Shuanghe
★★  

Spain,
2025-11-10 16:55
(214 d 17:13 ago)

Posting: # 24484
Views: 5,185
 

 new regression rule for lambda_z in Phoenix: bug or new feature? [Software]

Dear all,

A quick question regarding the best fit regression method to determine \(\lambda_z\) in Phoenix.

I vaguely recall that in older WinNonlin and early version of Phoenix, the rule to determine the \(\lambda_z\) is more or less like this: starting from the last 3 points, do the regression and calculate \(R^2_\textrm{adj}\), then include last 4 points and repeat the regression and calculate \(R^2_\textrm{adj}\), if \(R^2_\textrm{adj}\) improves (more than some tolerance limit such as 0.0001), continue move forward by including 1 more early point, ..., until your \(R^2_\textrm{adj}\) does not improve or improves but within tolerance limit, then you stop. Even when there are more time points after \(t_\textrm{mx}\), we do not proceed forward. Those are what I remembered and there were examples of some R code in this forum to implement such approaches.

However, I noticed that in the recent Phoenix Certara changed the method to the following: starting from the last 3 points, do the regression and calculate \(R^2_\textrm{adj}\), and repeat it until you reach the point after \(t_\textrm{mx}\) (for oral dose). Then you compare all the \(R^2_\textrm{adj}\), the biggest \(R^2_\textrm{adj}\) is the winner. If there are more than 1 \(R^2_\textrm{adj}\) within tolerance limit (e.g., 0.0001), then the winner is the one including the most time points.

I paraphrased the new procedure and left some details that's irrelevant to my question below. You can read it in details here.

When I noticed the difference of PK values that I got from my R code and the results received from CRO using Phoenix, I thought I might have found a bug in Phoenix. But then I saw the manual of the new version of Phoenix, it seems they did it deliberately.

I think that the previous method (even though with its limitations) is better. To cite Helmut's word, the previous best fit method is "greedy" (compared to the eyes of pharmacokineticist). Now I don't have a word for the new method. "Greeeeeeedy"? Does anyone know why they changed it? Simon, are you still here? Is this a bug or new feature? :-D


Edit: Category changed. [Helmut]

All the best,
Shuanghe
Helmut
★★★
avatar
Homepage
Vienna, Austria,
2025-11-11 09:42
(214 d 00:25 ago)

@ Shuanghe
Posting: # 24486
Views: 4,540
 

 new regression rule for lambda_z in Phoenix: bug or new feature?

Hi Shuanghe,

very strange. If you navigate up in the online-manual, can you access the ‘release notes’?

[image]


I can’t, though I have a license. I will drop Simon a note.

Dif-tor heh smusma 🖖🏼 Довге життя Україна! [image]
Helmut Schütz
[image]

The quality of responses received is directly proportional to the quality of the question asked. 🚮
Science Quotes
SDavis
★★  
Homepage
UK,
2025-11-11 15:24
(213 d 18:43 ago)

(edited on 2025-11-11 15:36)
@ Shuanghe
Posting: # 24488
Views: 4,544
 

 new regression rule for lambda_z in Phoenix: no change in implementation

hi, I happen to have a 4.1 user manual on my bookshelf, yes a real book (2003)... and you can see from the attached pages the text is almost the same as in 8.7. My understanding is the only change made was with v6.0 of WinNonlin, first Phoenix version, was not to allow Cmax.

I think the issue maybe the text is slightly ambiguous and people read into it, "and then you stop", however it always went all the way back to Tmax, (and from 6.0+; Tmax+1) before selecting the value with the largest Adj-R-Sq. I think "If the adjusted R2 does not improve, but is within 0.0001 of the largest adjusted R2 value, the regression with the larger number of points is used." infers that but I've asked that we make it clearer in future iterations of the help that all points are assessed.

Long story short, it hasn't changed for 20+years,
Regards, Simon. Manager, Support, Certara.
[image]
[image][image]

Simon
Senior Scientific Trainer, Certara™
[link=https://www.youtube.com/watch?v=xX-yCO5Rzag[/link]
https://www.certarauniversity.com/dashboard
https://support.certara.com/forums/
Shuanghe
★★  

Spain,
2025-11-11 23:38
(213 d 10:29 ago)

@ SDavis
Posting: # 24489
Views: 4,477
 

 new regression rule for lambda_z in Phoenix: no change in implementation

Hi Simon,

❝ hi, I happen to have a 4.1 user manual on my bookshelf, yes a real book (2003)... and you can see from the attached pages the text is almost the same as in 8.7.


Many thanks for the clarification. I check the Phoenix manual from 8.7 back to 8.2.2 (the earliest available manual that I could find online) and the description is more or less the same. I was about to ask Helmut if he has older WinNonlin version 5 manual to double check it, yours is better. I'm surprised, but glad, that you still has that old manual.

❝ Long story short, it hasn't changed for 20+years,


I stand corrected. I must've mixed up the source of the regression method that I described above before the Phoenix's approach.

Did any forum member recall the 1st approach of regression for \(\lambda_z\) I described in my post above? I could swear that I've read it somewhere.

All the best,
Shuanghe
Helmut
★★★
avatar
Homepage
Vienna, Austria,
2025-11-12 00:20
(213 d 09:47 ago)

@ Shuanghe
Posting: # 24490
Views: 4,510
 

 Old User’s Guides

Hi Shuanghe,

❝ […] I was about to ask Helmut if he has older WinNonlin version 5 manual to double check it, yours is better. I'm surprised, but glad, that you still has that old manual.


From my archive:

WinNonlin 3.3 (2001)
[image]

WinNonlin 5.0.1 (2005)
[image]

WinNonlin 5.3 (2009)
[image]

Phoenix WinNonlin 6.3 (2012)
[image]
Exactly the same wording in Phoenix WinNonlin 6.4 (2014), 7.0 (2016), 8.0 (2017), 8.1.1 (2018), 8.3.4 (2020), 8.4.0 (2023), 8.5.1 (2024).

Dif-tor heh smusma 🖖🏼 Довге життя Україна! [image]
Helmut Schütz
[image]

The quality of responses received is directly proportional to the quality of the question asked. 🚮
Science Quotes
Helmut
★★★
avatar
Homepage
Vienna, Austria,
2025-11-12 13:16
(212 d 20:51 ago)

@ Shuanghe
Posting: # 24492
Views: 4,461
 

 Misinterpretation

Hi Shuanghe,

❝ Did any forum member recall the 1st approach of regression for \(\lambda_z\) I described in my post above? I could swear that I've read it somewhere.

See there. This was also my interpretation and what I implemented in some [image]-scripts. I corrected the one in this article.
The package PKNCA was cross-validated with WinNonlin. IMHO, it’s vignette describes clearly that all fits are done and then one is selected..

Dif-tor heh smusma 🖖🏼 Довге життя Україна! [image]
Helmut Schütz
[image]

The quality of responses received is directly proportional to the quality of the question asked. 🚮
Science Quotes
Shuanghe
★★  

Spain,
2025-11-12 22:24
(212 d 11:43 ago)

@ Helmut
Posting: # 24494
Views: 4,420
 

 Misinterpretation

Hi Helmut,
Thanks. Yes, PKNCA's vignette is clear. I still cannot believe that I had the false memory for so many years, and I still think I must've have read it somewhere else about the step-wise regression approach :ponder:

I guess I probably should also thank myself for giving my Phoenix license to my colleague recently and was forced to use R instead of Phoenix for the first time to double check the CROs' results, otherwise I wouldn't have noticed it. :-D

All the best,
Shuanghe
Helmut
★★★
avatar
Homepage
Vienna, Austria,
2025-11-15 10:38
(209 d 23:29 ago)

@ Shuanghe
Posting: # 24499
Views: 4,319
 

 Misinterpretation

Hi Shuanghe,

❝ I still cannot believe that I had the false memory for so many years, and I still think I must've have read it somewhere else about the step-wise regression approach


You are not alone. Perhaps in this post?

An indirect a quote of WinNonlin (1.1 of 1996 ‼):

Linear regressions are repeated using the last three points, the last four points, the last five points etc. For each regression, an adjusted R² is computed:$$\small{Adjusted\,R^2=1-\frac{\left(1-R^2\right)\cdot(n-1)}{n-2}}$$where n is the number of data points in the regression and R² is the square of the correlation coefficient. The regression with the largest adjusted R² is selected to estimate the terminal half-life, with one caveat: if the adjusted R² does not improve, but is within .0001 of the largest value, the regression with the larger number of points is used.


Yet another approach. b

One problem is to find an algorithm that identifies an appropriate range over which the linear regression should be done. One such algorithm, which has long served this author well, but always needs to be supplemented by a sanity check, goes as follows:

  1. Use the last three points to make a linear regression on and compute the R2 to this linear regression.
  2. Successively add earlier time points and make the linear regression of these data points until R2 decreases (i.e., degree of explanation decreases).
  3. When a point for which a R2 decrease is encountered, remove that point and add the preceding one. Repeat the analysis. If R2 still decreases, the process is stopped and the data are included in the linear regression up to the first decrease in R2. If however R2 increases, we consider the removed point an outlier and continue the process with that individual point removed.

This is not a fool-proof method but it usually represents a good starting point. However, sometimes one needs to use only the last two points in the estimate of the terminal slope, which this method does not allow for (no R2 can be computed in that cases). Data must therefore always be inspected in order to allow for manual modification of the time interval over which the regression should be made.



  1. Gieschke R. Half-life. In: Cawello W, editor. Parameters for Compartment-free Pharmacokinetics. Aachen: Shaker Verlag; 2003. p. 55.
  2. Källén A. Computational Pharmacokinetics. Boca Raton: Chapman & Hall/CRC; 2008. p. 43

Dif-tor heh smusma 🖖🏼 Довге життя Україна! [image]
Helmut Schütz
[image]

The quality of responses received is directly proportional to the quality of the question asked. 🚮
Science Quotes
billdenney
☆    

USA,
2026-02-27 15:20
(105 d 18:47 ago)

@ Shuanghe
Posting: # 24581
Views: 2,056
 

 Misinterpretation

❝ Thanks. Yes, PKNCA's vignette is clear. I still cannot believe that I had the false memory for so many years, and I still think I must've have read it somewhere else about the step-wise regression approach :ponder:


❝ I guess I probably should also thank myself for giving my Phoenix license to my colleague recently and was forced to use R instead of Phoenix for the first time to double check the CROs' results, otherwise I wouldn't have noticed it. :-D


I just saw this post. For disclosure, I'm the primary author of PKNCA.

The only way that I know PKNCA's automation for lambda z estimation is different than WinNonlin is one very particular case. Both programs calculate all possible half-life estimates from third to last to the next point after Tmax. PKNCA next selects the best line by adjusted r-squared; WinNonlin next filters for negative slopes (non-increasing half-lives) then for best adjusted r-squared.

The difference is that when the best adjusted r-squared line is increasing, PKNCA will not report an automated half-life estimate. WinNonlin will report a half-life is there is any adjusted r-squared line that is decreasing. Usually, the estimated half-life would not be high-quality (the adjusted r-squared would be low, the span ratio would be low, and the percent extrapolated would be high), so as long as appropriate filtering for good values is in place, these estimates wouldn't make it to a final summary.
Helmut
★★★
avatar
Homepage
Vienna, Austria,
2026-02-27 16:29
(105 d 17:38 ago)

@ billdenney
Posting: # 24582
Views: 2,082
 

 Clarification

Hi Bill,

nice to see you here and THX for the clarification!

Dif-tor heh smusma 🖖🏼 Довге життя Україна! [image]
Helmut Schütz
[image]

The quality of responses received is directly proportional to the quality of the question asked. 🚮
Science Quotes
Helmut
★★★
avatar
Homepage
Vienna, Austria,
2025-11-12 11:30
(212 d 22:38 ago)

@ SDavis
Posting: # 24491
Views: 4,467
 

 Rewording appreciated

Hi Simon,

❝ I think the issue maybe the text is slightly ambiguous and people read into it, "and then you stop", however it always went all the way back […] before selecting the value with the largest Adj-R-Sq. I think "If the adjusted R2 does not improve, but is within 0.0001 of the largest adjusted R2 value, the regression with the larger number of points is used." infers that but I've asked that we make it clearer in future iterations of the help that all points are assessed.

Exactly. I also thought for ages (since PCNONLIN V03.0 in 1991 😜) that it is a stepwise procedure working backwards and not calculating all regressions and finally ‘play the winner’. Might be obvious for you… A rewording will help confused people who are not native speakers of English.

Dif-tor heh smusma 🖖🏼 Довге життя Україна! [image]
Helmut Schütz
[image]

The quality of responses received is directly proportional to the quality of the question asked. 🚮
Science Quotes
jag009
★★★

NJ,
2026-01-29 05:20
(135 d 04:47 ago)

@ Helmut
Posting: # 24559
Views: 2,821
 

 Reason for failing to calculate lambda and Kel in Phoenix

Hi Helmut and others,

Revisiting this thread.... Does Phoenix provide a reason somewhere in the result/output as to why kel cannot be estimated for a subject? I know the Use guide lists the reasons but I am looking for something in the result output relevant to the subject's profile.

I use the best fit method for slope calculation. The semi-log plot just showed the statement "cannot be determined" or something, the "warning and error" output showed "No parameters could be extrapolated to infinity". I didnt see anything in the core output.

thx
J
Helmut
★★★
avatar
Homepage
Vienna, Austria,
2026-01-29 12:04
(134 d 22:03 ago)

@ jag009
Posting: # 24560
Views: 2,818
 

 Reason for failing to estimate λz

Hi John,

❝ … why kel cannot be estimated …

<picky>

\(\begin{array}{lcl}
\color{Red}{\textsf{PK modeling}}&\longrightarrow& \widehat{k}_\text{el}\\
\color{Green}{\textsf{NCA}}&\longrightarrow&\widehat{\lambda}_\text{z}
\end{array}\)

</picky>

❝ … the Use guide lists the reasons [as to why kel cannot be estimated for a subject] but I am looking for something in the result output relevant to the subject's profile.

Not in the output.

❝ I use the best fit method for slope calculation. The semi-log plot just showed the statement "cannot be determined" or something, the "warning and error" output showed "No parameters could be extrapolated to infinity". I didnt see anything in the core output.

Right. After the “best fit” method you should check every single profile for plausibility (as not only recommended in the literature but also in the User’s Guide of PHX/WNL) and correct the selected data if necessary.
Most common cases for ‘failures’:
  • Long half live and last data point close to the LLOQ results in \(\lambda_\text{z}\leq0\). Exclude the increasing concentration. Optionally select an earlier starting point.
  • Low absorption of the drug in this subject. The profile ‘drops’ fast below the LLOQ. See the last paragraph of this post. If the half live estimated from only two data points is consistent with the others, use it.
Of course, you need an SOP for that. Even then, it could cause an assessor to frown.

Dif-tor heh smusma 🖖🏼 Довге життя Україна! [image]
Helmut Schütz
[image]

The quality of responses received is directly proportional to the quality of the question asked. 🚮
Science Quotes
UA Flag
Activity
 Admin contact
23,653 posts in 4,991 threads, 1,570 registered users;
130 visitors (0 registered, 130 guests [including 24 identified bots]).
Forum time: 11:08 CEST (Europe/Vienna)

To propose that poor design can be corrected by subtle analysis techniques
is contrary to good scientific thinking.    Stuart J. Pocock

The Bioequivalence and Bioavailability Forum is hosted by
BEBAC Ing. Helmut Schütz
HTML5