ElMaestro ★★★ Denmark, 2011-01-16 20:13 (5269 d 20:09 ago) Posting: # 6423 Views: 34,143 |
|
Ahoy bears and dlabes Wouldn't it be nice to include an option for (but not replace with) sample size calculation with adaptation to CI-widening ad modum the new BE guideline in Europe? It would be easy enough to internally adapt the theta's where the input CV is in the 30%-50% range and the design is ref-replicated. Best regards, EM. — Pass or fail! ElMaestro |
d_labes ★★★ Berlin, Germany, 2011-01-17 14:19 (5269 d 02:03 ago) @ ElMaestro Posting: # 6427 Views: 31,073 |
|
Ahoy my dear Capt'n, ❝ Wouldn't it be nice to include an option for (but not replace with) sample size calculation with adaptation to CI-widening ad modum the new BE guideline in Europe? THX for your suggestion. Theoretically the power of scaled ABE evaluation (if it could be done exact ![]() Of course the 50% CV cut-off of the EMA complicates the situation. Moreover the widened limits are only an approximation to the problem. The constraint "point estimate within 0.8 - 1.25" is not part of the classical power considerations. ❝ It would be easy enough to internally adapt the theta's where the input CV is in the 30%-50% range and the design is ref-replicated. require(PowerTOST) Easy enough for the programmer of the famous EFG ![]() Lets see what the results are: # Targetpower=0.8, theta0=0.95, alpha=0.05, design=2x3x3 partial replicate Regulatory body: EMA k=0.760
CV theta1 theta2 Sample size Achieved power Regulatory body: FDA k=0.893
CV theta1 theta2 Sample size Achieved power — Regards, Detlew |
Helmut ★★★ ![]() ![]() Vienna, Austria, 2011-01-17 15:38 (5269 d 00:44 ago) @ d_labes Posting: # 6429 Views: 31,168 |
|
Dear D. Labes! ❝ […] therefore the sample size is independent of the variability. […] For this reason the FDA has discussed 24 or32 subjects for scaled ABE. […] But don't know where these numbers exactly came from. 32 was FDA’s suggestion back in 2006 (see here). 24 is FDA’s current requirement (personal communication with Barbara Davit, Ljubljana, May 2010). ❝ ❝ # change 0.76 to log(1.25)/CV2se(0.3) if you prefer higher precision ❝ # for the regulatory constant ❝ ❝ Lets see what the results are: ❝ ❝ Regulatory body: EMA k=0.760 ❝ ❝ ❝ Yes, we discussed that already in #5249pp. As the grumpy old man I am I was already lamenting in one of my lectures (slides 32-33). Though I have only used exact widened limits of 0.75 and 0.75-1 in the past (not UL of 1.3333) I woudn’t bet on the acceptability of the exact value (which I will use anyway). As you mentioned in your commented code, one will get the ‘correct’ limits by calculating the regulatory constant based on a CV of 30%. theta1 <- exp(-log(1.25)/CV2se(0.3)*CV2se(CVr)) P.S.: Do you think that engineers would use 3.14 instead of π, because it’s written in a guideline? Bridges would collpase, trains derail, etc. ![]() — Dif-tor heh smusma 🖖🏼 Довге життя Україна! ![]() Helmut Schütz ![]() The quality of responses received is directly proportional to the quality of the question asked. 🚮 Science Quotes |
d_labes ★★★ Berlin, Germany, 2011-01-17 17:37 (5268 d 22:45 ago) @ Helmut Posting: # 6433 Views: 30,846 |
|
Dear Helmut, I correct my code as following:
... I do this because I wont take liability of collpasing bridges, derailing trains of Deutsche Bahn, etc. ![]() BTW: R > options(digits=30) I have learned in my early days during learning draftsman in agrar melioration engineering that 22/7=3.142857 is an good estimate. Very near to your 3.14! Maybe there were not so much circles involved, but more triangles and rectangles ![]() — Regards, Detlew |
Helmut ★★★ ![]() ![]() Vienna, Austria, 2011-01-17 19:06 (5268 d 21:16 ago) @ d_labes Posting: # 6434 Views: 32,800 |
|
Dear D. Labes, THX for supporting an ol’ fart. ❝ BTW: R ❝ ❝ ❝ ❝ ❝ ❝ ❝ Right. Interesting, that signif doesn’t throw an error for values > 22.options(signif=32) However pi Doesn’t really surprise me; that’s the numeric resolution of my 32 bit XP; I guess you have one of these fancy 64 bit operating systems? There’s a package for R being able to handle floating point operations in arbitrary precision – essentially by converting all numbers to character strings. ![]() ❝ I have learned in my early days during learning draftsman in agrar melioration engineering that 22/7=3.142857 is an good estimate. Very near to your 3.14! Maybe there were not so much circles involved, but more triangles and rectangles Yes, I’ve learned that one also. Another one is Archimedes’: 3+10/71 < π < 3+10/70 (the upper limit is your’s; the mean is just 0.008% off). Not so bad for 250 BCE! Maniacs might click here or there. Fascinating that calc.exe comes up with:Pi 3.1415926535897932384626433832795 Maybe it’s stored somewhere? Let’s try a true calculation: [o] Rad 1 [x] Inv tan 0.78539816339744830961566084581988 * 4 = 3.1415926535897932384626433832795 Amazing. According to the online help calculations are performed in 32 digit precision. - Pi = -1.393267707463821448075635979701e-37 Gotcha! If you have Java onboard, a true overkill for nerds (calling a Java implementation of Python ‘Jython’, library sympy for symbolic mathematics from R): require(rSymPy) Wow! Remember that one? factorial(170) But: sympy("factorial(170)") ![]() — Dif-tor heh smusma 🖖🏼 Довге життя Україна! ![]() Helmut Schütz ![]() The quality of responses received is directly proportional to the quality of the question asked. 🚮 Science Quotes |
d_labes ★★★ Berlin, Germany, 2011-01-18 09:32 (5268 d 06:50 ago) @ Helmut Posting: # 6437 Views: 31,112 |
|
Dear Helmut, Kudos to you! You can always top it ![]() BTW: Try ![]() — Regards, Detlew |
Helmut ★★★ ![]() ![]() Vienna, Austria, 2011-01-18 12:44 (5268 d 03:38 ago) @ d_labes Posting: # 6441 Views: 31,023 |
|
Dear D Labes, ❝ I didn't know the last one! Going ![]() For many years ![]() Unfortunately ![]() — Dif-tor heh smusma 🖖🏼 Довге життя Україна! ![]() Helmut Schütz ![]() The quality of responses received is directly proportional to the quality of the question asked. 🚮 Science Quotes |
ElMaestro ★★★ Denmark, 2011-01-19 18:50 (5266 d 21:32 ago) @ d_labes Posting: # 6451 Views: 30,877 |
|
Ahoy dlabes, always a pleasure reading posts from you; I understand so little... ❝ Theoretically the power of scaled ABE evaluation (if it could be done exact This thing about 'If it could be done exact ![]() If I get you right you are looking for the answer to this question: If we have a power of P at T/R=soandso and CV being thisorthat, then what should the accetance range be if the CV is not thisorthat but blahblah? In principle, the EMA guideline provides the answer to that Question as it allows a construct of a curve of acceptance range versus CV. But trying to decipher your encrypted babblings, you're saying this is an approximation given that power should be the constant. Allow me to return to the famous german desert and ingestion of clothing, but this time garnished with some heavy metal (I use the simple example of a 2,2,2-design although this is not the relevant one for scaling but it can be adapted pretty easily):
EatMyShorts <- function (x, V, T, W) Pwr222(30,0.05, 0.8, 1.25, 0.95, 0.3) is 0.6850399. Let's check: OK, what acceptance range matches a 2,2,2-design with T/R being 0.95, N=30 and power being 0.6850399? I_Eat_Cadmium (0.6850399, 0.95, 0.3, 0.05, 30) gives the answer. And if the CV is not 0.3 but 0.4 then which acc range can apply? I_Eat_Cadmium (0.6850399, 0.95, 0.4, 0.05, 30) Better approximation possible by this approach, no? EM. |
d_labes ★★★ Berlin, Germany, 2011-01-20 16:00 (5266 d 00:22 ago) @ ElMaestro Posting: # 6456 Views: 31,066 |
|
Ahoy dear "Der Meister", sorry but I must confess that I'm totally confused and that I can't follow your points and reasoning ![]() ❝ If I get you right you are looking for the answer to this question: ❝ If we have a power of P at T/R=soandso and CV being thisorthat, then what should the accetance range be if the CV is not thisorthat but blahblah? I never asked that question (they never came into my mind!) and I can't imagine what the results of I_Eat_Cadmium() are. Eventually hard poisoning with some side effects on the CNS ![]() What I meant with my sentence is that using the original scaled ABE criterion (µT-µR)2/s2WR < (theta/sw0)2 would let to power calculations that are independent from the variabilities, at least at true µT-µR=0. This is well known for using the "effect size" in superiority studies. But this can only be proven for BE within the classical 2x2 design because in case of others the distribution of the test statistics are not known exactly. See f.i. L. Tothfalusi, L. Endrenyi and A. Garcia Arieta "Evaluation of Bioequivalence for Highly Variable Drugs with Scaled Average Bioequivalence" Clin Pharmacokinet 2009; 48 (11): 725-743 The use of the widened acceptance limits according to the EMA guidance is only an approximation to the original problem (as well as the linearized criterion with approximate upper 95% CI in the FDA Progesterone guidance). Power calculations for the EMA widened acceptance limits can be done naively by inserting the widened limits depending on the assumed CV into the classical power formulas. I had thought that Your feature request is going in that direction. Lets see what happens with this approach: require(PowerTOST) Results:
GMR=1 You see (if not you need glasses ![]()
Thus my encrypted babblings are nothing more then the statement: "I'm not convinced that we are actually calculating somefink that can be called power of scABE if we do it that way". Long post but not any meaning ![]() "Power Calculation - A guess masquerading as mathematics." — Regards, Detlew |
Helmut ★★★ ![]() ![]() Vienna, Austria, 2011-01-20 16:25 (5265 d 23:57 ago) @ d_labes Posting: # 6457 Views: 30,966 |
|
Dear D. Labes! ❝ Lets see what happens with this approach: ❝ 0-8.3 ![]()
❝ Moreover the classical power formulas have nowhere the scABE constraint ❝ "point est. must be within 0.8-1.25" as presupposition. We can't incoporate a political constraint in an explicit statistical formula. The only way IMHO would be to go with Monte Carlo simulations. Power: That which is wielded by the priesthood of clinical trials, the statisticians, P.S.: Added one missing ' } ' to your code above.— Dif-tor heh smusma 🖖🏼 Довге життя Україна! ![]() Helmut Schütz ![]() The quality of responses received is directly proportional to the quality of the question asked. 🚮 Science Quotes |
d_labes ★★★ Berlin, Germany, 2011-01-20 17:15 (5265 d 23:07 ago) @ Helmut Posting: # 6460 Views: 30,795 |
|
Dear Helmut! ❝ 0-8.3 ![]() ❝ We can't incorporate a political constraint in an explicit statistical formula. The only way IMHO would be to go with Monte Carlo simulations. IMHO right. All the power calculations seen in the literature about scABE were done via simulations. But this is not so easy to implement for all the designs covered in PowerTOST. If anybody out there has an idea or code, feel free to contact me to incorporate it in the package. Edit: I'll try to shanghai Martin and Jack... [Helmut] — Regards, Detlew |
Helmut ★★★ ![]() ![]() Vienna, Austria, 2011-01-21 05:13 (5265 d 11:10 ago) @ d_labes Posting: # 6464 Views: 30,931 |
|
Dear D. Labes! ![]() — Dif-tor heh smusma 🖖🏼 Довге життя Україна! ![]() Helmut Schütz ![]() The quality of responses received is directly proportional to the quality of the question asked. 🚮 Science Quotes |
d_labes ★★★ Berlin, Germany, 2011-01-21 09:36 (5265 d 06:47 ago) (edited on 2011-01-21 10:48) @ Helmut Posting: # 6467 Views: 30,614 |
|
Dear Helmut! ❝ … no comment Helmut: Forget my post or delete it. I have totally misinterpreted the graphics ![]() Learned: First think, then post. BTW: Looking at the time of your post: did You suffer from senile Bettflucht (insomnia, Google translator "Bed run") ![]() — Regards, Detlew |
Helmut ★★★ ![]() ![]() Vienna, Austria, 2011-01-21 13:53 (5265 d 02:29 ago) @ d_labes Posting: # 6469 Views: 30,881 |
|
Dear D. Labes! ❝ BTW: Looking at the time of your post: did You suffer from senile Bettflucht (insomnia, Google translator "Bed run") Not quite. I rarely go to bed earlier than 03:00 am – so this was just an outlier. ![]() — Dif-tor heh smusma 🖖🏼 Довге життя Україна! ![]() Helmut Schütz ![]() The quality of responses received is directly proportional to the quality of the question asked. 🚮 Science Quotes |
d_labes ★★★ Berlin, Germany, 2011-01-21 13:12 (5265 d 03:10 ago) @ Helmut Posting: # 6468 Views: 31,188 |
|
Dear Helmut! Look at this! ![]() I have combined the the graphics from the paper L. Tothfalusi, L. Endrenyi Limits for the scaled average bioequivalence of highly variable drugs and drug products Pharm. Res. 20, p382-389 (2003) Figure 3 "Percentage of simulated two-period crossover studies in which BE was accepted ..." Wow! — Regards, Detlew |
Helmut ★★★ ![]() ![]() Vienna, Austria, 2011-01-21 14:01 (5265 d 02:22 ago) @ d_labes Posting: # 6471 Views: 30,656 |
|
Dear D. Labes! ❝ Look at this! ❝ Wow! ![]() — Dif-tor heh smusma 🖖🏼 Довге життя Україна! ![]() Helmut Schütz ![]() The quality of responses received is directly proportional to the quality of the question asked. 🚮 Science Quotes |
d_labes ★★★ Berlin, Germany, 2011-01-21 15:43 (5265 d 00:39 ago) @ Helmut Posting: # 6474 Views: 30,791 |
|
Dear Helmut, beyond belief for me simple minded but seems definitely TRUE. Have a look at Figs. 3 and 4 of Haidar et. al. Evaluation of a Scaling Approach for the Bioequivalence of Highly Variable Drugs The AAPS Journal, Vol. 10, No. 3, p450-454, 2008 online resource to see the same behaviour, also for the combined criteria scaled ABE + constraint on point estimate. See also Endrenyi, Tothfalusi Presentation at BASS 2010 Scaled Average bioequivalence: An Approach to Resolve a Difficult Program page 21-23. See also Page 28 of that presentation to notice increasing power after CV=30% if point estimator is 1.1, same as we have had here for point estimator = 0.95. It would be interesting to know how good the real values of the simulated power are in accordance with our naive power calculations, not only the qualitative behaviour. Seems I had read all this stuff, but never had this figured out well. What we here see is the mean vs. variance trade-off in the scaled ABE evaluation. Greater differences in the means µT-µR are allowed with increasing variability to have the same value for the scaled ABE criterion. Eventually this is one of the sources for introducing the GMR constraint? — Regards, Detlew |
ElMaestro ★★★ Denmark, 2011-01-21 13:59 (5265 d 02:23 ago) @ Helmut Posting: # 6470 Views: 30,632 |
|
Ahoy HS, ❝ ❝ Moreover the classical power formulas have nowhere the scABE constraint "point est. must be within 0.8-1.25" as presupposition. ❝ ❝ We can't incoporate a political constraint in an explicit statistical formula. The only way IMHO would be to go with Monte Carlo simulations. I can't imagine this is really a big deal, because the chance that a PE is outside 0.8 - 1.25 while the CI still evaluates to BE must be next to nothing. — Pass or fail! ElMaestro |
Helmut ★★★ ![]() ![]() Vienna, Austria, 2011-01-21 15:05 (5265 d 01:18 ago) @ ElMaestro Posting: # 6473 Views: 30,698 |
|
Ahoy ElMeastro! ❝ ❝ We can't incoporate a political constraint in an explicit statistical formula. The only way IMHO would be to go with Monte Carlo simulations. ❝ ❝ I can't imagine this is really a big deal, because the chance that a PE is outside 0.8 - 1.25 while the CI still evaluates to BE must be next to nothing. To quote from L. Tóthfalusi et al. (2009), p.739: It is interesting to note the effect of the GMR constraint on the overall performance, notably on the producer risk of bioequivalence determinations. This effect depends on the comparative positions of the three criteria: the GMR constraint, the regulatory limit for SABE, and the estimated scaling variance. Depending on the relative magnitudes of these parameters, either SABE or the GMR restriction can become the dominant regulatory criterion. For instance, with a constraint of 0.80–1.25 on the GMR, when both the switching and scaling variations are 30%, then the combined criteria of SABE and the GMR restriction have, at moderately high within-subject variability (CV = 35%), almost the same proportion of regulatory approvals as SABE alone. In contrast, at substantially higher within-subject variability (CV = 60%), the proportion of approvals with the combined criteria is very similar to that with just the GMR constraint alone. Have a look at Fig.2B of Endrényi L and L Tóthfalusi ![]() The GMR constraint was 1.25, filled diamonds are the EMA's method (filled circles conventional ABE, open squares unconstrained SABE, open triangles PE 0.8-125 without CI = Canada Cmax). Simulation: 10,000 studies, 3-period replicate, true GMR 1.0-1.6. Due to the variability inherent to HVDs almost 40% pass at 1.25 and still ≈20% at 1.35. — Dif-tor heh smusma 🖖🏼 Довге життя Україна! ![]() Helmut Schütz ![]() The quality of responses received is directly proportional to the quality of the question asked. 🚮 Science Quotes |
ElMaestro ★★★ Denmark, 2011-01-21 15:52 (5265 d 00:30 ago) @ Helmut Posting: # 6475 Views: 30,796 |
|
Thanks HS. My intuition -which is most often counteracting reasonable decisions- tells me something needs to be dug into here. E&T's paper says they do simulations and evaluate the simulated data with a simplified method by Hyslop. The reference (Hyslop et al. 2000) which I don't have is about IBE rather than ABE. Do you have it? Can you shed light onto this? Who's got some working R-code for 223-ABE evaluation (linear mixed model that specifies the covariance matrix with within-sigmas for the ref when the test is not replicated; I don't know how to do this)? — Pass or fail! ElMaestro |
d_labes ★★★ Berlin, Germany, 2011-01-21 16:28 (5264 d 23:55 ago) @ ElMaestro Posting: # 6476 Views: 30,725 |
|
Dear ElMaestro, ❝ E&T's paper says they do simulations and evaluate the simulated data with a simplified method by Hyslop. The reference (Hyslop et al. 2000) which I don't have is about IBE rather than ABE. Do you have it? Can you shed light onto this? See here and especially here for some discussions about how to calculate the scaled ABE criterion (linearized) including it's upper 95% CI according to Hyslop's method, sometimes the name Howe is also used. The SAS code given in the FDA Progesterone guidance can be rewritten for R with not so much difficulties I think. But a full working version I don't have. — Regards, Detlew |
ElMaestro ★★★ Denmark, 2011-01-21 16:45 (5264 d 23:37 ago) @ d_labes Posting: # 6477 Views: 30,624 |
|
Thanks dlabes, I am not sure I follow the code. As I indicated earlier, I'd be a little apprehensive at least for the case of imbalance between sequences; HOWEver ( ![]() Correction: Sorry, I meant something else than I initially wrote. It is missing values that is the concern issue because of the difference between a normal linear model and the reml approach by a mixed model. — Pass or fail! ElMaestro |
d_labes ★★★ Berlin, Germany, 2011-01-21 17:19 (5264 d 23:04 ago) @ ElMaestro Posting: # 6478 Views: 31,012 |
|
Hi ElMaestro ❝ Correction: Sorry, I meant something else than I initially wrote. It is missing values that is the concern issue because of the difference between a normal linear model and the reml approach by a mixed model. Good question! Next question please ![]() To be honest: Don't know exactly. In the Progesterone guidance it is written "... Further assume that there are no missing observations ...". But only under the Heading "Example SAS Codes: fully replicated 4-way design". My observation is that this guidance is in some instances rather vague or gives no sufficient explanation why what is done that way. Just to name "the unknown x = estimate2 - stderr2" or "use GLM for partial replicate and MIXED for full replicate". — Regards, Detlew |
Helmut ★★★ ![]() ![]() Vienna, Austria, 2011-01-21 19:41 (5264 d 20:41 ago) @ ElMaestro Posting: # 6479 Views: 30,740 |
|
Ahoy ElMaestro! ❝ My intuition -which is most often counteracting reasonable decisions- tells me something needs to be dug into here. I try not to think with my gut. If I’m serious about understanding the world, thinking with anything besides my brain, as tempting as that might be, is likely to get me into trouble. To be honest I'm struggling in preventing my little brain from overheating. My rumbling gut is uncomfortable with the 'fact' (?) that with any given sample size I have a higher chance of success for higher CVs. ❝ The reference (Hyslop et al. 2000) which I don't have is about IBE rather than ABE. Do you have it?
T Hyslop and B Iglewicz — Dif-tor heh smusma 🖖🏼 Довге життя Україна! ![]() Helmut Schütz ![]() The quality of responses received is directly proportional to the quality of the question asked. 🚮 Science Quotes |
ElMaestro ★★★ Denmark, 2011-01-20 16:34 (5265 d 23:48 ago) @ d_labes Posting: # 6458 Views: 31,573 |
|
Dear dlabes, Thanks for your continued support to my lack of understanding ![]() I cannot rule out the possibility that my metal poisoning is a factor in my posts. ❝ What I meant with my sentence is that using the original scaled ABE criterion ❝ ❝ would let to power calculations that are independent from the variabilities, at least at true µT-µR=0. At T/R equal to 1 power and variability is independent. At T/R not equal to 1 it becomes unclear. As the intention with scaling is to keep power constant with varying CV one should/can not always assume T/R being one, I mean. My code would be attempt at calculating the limits that preserve constant power with varied CV where T/R is anything the user specifies. With this function, curves for acceptance range as function of CV can be constructed for any combo of T/R and N while the power is kept constant. — Pass or fail! ElMaestro |
d_labes ★★★ Berlin, Germany, 2011-01-20 17:06 (5265 d 23:16 ago) @ ElMaestro Posting: # 6459 Views: 30,755 |
|
Dear ElMaestro, ❝ With this function, curves for acceptance range as function of CV can be constructed for any combo of T/R and N while the power is kept constant. Oh I see! But for what end? I understand the EMA widened acceptance ranges as applicable regardless of point estimate obtained. — Regards, Detlew |
ElMaestro ★★★ Denmark, 2011-01-20 17:52 (5265 d 22:31 ago) @ d_labes Posting: # 6461 Views: 30,735 |
|
Dear dlabes, ❝ Oh I see! But for what end? ❝ I understand the EMA widened acceptance ranges as applicable regardless of point estimate obtained. A good point to make. On the other hand some people find pleasure in discussing if 0.760 is meaningful or if it should be 17 decimals. In this case I think it is a matter of doing what the guidelines says or doing what the guideline intends. I would expects this to be something you can discuss. The code above does not assume anything other than what the ordinary (and widely accepted power routines) do. I envisage Helmut examining the acceptance range for T/R deviating from 1 and comparing it to the EMA curve and presenting such stuff at a conference. Heck if he doesn't do it, I will. Here is a divide-and-conquer approach to the problem (simpler but not really faster ![]() I_Eat_Cadmium2 <- function(P, TR, CV, alpha, N) — Pass or fail! ElMaestro |
ElMaestro ★★★ Denmark, 2011-01-20 18:53 (5265 d 21:29 ago) @ d_labes Posting: # 6462 Views: 31,140 |
|
OK, here what I meant with my last posts and the gibberish code. ![]() We have for instance 24 patients in a 2,2,3-BE trial, expected T/R being 1.0 and expected CV is 30%. This gives us a power of 0.87. Scaling preserves power. Thus should the variability be 50% and with preserved power we have 0.6983 as our lower acceptance range and 1.4320 as the upper. Wunderbar. This is the blue line, or shall we call it the EMA line. I propose to check how the behavior is once the expected R/T diviates from 1.0 and being 0.95 (red), 0.90 (yellow) and 0.85 (green); these curves preserve the power within themselves but of course not between them. It was intended as a graphical illustration of the limitation of the EMA approximation when T/R differs from 1.0. Note: We are now no longer so intimately connected with the original thread topic. — Pass or fail! ElMaestro |