Dear All,

What is the importance of Null ratio (=test/reference) in sample size calculations?

It should be 1 (test/reference=1) or 0.95 or 1.05 (5% variability in test/reference).

I am using paried equivalence test in SAS for sample size calculation where it asked for null ratio. Here I am confused :confused: whether to use 1 or 0.95 or to be consider actual variability (geometric mean ratio of test/ref untransformed data) received from the reference/pilot data.

I gone through many guidelines but none has mentioned clearly about null ratio. ANVISA says take difference 0% or 5%. USFDA suggested to consider 5% variability. pls guide what is appropriate?
Is it appropriate to consider actual variability as a null ratio :ponder:

regards,
Nirali.
Helmut
★★★
avatar
Homepage
Vienna, Austria,
2007-11-02 18:05
(6399 d 15:05 ago)

@ Nirali
Posting: # 1266
Views: 7,814
 

 Expected deviation of test from reference; actual content

Dear Nirali!

❝ What is the importance of Null ratio (=test/reference) in sample size calculations?


Since power curves become very steep if moving the test/reference ratio away from 1 (for examples see this post) it's quite important to use reasonable estimates (of both the expected deviation of test from reference and CVintra).

❝ I gone through many guidelines but none has mentioned clearly about null ratio. ANVISA says take difference 0% or 5%. USFDA suggested to consider 5% variability. pls guide what is appropriate?

❝ Is it appropriate to consider actual variability as a null ratio


The magical number ±5% has its origin in the following reasoning:
Specifications for batch release commonly are set to ±5% of the declared content. If both products show identical BA, you may expect in your study a deviation of ±5% (let’s say, test’s content -2.5% and reference’s content +2.5% of nominal).

Nwakama P, Haidar S, Yang Y, Davit B, Conner D, Yu L. Generic Drug Products Demonstrate Small Differences in Bioavailability Relative to the Brand Name Counterparts: A Review of ANDAs Approved 1996 – 2004. AAPS J. 2005;7/S2:Abstract M1262. online abstract


... report from 1411 studies submitted to the FDA deviations of 3.13% (±2.71) for AUCt, 3.05% (±2.62) for AUCinf, and 4.52% (±3.57) for Cmax.
A more recent metaanalysis of 1636 BE studies submitted to the FDA within 1996-2005 showed deviations of 3.19% (±2.72) for AUCt, 3.12% (±2.66) for AUCinf, and 4.50% (±3.57) for Cmax.
If you have no clue, expect the true ratio to be 1 and want to be very conservative you may even go with an expected deviation of ±10%.

Power curves are not symmetrical; if you want to go with a 5% deviation of test from reference and have no clue about which direction the deviation will point to, it's wise to use a ratio of 0.95 instead of 1.05. Power is lower for ratios <1 than for ratios >1; therefore you should apply the most conservative approach.

At least some guidelines are specific:
Canada calls for a sample size estimation based on the expected deviation, but wants to see two evaluations - one with original data and one with data corrected for actual content of study batches.
The European guideline calls for a sample size estimation based on the expected deviation; a potency correction is currently under discussion for the next revision.
The WHO’s guideline also suggests the expected deviation; a dose correction (which was suggested in earlier versions) is currently not recommended.

As guidelines are ambiguous in their recommendations about potency correction, some people opt for this procedure:
  • No potency correction if deviation of actual contents ≤5%,
  • Potency corrected analysis primary if deviation >5%; original data supportive only.

Dif-tor heh smusma 🖖🏼 Довге життя Україна! [image]
Helmut Schütz
[image]

The quality of responses received is directly proportional to the quality of the question asked. 🚮
Science Quotes
UA Flag
Activity
 Admin contact
23,424 posts in 4,927 threads, 1,668 registered users;
133 visitors (0 registered, 133 guests [including 7 identified bots]).
Forum time: 10:11 CEST (Europe/Vienna)

Lack of clarity is always a sign of dishonesty.    Celia Green

The Bioequivalence and Bioavailability Forum is hosted by
BEBAC Ing. Helmut Schütz
HTML5