Imran
☆    

Mumbai,
2007-06-27 08:02
(6927 d 06:44 ago)

Posting: # 835
Views: 31,376
 

 Steady state calculation using WinNonlin [Software]

Dear PK Colleagues

I have come across a problem while runing PK of steady state using WinNonline software. My queries are:
  1. should trough concentrations and concentration of profiling day should be run separately or in the same set of data to generate PK parameters?
  2. Should only trough values should be used for calculation of % fluctuation?
Please suggest, if anyone having the insight in running the PK of Steady state using WinNonlin.

Thanking you in Anticipation

Regards
Dr.Imran Khan

Dr.Imran Khan
Helmut
★★★
avatar
Homepage
Vienna, Austria,
2007-06-27 13:56
(6927 d 00:50 ago)

@ Imran
Posting: # 840
Views: 23,946
 

 Steady state calculation using WinNonlin

Dear Imran!

❝ 1. should trough concentrations and concentration of profiling day should

❝ be run separately or in the same set of data to generate PK parameters?

  • If you are talking about PK modeling, all values should be included in the data set. But beware: in most (all?) situations WinNonlin's stripping procedure does not provide intial estimates; they have to be provided by the user.
  • If you are talking about NCA, beware of 'PK/PD/NCA Modeling Questions FAQ #2' form Pharsight's support site:
    Q: When WinNonlin gives an AUC following multiple doses, is this the total AUC or the AUC for the first dose?
    A: When using Noncompartmental Modeling, the AUC given in the Final Parameters table is the AUC from 0 to infinity that would occur if only the first dose had been given.
Also beware of (quoting the manual):

Time deviations in steady-state data
When using steady-state data, WinNonlin computes AUC from dose time to dose time + tau, based on the tau value set above. However, in most studies, there will be sampling time deviations. That is, if dose time = 0 and tau = 24, the last sample might be at 23.975 or 24.083 hours. In this instance, the pro­gram will estimate the AUC based on the estimated concentration at 24 hours, and not the concentration at the actual observation time. For steady state data, Cmax, Tmax, Cmin and Tmin are found using observations taken at or after the dose time, but no later than dose time+tau.


❝ 2. Should only trough values should be used for calculation of % fluctuation?


:confused: %fluctutation (or %Peak-to-Trough Fluctuation, %PTF) is defined as:
\(\small{\%PTF=100\frac{C_{\textrm{max}}-C_{\textrm{min}}}{C_{\textrm{av}}}}\)
where \(\small{C_{\textrm{av}}=AUC_{\tau}/\tau}\), and \(\small{C_{\textrm{max}}}\), \(\small{C_{\textrm{min}}}\) refer to extremes of the profile in steady state.

In WinNonlin you have to calculate %PTF manually (it's not given in the result workbook).

Dif-tor heh smusma 🖖🏼 Довге життя Україна! [image]
Helmut Schütz
[image]

The quality of responses received is directly proportional to the quality of the question asked. 🚮
Science Quotes
Imran
☆    

Mumbai,
2007-06-29 07:13
(6925 d 07:33 ago)

(edited on 2007-06-29 10:01)
@ Helmut
Posting: # 844
Views: 23,830
 

 Steady state calculation using WinNonlin

Dear HS
Thanks for ur mail and valuable suggestion on the issue. I data I want to estimate for SS study is given below.
Subject Period Trough sample      Profiling till 72.00 hrs
                  Day 1   Day-5   Day-6   Day-7  Day-7
                  0.0 hr  0.0 hr  0.0 hr  0.0 hr  1.0 hr

  2        2      BLQ    13.344  13.689  12.272   8.427
  3        1      BLQ     2.408   2.88    3.299   2.944
  4        1      BLQ     6.068   7.562   5.267   4.734
  5        1      BLQ     4.711   6.01    6.407   5.082
  6        2      BLQ     5.594   5.134   4.733   3.691
  7        2      BLQ     4.043   5.055   3.842   3.867
  8        1      BLQ     6.104   6.033   6.144   5.251
  9        2      BLQ     7.272   8.825  10.937   6.892
 10        2      BLQ     2.671   3.53    3.655   2.744
 11        1      BLQ     2.238   6.671   3.635   3.092
 12        2      BLQ     8.961   4.591   5.078   4.127
 13        1      BLQ     9.957  11.607   8.85    7.571
 14        2      BLQ     2.596   3.972   2.756   2.696
 15        1      BLQ     5.578   5.959   4.593   3.244
 16        2      BLQ     4.135   5.602   6.656   5.397
 17        1      BLQ     3.339   3.985   2.957   3.241
 18        2      BLQ     7.984   6.268   4.776   3.843
 19        1      BLQ     5.478   5.781   5.357   4.043
 20        2      BLQ    17.672  21.912  24.53   22.566


The step I followed calculation of PK Parameters is as:
PK/PD/NCA Analysis Wizard , noncompartment modeling, select model 200, data variables, dosage regimen and selecting steady state followed by modeling, generation of PK output.

The XL file I have attached, the data processed is excluding trough sample (day1, 5, 6 and 7), should I process trough data and profiling data separately.

As u wrote in ur mail that in WinNonlin %PTF to be calculate manually, but here in my case I calculated all of the primary and secondary parameters (Excel file attached).

I have troubleshooting how the data to be presented in the WinNonline, I tried in all the possible way, but I am not sure about the output.

Primary Efficacy Parameters: Cmin, Cmax, AUC0 - 

Secondary Efficacy Parameters: Tmax, % Fluctuation

Please suggest.

Regards
Dr.Imran Khan

Dr.Imran Khan
Jaime_R
★★  

Barcelona,
2007-06-29 12:29
(6925 d 02:17 ago)

@ Imran
Posting: # 845
Views: 23,555
 

 Please use the Preview!

Dear Imran,

I edited your post (took me half an hour), because it came out as a mess.
Please use the 'Preview'-button before posting (and read the section about formating in the Forum's Policy).
Don't use <tabs>, use [ code]with blanks between columns[/code] instead.

❝ The XL file I have attached [...]



What's 'XL'? Do you mean Micro$oft's Excel?
Although the Forum's software supports Uploads, this function is deactivated.
Attachments are not possible anyhow (this is not an e-mail list).

❝ [...] (Excel file attached).


OK, I was right, but just out of couriosity: what did you actually try, to 'attach a file'?

❝ I have troubleshooting how the data to be presented in the WinNonline, I tried in all the possible way, but I am not sure about the output.


Why don't you contact Pharsight's Support?

❝ Primary Efficacy Parameters: [...] AUC0 - &#61556;


'#61556' is in the private code range of UTF-8. No greek letters, Unicode, or ISO-8859 above DEC 255 are supported. You would have noticed this during Preview. Maybe you mean the greek letter 'tau' denoting a dosage interval (Unicode #964).

Don't take it personally, but I think the Forum should not become a Fastfood Restaurant (quickanddirty).

Regards, Jaime
Imran
☆    

Mumbai,
2007-06-30 14:50
(6923 d 23:56 ago)

@ Jaime_R
Posting: # 848
Views: 23,547
 

 Please use the Preview!

Dear Jaime

Extremely sorry! I tried to send the data in best possible way, even checked the preview, but I could not trace attachement option in the mail

I have definately wrote a mail to Pharsight, but as a matter of discussion I wrote to Forum.

Even I am agree that the Forum should not become a Fastfood Restaurant

Best Regards,
Imran

--
Edit: Full quote removed. [Jaime]
Jaime_R
★★  

Barcelona,
2007-06-30 15:33
(6923 d 23:13 ago)

@ Imran
Posting: # 849
Views: 23,450
 

 no attachments and uploads...

Dear Imran!

Extremely sorry!


OK, my reaction was also a little bit too harsh...

❝ [...] but I could not trace attachement option in the mail


Yes, because this is not a Mailing List, but a Forum.
File uploads may be possible in the new version.

❝ I have definately wrote a mail to Pharsight,[...]



Expect an 'RTFM'... :rotfl:

❝ [...] but as a matter of discussion I wrote to Forum.


Yes, that's fine!

Regards, Jaime
Helmut
★★★
avatar
Homepage
Vienna, Austria,
2007-06-29 23:08
(6924 d 15:38 ago)

@ Imran
Posting: # 847
Views: 23,999
 

 Steady state calculation using WinNonlin

Dear Imran!

❝ The step I followed calculation of PK Parameters is as: PK/PD/NCA Analysis Wizard , noncompartment modeling, select model 200, data variables, dosage regimen and selecting steady state followed by modeling, generation of PK output.


Yes fine, but the order in Dosage Regimen is:
[x] Steady state
Dose
Time of Last Dose
Tau

❝ The XL file I have attached, the data processed is excluding trough sample (day1, 5, 6 and 7), [...]


Sure, you are interested in the profile (WinNonlin simply ignores the saturation phase, which is fine).

❝ should I process trough data and profiling data separately.


In versions 5.1+ you may run a linear regression of e.g., the last three trough values to check whether you have reached steady state (95% CI of the slope should include zero).
Model 502 (Linear) of example's data below give:
Last 4 troughs (24 h - 96 h): slope 0.063, CI -0.049, +0.176
Last 3 troughs (48 h - 96 h): slope 0.019, CI -0.120, +0.158
...so you may assume steady state conditions.

❝ As u wrote in ur mail that in WinNonlin %PTF to be calculate manually, but here in my case I calculated all of the primary and secondary parameters [...]



Oh, you are right!

❝ I have troubleshooting how the data to be presented in the WinNonline, I tried in all the possible way, but I am not sure about the output.


Why don't you keep it simple and check the results manually?
OK, I've done that for you... ;-)

D=100, V/F=1, k01=0.6931, k10=0.06931, tau=24h for five days gives (rounded to 1 decimal place):
time conc
 0    0.0
 24  21.1
 48  25.0
 72  25.8
 96  25.9
 98  91.5
 99  97.4
100  96.9
102  88.7
104  78.3
106  68.4
109  55.7
112  45.2
116  34.3
120  26.0


Now let's have a look at the output (Non-Transposed Final Parameters Tab), and a 'manual' calculation
          WinNonlin5.2  'manual'
AUClast       1425.4     1425.4
Cmax            97.4      97.4
Cmin            25.9      25.9
Tmin            96        96
Cav             59.4      59.4
%Fluctuation   120.3873  120.3873
AUCtau        1425.4    1425.4
AUCall        1425.4    1425.4


Fine.
Please note, that AUClast = AUCtau = AUCall.
Generally nobody is interested in the sampling beyond the last administration + tau, but in your case obviously this was the case (+72 hours).

If we add two more samples (144 h, 168 h), WinNonlin's output might get a little bit confusing for a novice, since:
time conc
  0   0.0
...   ...
120  26.0
144   4.9
168   0.9


          WinNonlin5.2  'manual'
AUClast       1865.8    1865.8
Cmax            97.4      97.4
Cmin            25.9      25.9
Tmin            96        96
Cav             59.4      59.4
%Fluctuation   120.3873  120.3873
AUCtau        1425.4    1425.4
AUCall        1865.8    1865.8


WinNonlin is 'clever' enough to keep AUCtau (and therefore Cav, and %Fluctuation correct within tau), but please note AUClast # AUCtau and AUCall!

If we sample for another 48 hours until we get a BLQ:
time conc
  0   0.0
...   ...
168   0.9
192   0.2
216   BLQ


          WinNonlin5.2  'manual'
AUClast       1879.0    1879.0
...              ...       ..
AUCtau        1425.4    1425.4
AUCall        1881.4    1881.4


Now finally we ran into AUClast # AUCtau # AUCall!
AUCall adds a triangle from the Tlast to the next sampling time point, assuming C=0. If you know of any serious reference using AUCall, please tell me. It's Pharsight's invention and IMHO useless - I removed it from my templates... :angry:
It's also funny that Pharsight's business rival ThermoScientific added this 'metric' to Kinetica as well.

In short:
  • Sample only until last administration + tau,
  • WinNonlin gives the correct result for %Fluctuation,
  • Don't get confused; it's only AUCtau you are interested in.

❝ Primary Efficacy Parameters: Cmin, Cmax, AUC0 - &#61556;

❝ Secondary Efficacy Parameters: Tmax, % Fluctuation


In BA/BE we should not talk about Efficacy...

Dif-tor heh smusma 🖖🏼 Довге життя Україна! [image]
Helmut Schütz
[image]

The quality of responses received is directly proportional to the quality of the question asked. 🚮
Science Quotes
Imran
☆    

Mumbai,
2007-08-02 15:52
(6890 d 22:54 ago)

@ Helmut
Posting: # 957
Views: 23,417
 

 Steady state calculation using WinNonlin

Dear HS

Thanks for the solutions. I worked with WinNonlin and I am through with my results.

One more Issue I would like to discuss with u.
How can we calculate or prove that SS is achieved.
I had used SAS for the same by repeated measures of ANOVA for Cpd calculation, and withdrawn the conclusion based on time* formulation interaction (p value.)

What procedure to be followed by using Winnonlin 5.0.1. or any other suitable method can be used to prove SS is achived.


Pls comment.

Best Regards
Dr.Imran

--
Edit: Full quote removed. Please see this post! [HS]

Dr.Imran Khan
Helmut
★★★
avatar
Homepage
Vienna, Austria,
2007-08-02 17:07
(6890 d 21:40 ago)

@ Imran
Posting: # 958
Views: 23,549
 

 Steady state calculation using WinNonlin

Dear Imran!

❝ How can we calculate or prove that SS is achieved.

❝ I had used SAS for the same by repeated measures of ANOVA for Cpd calculation, and withdrawn the conclusion based on time* formulation interaction (p value.)


In never understood this procedure. OK, statistically yes, but would you really throw away an entire study based on such a result? I always preferred it to look at the time course of the subjects and eventually exclude only ones who were not in steady state on the profile day. Of course you have to state this procedure in the protocol.

❝ What procedure to be followed by using Winnonlin 5.0.1. or any other suitable method can be used to prove SS is achived.


For my personal suggestions see this tread.

If you want to perform the analogue of SAS' repeated measures ANOVA, you will have to fight yourself through WinNonlin's Linear Mixed Effects Modeling feature - and if you succeed, come back and tell us how you did it! :cool:

Dif-tor heh smusma 🖖🏼 Довге життя Україна! [image]
Helmut Schütz
[image]

The quality of responses received is directly proportional to the quality of the question asked. 🚮
Science Quotes
Imran
☆    

Mumbai,
2007-08-06 13:48
(6887 d 00:58 ago)

@ Helmut
Posting: # 965
Views: 23,536
 

 Steady state calculation using WinNonlin

Dear HS and Colleagues
Hi!

In SS study Cpd Analysis should we calculate SS acheived for each subject, or it should be calulated as a mean of all observation.

What if a subject 'A' has not acheived SS, sholud the subject be included in the PK analysis.

I understand we calculate ABE, but would be it questioned if any subject is not in SS state.


Please comment on the issue.
Best regards

Dr.Imran

--
Edit: Full quote removed. Please see this post! [Jaime]

Dr.Imran Khan
Helmut
★★★
avatar
Homepage
Vienna, Austria,
2007-08-07 00:16
(6886 d 14:30 ago)

@ Imran
Posting: # 966
Views: 23,672
 

 Exclusion of subjects who are not in SS

Dear Imran!

❝ In SS study Cpd Analysis should we calculate SS acheived for each subject, or it should be calulated as a mean of all observation.


As already told in this thread, my personal preferences are always to assess achievement of SS for individual subjects.

❝ What if a subject 'A' has not acheived SS, sholud the subject be included in the PK analysis.


IMHO this subject should be excluded, because if the subject was not in steady state, the superposition principle of pharmacokinetics is not fulfilled.
You may only make valid statements (about accumulation, etc.) if
AUCinf (single dose) = AUCtau (steady state).

❝ I understand we calculate ABE, but would be it questioned if any subject is not in SS state.


At least I would do so! :-D

Dif-tor heh smusma 🖖🏼 Довге життя Україна! [image]
Helmut Schütz
[image]

The quality of responses received is directly proportional to the quality of the question asked. 🚮
Science Quotes
KR
★    

India,
2009-04-13 16:34
(6270 d 22:12 ago)

(edited on 2009-04-13 22:30)
@ Helmut
Posting: # 3516
Views: 22,330
 

 Exclusion of subjects who are not in SS

Dear Helmut,

Adding to that, if dosing interval for reference product is 24 hours and that of test product is 12 hours i.e. tau=24 hrs. for reference and tau=12 hrs. for test product then how can we calculate AUCtau ? means if we put tau=12 hrs. for test product in the model then which AUC area will be estimated by WinNonlin i.e first 0-12 hrs or last 0-12 hrs ?

Moreover, if we calculate AUC0-24 for both test and reference in above case then how WinNonlin calculates ? Should i give common dosing interval i.e. tau=24 hrs. for both test and reference products and how much AUC0-24 calculation is reliable for test product?

Thanks and regards,

KR
--
Edit: Full quote removed. Please see this post! [Ohlbe]
vixen
☆    

Czechia,
2020-04-27 12:10
(2239 d 02:36 ago)

@ Helmut
Posting: # 21350
Views: 11,976
 

 Exclusion of subjects who are not in SS

Hello,

I like Helmut's suggestion to assess achievement of SS for individual subjects. I wonder though whether I should evaluate the Cpds on a normal or logarithmic scale. What is your approach?

Thanks,
vixen
Helmut
★★★
avatar
Homepage
Vienna, Austria,
2020-04-27 13:45
(2239 d 01:02 ago)

@ vixen
Posting: # 21351
Views: 11,966
 

 Common sense instead of statistics

Hi vixen,

❝ I like Helmut's suggestion to assess achievement of SS for individual subjects.


My suggestion dates back to 2007. A while ago…
I don’t use it since 2010 and don’t recommend it any more (see this presentation, slide 14). Furthermore, the EMA stated:*

A specific statistical method to assure that steady state has been reached is not considered necessary in bioequivalence studies. Descriptive data is sufficient.


❝ I wonder though whether I should evaluate the Cpds on a normal or logarithmic scale.


Normal – otherwise you cannot deal with BQLs and missings.

❝ What is your approach?


See the presentation.


  • EMEA. Overview of Comments received on Draft Guideline on the Investigation of Bioequivalence. EMA/CHMP/EWP/26817/2010. London. 20 January 2010. Online.

Dif-tor heh smusma 🖖🏼 Довге життя Україна! [image]
Helmut Schütz
[image]

The quality of responses received is directly proportional to the quality of the question asked. 🚮
Science Quotes
UA Flag
Activity
 Admin contact
23,653 posts in 4,991 threads, 1,570 registered users;
134 visitors (0 registered, 134 guests [including 25 identified bots]).
Forum time: 14:47 CEST (Europe/Vienna)

I have never in my life learned anything
from any man who agreed with me.    Dudley Field Malone

The Bioequivalence and Bioavailability Forum is hosted by
BEBAC Ing. Helmut Schütz
HTML5