Including duplicates of Low, medium and High QCs during analytical method validation [Regulatives / Guidelines]
Hello Ayman!
ICH provides this information in QC sample preparation.
It is not necessary to have replicate QC levels but they help in acceptance/rejection of run.
Regards,
Divyen
ICH provides this information in QC sample preparation.
❝ "For non-accuracy and precision validation runs, low, medium and high QCs may be analysed in duplicate. These QCs, along with the calibration standards, will provide the basis for the acceptance or rejection of the run."
❝
❝ are the stability runs which contains low and high QC levels are considered non-accuracy run?
❝ what are the non-accuracy runs during the method validation and how can the duplicates of the low, med and high QCs be assessed to decide the run accepted or rejected if the run itself not designed for the accuracy (recovery as example or carry over).
It is not necessary to have replicate QC levels but they help in acceptance/rejection of run.
Regards,
Divyen
Complete thread:
- Including duplicates of Low, medium and High QCs during analytical method validation Ayman Rabayah 2023-06-15 07:14 [Regulatives / Guidelines]
- Including duplicates of Low, medium and High QCs during analytical method validationdshah 2023-06-15 08:36