Helmut ★★★ ![]() Vienna, Austria, 2018-07-12 19:47 (2883 d 07:38 ago) Posting: # 19039 Views: 31,769 |
|
|
Dear all, I was asked whether it is possible to validate PhEq_bootstrap (see this post). Sure. Since the source-code is available and if one is familiar with Object Pascal even a “white-box” validation is doable. ![]() I had a quick look based on the example data sets given by Shah et al.* (which is referred in the software’s documentation). I got for the 90% CI (PhEq_bootstrap v1.2, 2014-06-22, 64bit Windows):
Good news. With one exception (test batch 1, 500 bootstraps) the lower CL is more conservative than the ones reported by Vinod. With 1,000 bootstraps results are always slightly more conservative. Now for the bad news. There is one thing stupid in the software when it comes to validation (line 535 of the source Conv_to_learn1.pas):
IMHO, that’s bad coding practice. In our PowerTOST, Power2Stage) we have an option to work either with a fixed or a random seed. With the former (which is the default) it is possible to reproduce a given run. Detlew’s randomizeBE is even more flexible. You can work with a random seed and the output gives the seed used. Then you can use this seed as an argument to reproduce the run. Clever.OK, how reproducible are runs of PhEq_bootstrap?
Imagine that you claimed similarity (ƒ2 50) and an assessor asks you to demonstrate how you obtained it. You fire up the software with its default of 5,000 bootstraps and are slapped in the face with: f2* = 49.93 Similarity NOT confirmed: Lower CI is below the limit (f2 = 50) Oops!Bootstrapping converges to the true value with increasing repetitions. If you want to get reproducible results you have to use a workaround and go for 25,000+ bootstraps. Then the results are stable but you have to accept ~80MB result-files… Of course, this is only valid for these data sets (12 units, 4 sampling time points). Good luck with more time points.
— Dif-tor heh smusma 🖖🏼 Довге життя Україна! ![]() Helmut Schütz ![]() The quality of responses received is directly proportional to the quality of the question asked. 🚮 Science Quotes |
|
ElMaestro ★★★ Denmark, 2018-07-13 00:21 (2883 d 03:04 ago) @ Helmut Posting: # 19040 Views: 28,053 |
|
|
Hi Hötzi, ❝ Bootstrapping converges to the true value with increasing repetitions. If you want to get reproducible results you have to use a workaround and go for 25,000+ bootstraps. Then the results are stable but you have to accept ~80MB result-files… ❝ Of course this is only valid for these data sets (12 units, 4 sampling time points). Good luck with more time points. The magnitude of the variability of the estimate is dependent on the number of bootstraps and the within- and between-variability. This is just the nature of it. What I don't quite get is how the need for 80 Mb files (files, really???) arises. For 25000 bootstraps you should only need to allocate e.g. 25000x a double datatype (8 bytes) for the storage of f2*, which isn't even a megabyte from the heap. But note I did not care to read any of the Pascal source code. However, speed or memory consumption is a mundane issue. If the calculation is proper then no need to whine at all. When I do this in C, I allocate 8 Mbs from the heap, and it takes a blink of an eye or two with a million bootstraps and sorting, and no intermediary files anywhere, and it provides derivation of the bootstrap CI in various flavours: Raw, bias corrected, and bias corrected and accelerated. — Pass or fail! ElMaestro |
|
Helmut ★★★ ![]() Vienna, Austria, 2018-07-13 03:29 (2882 d 23:56 ago) @ ElMaestro Posting: # 19041 Views: 27,927 |
|
|
Hi ElMaestro, ❝ The magnitude of the variability of the estimate is dependent on the number of bootstraps and the within- and between-variability. This is just the nature of it. ❝ What I don't quite get is how the need for 80 Mb files (files, really???) arises. ❝ For 25000 bootstraps you should only need to allocate e.g. 25000x a double datatype (8 bytes) for the storage of f2*, which isn't even a megabyte from the heap. ❝ But note I did not care to read any of the Pascal source code. ❝ However, speed or memory consumption is a mundane issue. If the calculation is proper then no need to whine at all. ❝ When I do this in C, I allocate 8 Mbs from the heap, and it takes a blink of an eye or two with a million bootstraps and sorting, … My Pascal is rusty but lines 162–181 of the source stinks of Bubble Sort. In ❝ … and no intermediary files anywhere, … ❝ … and it provides derivation of the bootstrap CI in various flavours: Raw, bias corrected, and bias corrected and accelerated. — Dif-tor heh smusma 🖖🏼 Довге життя Україна! ![]() Helmut Schütz ![]() The quality of responses received is directly proportional to the quality of the question asked. 🚮 Science Quotes |
|
ElMaestro ★★★ Denmark, 2018-07-13 12:14 (2882 d 15:11 ago) @ Helmut Posting: # 19043 Views: 27,833 |
|
|
Hi Hötzi, ❝ Dokumentation, darling! The file contains not only the four estimates (ƒ1, ƒ2, unbiased ƒ2, ƒ2*) in full precision for the X bootstraps but also the raw data (+average, variance) of all samples. That’s a fucking lot. Yes, that is a lot. Perhaps a lot of time is spent writing the datasets to file streams and flushing file buffers. Still, let us say 6 time points, 12 chambers, 2 formulation: ~144 Mb, plus the array of f2*. Allocating that sort of memory should not be an issue ever on newer hardware. The browser you have opened right now to read this post consumes a lot more, probably by a factor 4-10 ![]() What is the use of having all that info cached?? Under bootstrapping the i'th dataset is as valid as the j'th dataset where (i,j) is in ([0;B], [0;B]). When I write resampling code I usually just extract a few dataset and print e.g. the 111'th sample or whatever and check that it has the right set of elements (i.e. that the 111'th dataset (or whatver) for Test only has Test values at the 'right' time points and so forth). And I can check that the derived f2* is one I can reproduce by manual calculation in another application. It isn't such a lot of work. But if I had to do that for a million datasets, then I'd be very, very late for dinner. What I find little more tricky is to convince myself that the derivation of CI's from the vector of f2* values is valid. A key element here is the sorting. What I did once was to output the sorted first 65536 rows of the sorted array and import them in a spreadsheet (if I recall correctly 65536 is the maximum number of rows in my spreadsheet). Then copy those values to another column, sort this column, and check if row difference exist. A row difference would indicate a sorting mishap in one of the applications. Never observed it, fortunately. — Pass or fail! ElMaestro |
|
Helmut ★★★ ![]() Vienna, Austria, 2018-07-13 14:54 (2882 d 12:31 ago) @ ElMaestro Posting: # 19045 Views: 27,943 |
|
|
Hi ElMaestro, ❝ ❝ Dokumentation, darling! […] ❝ Yes, that is a lot. Perhaps a lot of time is spent writing the datasets to file streams and flushing file buffers. ❝ […] ~144 Mb, plus the array of f2*. Allocating that sort of memory should not be an issue ever on newer hardware. ❝ What is the use of having all that info cached?? ❝ Under bootstrapping the i'th dataset is as valid as the j'th dataset where (i,j) is in ([0;B], [0;B]). When I write resampling code I usually just extract a few dataset and print e.g. the 111'th sample or whatever and check that it has the right set of elements (i.e. that the 111'th dataset (or whatver) for Test only has Test values at the 'right' time points and so forth). ❝ What I find little more tricky is to convince myself that the derivation of CI's from the vector of f2* values is valid. A key element here is the sorting. ❝ What I did once was to output the sorted first 65536 rows of the sorted array and import them in a spreadsheet (if I recall correctly 65536 is the maximum number of rows in my spreadsheet). Then copy those values to another column, sort this column, and check if row difference exist. A row difference would indicate a sorting mishap in one of the applications. Tricky is the CI. Try it with a very small data set and use the spreadsheet’s PERCENTILE() function. I was suspicious when I looked at the source of PhEq_bootstrap whether it will be the same.Example in
PCTLDEF=5). All spreadsheets I have tested (Excel, OO Calc, Gnumeric) use “Type 7”. Of course, with already moderate samples differences will practically disappear. But checking sumfink like this is the fun of software validation, right?— Dif-tor heh smusma 🖖🏼 Довге життя Україна! ![]() Helmut Schütz ![]() The quality of responses received is directly proportional to the quality of the question asked. 🚮 Science Quotes |
|
ElMaestro ★★★ Denmark, 2018-07-13 16:28 (2882 d 10:57 ago) @ Helmut Posting: # 19048 Views: 27,534 |
|
|
Hi again, re. the index: A simple solution is not to sample 1000000 times but 999999. Another solution is to sample so many times that the result is the same regardless of whether you round up or down when finding the index of the 5% percentile etc. This, however, takes a bit of time to figure out. — Pass or fail! ElMaestro |
|
Shuanghe ★★ Spain, 2018-07-17 17:54 (2878 d 09:31 ago) @ Helmut Posting: # 19077 Views: 27,641 |
|
|
Hi Helmut! ❝ Now for the bad news. There is one thing stupid in the software when it comes to validation (line 535 of the source
What a coincidence! I was in a meeting recently with a regulatory agency and the issue of reproducibility came up since the agency couldn't reproduce the bootstrap f2 values we reported in the dossier. My first thought was the different seed number. I opened the source files one by one and searched words like random, seed, to get the line. ❝ IMHO, that’s bad coding practice. Indeed. End user should be able to specify the seed number to be able to reproduce the analysis. I always leave a variable for seed number in my SAS code for bootstrap of BE data. — All the best, Shuanghe |
|
d_labes ★★★ Berlin, Germany, 2018-08-15 11:45 (2849 d 15:41 ago) @ Helmut Posting: # 19168 Views: 27,682 |
|
|
Dear Helmut, dear All, just found: bootf2BCA, a R equivalent of PhEq_bootstrap. Contained are scripts for the R console and a GUI based on the shiny package. May be better suited for validation since the source code may be modified to fit your needs, without access to the rarely used Object Pascal. E.g. introduce a setting of the seed to assure the same results between different runs. Be warned: The source code is hard to read, nearly no comments at all. Be further warned: If you rise the number of bootstrap samples (default is 1000, too low IMHO) the result file will contain a huge number of entries, each bootstrap sample is recorded! The same mess as in the 'Object Pascal' implementation. But could of course also changed with a little knowledge of R. — Regards, Detlew |
|
Helmut ★★★ ![]() Vienna, Austria, 2019-10-05 02:59 (2434 d 00:27 ago) @ d_labes Posting: # 20665 Views: 25,974 |
|
|
Dear all, we discussed the code at BioBridges… The problem with reproducibility can easily be solved. In the script F2_boot.R after line 270 #main algorithm add a new line
X is an integer (specified in the protocol and report). Use
❝ Be warned: The source code is hard to read, nearly no comments at all. The current version 1.2 is a little bit better. Still heavy stuff. The cut-off preferred by the EMA is implemented:
Still missing: One value >85% for the test (FDA) or one value >85% for the comparator (WHO). ❝ Be further warned: If you rise the number of bootstrap samples (default is 1000, too low IMHO) the result file will contain a huge number of entries, each bootstrap sample is recorded! The same mess as in the 'Object Pascal' implementation. But could of course also changed with a little knowledge of R. Hint: In F2_boot.R comment out the loop in lines 317–329. Then the report-file shrinks from a couple of megabytes to some kilobytes. I would keep the rest for documentation. Useful to read the bootstrapped ƒ2 values from the file and generate your own histogram (density instead of counts). I don’t like the one of plotly. Way too many bins for my taste (here an example with 5,000 bootstraps). \(\small{\textrm{ceiling}(\sqrt n)}\) gives 71 bins and smells of Excel.
I prefer the Freedman–Diaconis rule* and show additionally the median and 90% CI (green line if the lower CL ≥50).
— Dif-tor heh smusma 🖖🏼 Довге життя Україна! ![]() Helmut Schütz ![]() The quality of responses received is directly proportional to the quality of the question asked. 🚮 Science Quotes |
|
Helmut ★★★ ![]() Vienna, Austria, 2020-09-25 21:37 (2077 d 05:49 ago) @ Helmut Posting: # 21940 Views: 22,989 |
|
|
Dear all, Alfredo García-Arieta (Spanish agency AEMPS) notified me that in v1.2 there is no need any more for the workaround to get a fixed seed. Don’t from the top-folder at SourceForge. That’s still v1.1 of October 2018. Instead
At the end of report.txt the seed is given for reproducibility.— Dif-tor heh smusma 🖖🏼 Довге життя Україна! ![]() Helmut Schütz ![]() The quality of responses received is directly proportional to the quality of the question asked. 🚮 Science Quotes |
|
PTM139 ☆ India, 2021-04-16 09:53 (1874 d 17:33 ago) @ Helmut Posting: # 22309 Views: 21,024 |
|
|
Dear Helmut, Although downloaded v 1.2 but it is not working from me. GUI is not opening as it happens with Ph_Eq bootstrap. Any specific steps to be followed? Getting below message when I open GUI file: #!/bin/shRegards, Jayant Edit: Full quote removed. Please delete everything from the text of the original poster which is not necessary in understanding your answer; see also this post #5! [Helmut] |
|
Helmut ★★★ ![]() Vienna, Austria, 2021-04-16 12:50 (1874 d 14:35 ago) @ PTM139 Posting: # 22310 Views: 20,933 |
|
Hi Jayant,❝ […] not working … … is not a proper description of a software-related problem. I don’t have a crystal ball. Operating system, release of Install devtools from CRAN and run this script:
boot, ggplot2, plotly, rlang, and shiny. Their current versions are: boot 1.3-27, ggplot2 3.3.3, plotly 4.9.3, rlang 0.4.10, shiny 1.6.0. If one or more are missing our outdated, see this article about installation/updating of packages.Alternatively you can execute this script:
pops up, click on (the central button – it’s actual label depends on the locale of your operating system). Your working directory (as reported by getwd()) has to be the one where the scripts of bootf2BCA are (the data files can be somewhere else). If not, in the menu bar of the
— Dif-tor heh smusma 🖖🏼 Довге життя Україна! ![]() Helmut Schütz ![]() The quality of responses received is directly proportional to the quality of the question asked. 🚮 Science Quotes |
|
PTM139 ☆ India, 2021-04-27 15:02 (1863 d 12:23 ago) @ Helmut Posting: # 22330 Views: 20,457 |
|
|
Thank you for your prompt reply. I will follow the instructions and try to install all the packages. Regards, Jayant |
|
Rayhope ☆ India, 2022-03-28 16:35 (1528 d 10:50 ago) @ Helmut Posting: # 22881 Views: 17,958 |
|
|
❝ Alfredo García-Arieta (Spanish agency AEMPS) notified me that in v1.2 there is no need any more for the workaround to get a fixed seed. Currently V 1.3 is available and i guess they fixed lot of issues in current version. @Helmut I am using V1.3 and in software bundle source code was not found. any chance to get source code if its available with you. I have verified the results with Pheq 1.2 Vs Bootf2bca v 1.3 and found similar result using article examples (Comparison of free software platforms for the calculation of the 90% confidence interval of f2 similarity factor by bootstrap analysis L. Nocea,b, L. Gwazac, V. Mangas-Sanjuana,b,⁎, A. Garcia-Arietad. European Journal of Pharmaceutical Sciences.) Very minor differences which corrected with rounding off. |
|
Helmut ★★★ ![]() Vienna, Austria, 2022-03-28 16:48 (1528 d 10:37 ago) @ Rayhope Posting: # 22882 Views: 18,272 |
|
|
Hi Rayhope, ❝ Currently V 1.3 is available and i guess they fixed lot of issues in current version. NEWS-file (which would be common on CRAN and GitHub). If you are interested, you could only compare the source with the one of v1.2.❝ @Helmut I am using V1.3 and in software bundle source code was not found. any chance to get source code if its available with you. — Dif-tor heh smusma 🖖🏼 Довге життя Україна! ![]() Helmut Schütz ![]() The quality of responses received is directly proportional to the quality of the question asked. 🚮 Science Quotes |
|
Rayhope ☆ India, 2022-03-30 08:34 (1526 d 18:51 ago) @ Helmut Posting: # 22885 Views: 17,829 |
|
|
❝ Use this link to download the zip-file. Thanks Helmut for the link. |
|
Flam ☆ New Zealand, 2023-06-21 05:41 (1078 d 21:44 ago) @ Helmut Posting: # 23609 Views: 15,756 |
|
|
❝ Hint: In I am a first time R user trying to use bootf2BCA v1.3, how can I access the bootstrapped f2 values to draw the histogram etc.? |
|
Helmut ★★★ ![]() Vienna, Austria, 2023-06-23 15:47 (1076 d 11:39 ago) @ Flam Posting: # 23613 Views: 15,867 |
|
|
Hi Flam, ❝ I am a first time R user trying to use bootf2BCA v1.3, how can I access the bootstrapped f2 values to draw the histogram etc.? bootf2BCA_v1.3.1, did you succeed in specifying the result file in shiny?— Dif-tor heh smusma 🖖🏼 Довге життя Україна! ![]() Helmut Schütz ![]() The quality of responses received is directly proportional to the quality of the question asked. 🚮 Science Quotes |
|
Flam ☆ New Zealand, 2023-06-27 01:16 (1073 d 02:09 ago) @ Helmut Posting: # 23615 Views: 15,727 |
|
|
❝ Before I start to adapt my old workaround to No, I have not been able to install shiny. Edit: Full quote removed. Please delete everything from the text of the original poster which is not necessary in understanding your answer; see also this post #5! [Helmut] |
|
Helmut ★★★ ![]() Vienna, Austria, 2023-06-27 13:47 (1072 d 13:39 ago) @ Flam Posting: # 23620 Views: 15,519 |
|
|
Hi Flam, ❝ No, I have not been able to install shiny. For the future, see Eric Steven Raymond’s essay “How To Ask Questions The Smart Way”. ![]() Execute this script in the
pops up, click on . — Dif-tor heh smusma 🖖🏼 Довге життя Україна! ![]() Helmut Schütz ![]() The quality of responses received is directly proportional to the quality of the question asked. 🚮 Science Quotes |
|
Rayhope ☆ India, 2021-07-29 10:49 (1770 d 16:37 ago) @ Helmut Posting: # 22485 Views: 20,085 |
|
|
❝ I was asked whether it is possible to validate PhEq_bootstrap (see this post). Sure. Since the source-code is available and if one is familiar with Object Pascal even a “white-box” validation is doable. Hi Helmut, similar query received from agency for validation of PhEq results. Can you help me out with simple steps to do that. I dont know about programming or validation of software. I read all conversations related to your post, I came to conclusion that Validation of PhEq and BootF2CA is not at all my cup of tea. But need to reply on Agency query how to proceed. Please guide me. Also let me know if you can provide SAS code for bootstrap mentioned in article of Shah et al. (In vitro Dissolution Profile Comparison-statistics and Analysis of similarity Factor F2, Pharmaceutical Research Vol 15, no.6 1998). Disclaimer: I don't have any basic knowledge of Coding in any language. |
|
Helmut ★★★ ![]() Vienna, Austria, 2021-08-02 20:12 (1766 d 07:14 ago) @ Rayhope Posting: # 22499 Views: 19,824 |
|
|
Hi Rayhope, ❝ similar query received from agency for validation of PhEq results. ❝ Can you help me out with simple steps to do that. I dont know about programming or validation of software. I read all conversations related to your post, I came to conclusion that Validation of PhEq and BootF2CA is not at all my cup of tea. ❝ But need to reply on Agency query how to proceed. […] the results demonstrate the adequacy of PhEq Bootstrap and Bootf2bca software to replicate the observed results (Shah et al., 1998). As expected, the 90% CI were more similar among the software considered when the number of bootstrap replicates increased, although the observed differences could be explained by the seed number considered by each computer and the computer processor. The external validation shows that similarity conclusion for each comparison was irrespective of the software considered. As shown already in this post, in the current version of bootf2BCA a fixed seed of 100 is used by default and documented in the output. Hence, results are reproducible – independent from the machine and operating system.❝ Also let me know if you can provide SAS code for bootstrap […] Sorry, I’m not equipped with![]()
— Dif-tor heh smusma 🖖🏼 Довге життя Україна! ![]() Helmut Schütz ![]() The quality of responses received is directly proportional to the quality of the question asked. 🚮 Science Quotes |


![[image]](https://static.bebac.at/pics/Blue_and_yellow_ribbon_UA.png)
![[image]](https://static.bebac.at/img/CC by.png)
![[image]](img/uploaded/image618.png)
![[image]](img/uploaded/image619.png)
![[image]](img/uploaded/image620.png)
and then![[image]](img/uploaded/image730.png)
Hi Jayant,
Do you want to install from sources the packages which need
![[image]](img/uploaded/image2.gif)