 1 2 2329.852 2928.550
20 3 2 2 3633.168 3679.938
21 1 2 2 5309.433 6265.735
21 1 3 2 3486.541 3589.547
22 1 2 2 2300.887 2612.660
22 1 3 2 2822.703 2961.982
23 2 1 2 2932.020 3026.017
23 2 3 2 3180.072 3275.972
24 3 1 2 4939.731 5099.487
24 3 2 2 6323.412 6404.058
25 3 1 2 5245.106 5302.180
25 3 2 2 5835.787 5992.491
26 1 2 2 1825.066 2010.360
26 1 3 2 3176.108 3287.679
27 2 1 2 1246.217 1897.647
27 2 3 2 1483.842 2159.919
28 1 2 2 3339.691 3549.645
28 1 3 2 2828.209 2959.100
29 3 1 2 863.188 890.382
29 3 2 2 650.852 675.257
30 2 1 2 1668.825 2028.178
30 2 3 2 2208.537 2561.781
31 3 1 2 4622.568 5118.419
31 3 2 2 4076.484 4564.566
32 2 1 2 3806.165 4177.127
32 2 3 2 3107.501 4118.659
33 1 2 2 2800.694 3202.677
33 1 3 2 3571.575 3772.260
34 1 2 2 4013.862 4091.954
34 1 3 2 3715.673 3833.044
35 2 1 2 2436.009 2782.867
35 2 3 2 3939.625 4241.525
36 3 1 2 2945.222 3078.332
36 3 2 2 4257.185 4311.668
37 1 2 2 4357.906 4472.984
37 1 3 2 3457.290 4137.185
38 3 1 2 3342.971 3439.184
38 3 2 2 4758.602 5012.171
40 3 1 2 3293.343 3709.858
40 3 2 2 1754.860 1816.881
41 1 2 2 2317.643 3100.998
41 1 3 2 2203.879 2497.227
43 1 2 2 1545.947 2316.025
43 1 3 2 1286.978 1418.360
44 3 1 2 4243.010 6262.912
44 3 2 2 3644.791 4708.170
45 2 1 2 2120.351 2239.488
45 2 3 2 2288.949 2401.353
46 1 2 2 5695.575 5798.696
46 1 3 2 6462.232 6657.683
47 2 1 2 2680.581 2762.981
47 2 3 2 2478.744 2888.345
48 3 1 2 4035.800 4110.881
48 3 2 2 2688.206 2827.054
49 2 1 2 1714.493 4334.532
49 2 3 2 2119.548 2453.837
50 1 2 2 5800.021 5888.889
50 1 3 2 1949.246 2192.628
51 3 1 2 4231.754 4738.621
51 3 2 2 3986.447 4385.957
52 2 1 2 3029.317 3674.960
52 2 3 2 4321.629 4592.441
53 1 2 2 7883.749 8101.335
53 1 3 2 5953.973 6241.813
54 3 1 2 6055.951 6502.790
54 3 2 2 5602.754 5859.172
55 2 1 2 3339.720 3426.866
55 2 3 2 3288.800 3391.971
56 1 2 2 4609.794 4841.331
56 1 3 2 3041.170 3261.202
57 3 1 2 2413.777 2767.540
57 3 2 2 2056.104 2255.159
59 3 1 2 3297.598 3764.997
59 3 2 2 2597.781 2743.093
60 2 1 2 6415.028 6636.392
60 2 3 2 6536.282 6912.101
61 1 2 2 1977.866 2070.410
61 1 3 2 1992.784 2083.171
62 2 1 2 4262.586 5392.059
62 2 3 2 4156.121 4284.027
63 3 1 2 1411.091 1473.652
63 3 2 2 1347.625 1522.727
64 3 1 2 814.878 991.637
64 3 2 2 1017.093 1101.514
65 2 1 2 6888.502 6959.243
65 2 3 2 8521.035 8602.048
66 1 2 2 3502.722 3712.982
66 1 3 2 3003.672 3202.768
67 2 1 2 4786.897 4921.378
67 2 3 2 2451.677 2517.846
68 1 2 2 3734.920 4015.660
68 1 3 2 2799.735 3043.947
69 3 1 2 4167.764 4316.579
69 3 2 2 5235.564 5297.359
70 3 1 2 2919.263 3048.971
70 3 2 2 2298.465 2520.636
71 1 2 2 6084.365 6157.185
71 1 3 2 5605.379 5634.624
72 2 1 2 2238.601 2377.907
72 2 3 2 2065.535 2147.259
73 1 2 2 2357.455 2451.162
73 1 3 2 3319.716 3466.406
74 3 1 2 1805.151 3700.507
74 3 2 2 3633.153 3666.874
76 1 2 2 4431.239 4492.829
76 1 3 2 3306.066 3368.552
77 3 1 2 5101.865 6838.827
77 3 2 2 3463.045 3787.538
78 2 1 2 3672.702 3884.331
78 2 3 2 2935.212 3729.577
79 1 2 2 5364.212 5441.344
79 1 3 2 8249.717 8554.513
80 2 1 2 3998.482 4184.930
80 2 3 2 4080.722 4298.298
81 3 1 2 5230.484 5339.276
81 3 2 2 5588.193 6372.836
82 1 2 2 2076.243 2254.340
82 1 3 2 1830.994 2790.836
83 3 1 2 1578.507 2105.984
83 3 2 2 2229.508 3109.140
84 2 1 2 2797.427 2880.782
84 2 3 2 2197.208 2429.935
85 1 2 2 5033.550 5168.372
85 1 3 2 3192.892 3358.880
86 2 1 2 1233.338 1322.313
86 2 3 2 2057.456 2341.459
87 3 1 2 2707.480 2782.544
87 3 2 2 2787.796 2894.282
88 1 2 2 1010.611 1269.911
88 1 3 2 1301.015 1393.945
89 3 1 2 5737.302 5818.926
89 3 2 2 6840.293 6910.318
90 2 1 2 2254.840 2708.719
90 2 3 2 1909.054 1984.713
91 1 2 2 6091.191 6289.288
91 1 3 2 8476.264 8943.856
92 3 1 2 5577.091 5986.726
92 3 2 2 4511.606 4669.532
93 2 1 2 831.347 1478.894
93 2 3 2 1404.214 1759.693
94 3 1 2 3877.871 3965.294
94 3 2 2 3317.100 3429.038
95 2 1 2 1948.567 2021.957
95 2 3 2 2440.997 2540.043
96 1 2 2 4032.920 4082.632
96 1 3 2 5007.126 5081.079
97 2 1 2 4966.626 5123.627
97 2 3 2 3792.665 3959.180
98 3 1 2 3390.648 3612.219
98 3 2 2 4496.329 4710.963
99 1 2 2 1821.359 2069.192
99 1 3 2 1328.736 1458.837
100 2 1 2 2254.175 2347.780
100 2 3 2 3602.854 3653.128
103 2 1 2 667.938 994.301
103 2 3 2 1881.014 1996.641
104 1 2 2 4412.429 4807.663
104 1 3 2 7269.143 7710.601
106 1 2 2 5727.482 5869.246
106 1 3 2 2489.764 2842.016
107 3 1 2 2522.239 2957.417
107 3 2 2 3220.742 3709.728
109 1 2 2 1543.887 1814.762
109 1 3 2 1310.696 1408.024
110 3 1 2 3831.757 4083.170
110 3 2 2 3059.492 3752.390
111 2 1 2 2419.770 2549.026
111 2 3 2 4323.227 4395.120
112 3 1 2 2309.067 2424.960
112 3 2 2 2908.746 3013.112
113 2 1 2 3003.194 3337.953
113 2 3 2 3831.171 4018.417
114 1 2 2 2856.754 2929.591
114 1 3 2 3209.380 3390.930
115 1 2 2 2751.360 2856.454
115 1 3 2 5202.542 5586.009
116 3 1 2 2157.130 2331.448
116 3 2 2 2301.391 2813.132
117 2 1 2 2702.549 2801.745
117 2 3 2 2083.111 2226.822
118 3 1 2 4113.503 4268.438
118 3 2 2 4470.444 4773.996
119 1 2 2 2577.045 2783.993
119 1 3 2 3534.523 3606.611
120 2 1 2 2235.692 2274.964
120 2 3 2 1367.220 1480.401
121 3 1 2 3383.035 3466.498
121 3 2 2 5317.416 6265.831
122 1 2 2 4307.590 4626.598
122 1 3 2 6508.045 7454.215
123 2 1 2 4569.969 4952.604
123 2 3 2 4789.485 5290.609
124 1 2 2 5837.689 7092.132
124 1 3 2 5066.133 6194.666
125 3 1 2 5285.137 5346.401
125 3 2 2 2468.586 3700.238
126 2 1 2 4807.900 5011.201
126 2 3 2 4517.280 4559.176


Thanks

John
Helmut
★★★
avatar
Homepage
Vienna, Austria,
2013-12-20 05:13
(4575 d 02:54 ago)

@ jag009
Posting: # 12083
Views: 10,017
 

 Phoenix/WinNonlin: not confirmed

Hi John,

seems to be a pivotal study this time? PHX/WNL6.3: No problem with FA0(2), converged after seven iterations, no warnings.

        PE        90% CI      CVWR
AUCt: 96.77%  91.88–101.93%  27.45%
AUCi: 99.71%  95.25–104.38%  23.23%


Take the CVs with a grain of salt. Exactly the same results with FA0(1). Would you mind trying that in SAS as well?

Dif-tor heh smusma 🖖🏼 Довге життя Україна! [image]
Helmut Schütz
[image]

The quality of responses received is directly proportional to the quality of the question asked. 🚮
Science Quotes
jag009
★★★

NJ,
2013-12-20 18:12
(4574 d 13:55 ago)

@ Helmut
Posting: # 12089
Views: 9,917
 

 Phoenix/WinNonlin: not confirmed

Hi Helmut,

When I switched to FA0(1) then it worked. using FA0(1):

AUCt: 96.77% 91.88-101.93%
AUCi: 99.71% 95.25-104.38%

So same as yours :-)

FA0(2) works for AUCi but no convergence for AUCt.
For AUCi, results with FA0(2) was the same as FA0(1).

Question is why FA0(2) didn't work? :confused: I am not a saint in statistics yet so I don't know :-D

❝ seems to be a pivotal study this time?


Yes, they didn't listen and went with 3/4 of the subjects (of this current panel) last time, guess what happened...

Thanks
John
Helmut
★★★
avatar
Homepage
Vienna, Austria,
2013-12-20 18:37
(4574 d 13:30 ago)

@ jag009
Posting: # 12091
Views: 10,100
 

 TRT|RTR or FA0(1)

Hi John,

❝ When I switched to FA0(1) then it worked. […] same as yours.


❝ Question is why FA0(2) didn't work?


Overspecified model – see this lengthy thread. I think you have two options:
  1. Avoid the partial replicate in the future. Run the fully replicated three-period TRT|RTR instead. That’s what the Lászlós recommend. Go ahead with FA0(2); no convergence issues on the way. Compliant with the guidance? Duno. You can stick to the code, but the design is not mentioned…

  2. If you already have the data on your desk, switch to FA0(1). Unlike ElMaestro I don’t think that you will be successful tweaking the side conditions in SAS (convergence criteria, no. of iterations, whatsoever). At least in PHX/WNL it never worked. Compliant with the guidance? If you believe devoutly that you have to follow the code literally – no.
    But: That’s your only chance to get a result in SAS. For your current study you could additionally show the results obtained in PHX/WNL with FA0(2) – which worked like a charm this time (by luck).

❝ I am not a saint in statistics yet so I don't know.


Me not either. But I’m pretty sure the partial replicate is crap.

Dif-tor heh smusma 🖖🏼 Довге життя Україна! [image]
Helmut Schütz
[image]

The quality of responses received is directly proportional to the quality of the question asked. 🚮
Science Quotes
jag009
★★★

NJ,
2013-12-20 22:24
(4574 d 09:43 ago)

@ Helmut
Posting: # 12092
Views: 9,901
 

 TRT|RTR or FA0(1)

Thanks Helmut,

Now I wanna see what the CRO comes back with the SAS stat output...

John
d_labes
★★★

Berlin, Germany,
2013-12-23 10:18
(4571 d 21:50 ago)

@ Helmut
Posting: # 12093
Views: 9,807
 

 Partial replicate crap? Really?

Dear Helmut!

❝ ... But I’m pretty sure the partial replicate is crap.


That's a sort of political incorrect statement about someone who can't answer back :-D.

I'm not "pretty sure" if you are right here. In my opinion it is crap that the FDA insists on using the SAS Proc Mixed code for full replicate designs also for the partial replicate design.

I always wondered why. When it comes to the scaled ABE evaluation they don't use it but revert to the evaluation of the scaled ABE criterion via intra-subject contrasts. Why not use them also for ABE if the criteria for scaling are not fulfilled? Ok, missings or imbalance may have an impact on the ABE decision if one uses this (method of moments) method. But this also applies if scaling is allowed since the ABE criterion is integral part of the scaled ABE criterion.

So why do they insist on using the SAS Proc Mixed code for the ABE decision? Sort of political correctness because it has a long history being published already in 2001 for ABE?
And do they really insist? Or are other models (beside FA0(1) we have additionally this one) also acceptable if described properly? As I see this: No one knows at the moment. May be it would help if not only John is asking them.

IMHO the partial replicate design itself is not worse or better than others, if evaluated properly and not literally.

Regards,

Detlew
Helmut
★★★
avatar
Homepage
Vienna, Austria,
2013-12-23 15:06
(4571 d 17:01 ago)

@ d_labes
Posting: # 12097
Views: 9,848
 

 Partial replicate crap? Really?

Dear Detlew!

❝ ❝ ... But I’m pretty sure the partial replicate is crap.


❝ That's a sort of political incorrect statement about someone who can't answer back :-D.


Okay, okay. Though they could. After all guys from the FDA were able to send me private e-mails asking for R-code evaluating replicate design studies. ;-)

Agree with all you other points.

Dif-tor heh smusma 🖖🏼 Довге життя Україна! [image]
Helmut Schütz
[image]

The quality of responses received is directly proportional to the quality of the question asked. 🚮
Science Quotes
ElMaestro
★★★

Denmark,
2013-12-20 15:32
(4574 d 16:35 ago)

@ jag009
Posting: # 12087
Views: 9,920
 

 Tweak & Fiddle

Hi John,

Generally when convergence issues pop up: Rescale, apply a different minimization Al Gore Rhythm, try to manually select appropriate starter values.

But isn't all this due to SAS trying to create a within-s for A in the covariance matrix, which can't be optimized? It might very well be syntax-dependent, but I am not a SAS user so I do not know which syntax to apply to make SAS understand that only a within-s for B makes sense in V.

Oh, and on a completely different note: Check AUC-extrapolations of subjs 67, 82.

Pass or fail!
ElMaestro
jag009
★★★

NJ,
2013-12-20 18:16
(4574 d 13:51 ago)

@ ElMaestro
Posting: # 12090
Views: 9,890
 

 Tweak & Fiddle

Hi ElMaestro,

❝ But isn't all this due to SAS trying to create a within-s for A in the covariance matrix, which can't be optimized? It might very well be syntax-dependent, but I am not a SAS user so I do not know which syntax to apply to make SAS understand that only a within-s for B makes sense in V.


So the SAS code is deficient? FDA should point out the need for adjustment I think. Maybe I am too much of a novice in this aspect...

❝ Oh, and on a completely different note: Check AUC-extrapolations of subjs 67, 82.


Yeah I know, the PK data was pulled from the CRO so it was just a snapshot on my behalf for now. They just use WinNonlin to estimate t1/2 and roll out the results.

Oh I encountered another sneaky thing with a CRO, on the bioanalytical side. That's another story, maybe after Xmas I will bring it up...

John
d_labes
★★★

Berlin, Germany,
2013-12-23 11:14
(4571 d 20:53 ago)

@ jag009
Posting: # 12094
Views: 10,001
 

 Not confirmed - SAS

Dear John!

Evaluation for both AUCt and AUCi converged without any change to the FDA code within 6-7 iterations (SAS 9.2).

In both cases:
NOTE: Convergence criteria met but final hessian is not positive definite.
NOTE: Asymptotic variance matrix of covariance parameter estimates has been found
      to be singular and a generalized inverse was used. Covariance parameters
      with zero variance do not contribute to degrees of freedom computed by
      DDFM=SATTERTH.


90% CIs exactly the same as Helmut's.

BTW: FA0(1) gives astonishing the same 90% CI's as FA0(2), without any (small) deviation. Even the CVwR are identical to the second decimal.

Regards,

Detlew
UA Flag
Activity
 Admin contact
23,656 posts in 4,994 threads, 1,570 registered users;
312 visitors (0 registered, 312 guests [including 29 identified bots]).
Forum time: 09:08 CEST (Europe/Vienna)

It requires a very unusual mind
to undertake the analysis of the obvious.    Alfred North Whitehead

The Bioequivalence and Bioavailability Forum is hosted by
BEBAC Ing. Helmut Schütz
HTML5