acceptance and bonus info [Bioanalytics]

posted by ElMaestro  – Denmark, 2022-11-13 21:57 (522 d 13:51 ago) – Posting: # 23362
Views: 1,759

Hi Ladi,

❝ What should be the acceptance criteria for the blank and zero samples that we inject before calibration curve standards in analytical run?


❝ According to the ICH M10 guideline on bioanalytical method validation, section 3.3.1 Analytical run, states that "An analytical run consists of a blank sample (processed matrix sample without analyte and without IS), a zero sample (processed matrix with IS)".


❝ However, I couldn't find the acceptance criteria for the blank and zero sample in the guideline.


Technically that seems correct.


❝ Our current SOP is to the accept the analytical run if interference at RT of analyte in blanK and zero sample is less than 20% compare to LLOQ and if interference at RT of IS is less than 5% compare to IS response in LLOQ (basically follow Selectivity criteria).


Sounds good to me.

❝ I have seen many cases where some runs are rejected because the blank sample is contaminated from sample preparation error (zero sample and pre-dose samples are clean).


❝ I personally don't think it make sense to reject a whole run from one contaminated sample!


I guess this is your major point.
On the other hand, one can also argue it does not make sense to inject a run if there are contaminated samples. Much of it boils down to the time by which you conclude there is contamination.
If you observe interference (unwanted signal in the blank or zero) at the RT, and you do an investigation and conclude that there was contamination in the blank / zero but certainly not in any other injection then you are almost begging for an inspection that is very difficult for you from the very beginning. Secondly, if you conclude there was contamination, how is that conclusion actually reached? If there is a source document stating that someone put ref std in the blank then we're back to the first counter-argument. If you conclude that contamination was present because you think you have ruled out all other potential explanations then we're back at the second counter-argument.

To me, it makes perfect sense to reject a run if there is "contamination" in the blank / zero. The alternative is much worse.


❝ Any suggestion or guidelines, papers I can use to use as reference to modify our SOP?


Your SOP is good. If you too often can't comply with it then it may be something else entirely that is your real problem.


Along the same lines, bonus info:
Quite a few CROs are currently experiencing sporadic carry-over issues with analytes measured in the the pg/mL range. I recommend sponsors to always insist on also having a double blank right after the highest calibrator in the PK-runs and to have some rules in-house for handling situations when there is CO in this injection. CROs really do not like it, simply because they realise their assays at study-time is not behaving as well as it may have been at the time of validation. The solution is certainly not to only insist that CO is assessed as part of the validation, not as part of the study.

Pass or fail!
ElMaestro

Complete thread:

UA Flag
Activity
 Admin contact
22,987 posts in 4,824 threads, 1,661 registered users;
71 visitors (0 registered, 71 guests [including 3 identified bots]).
Forum time: 12:48 CEST (Europe/Vienna)

The only way to comprehend what mathematicians mean by Infinity
is to contemplate the extent of human stupidity.    Voltaire

The Bioequivalence and Bioavailability Forum is hosted by
BEBAC Ing. Helmut Schütz
HTML5