Including duplicates of Low, medium and High QCs during analytical method validation [Regulatives / Guidelines]

posted by dshah  – India, 2023-06-15 10:36 (609 d 02:00 ago) – Posting: # 23597
Views: 1,922

Hello Ayman!

ICH provides this information in QC sample preparation.

❝ "For non-accuracy and precision validation runs, low, medium and high QCs may be analysed in duplicate. These QCs, along with the calibration standards, will provide the basis for the acceptance or rejection of the run."


❝ are the stability runs which contains low and high QC levels are considered non-accuracy run?

❝ what are the non-accuracy runs during the method validation and how can the duplicates of the low, med and high QCs be assessed to decide the run accepted or rejected if the run itself not designed for the accuracy (recovery as example or carry over).

The acceptance criteria is: For non-accuracy and precision validation runs, at least 2/3 of the total QCs and at least 50% at each concentration level should be within ±20% of the nominal values.

It is not necessary to have replicate QC levels but they help in acceptance/rejection of run.

Regards,
Divyen

Complete thread:

UA Flag
Activity
 Admin contact
23,380 posts in 4,914 threads, 1,661 registered users;
33 visitors (0 registered, 33 guests [including 2 identified bots]).
Forum time: 11:36 CET (Europe/Vienna)

Everything is trivial, if you know the answer.    Thomas Jaki

The Bioequivalence and Bioavailability Forum is hosted by
BEBAC Ing. Helmut Schütz
HTML5