Hi Loky do,
My personal points of view…
Why? You measured them (confirmed even in a second analysis).
Do you mean setting them to zero? If yes, that would not make sense.
BTW, replacing BQLs by LLOQ/2 does not make sense either.1
Any nonnumeric code would be fine.2 However, I recommend to give them with the more common code ‘BQL’ and use the linear-up / logarithmic-down trapezoidal rule. If you use the linear trapezoidal rule, the AUC will be positively biased (i.e., the value ignored and a trapezoid calculated connecting the time point before and after with a straight line). See this article (esp. the example about missing values).
Of note, the lin-up/log-down trapezoidal is recommended in the trainings for Phoenix/WinNonlin (Certara) and PKanalix (LIXOFT). It is the default in the -packages
❝ Regarding BLQ samples in the middle of plasma concentration curves, which remain BLQ after reanalysis, what is the best practice for those samples in calculations?
My personal points of view…
❝ To be considered as (missed or Non-reported samples …
Why? You measured them (confirmed even in a second analysis).
❝ … or blank)?
Do you mean setting them to zero? If yes, that would not make sense.
BTW, replacing BQLs by LLOQ/2 does not make sense either.1
❝ Or to keep it as it is <LLOQ?
Any nonnumeric code would be fine.2 However, I recommend to give them with the more common code ‘BQL’ and use the linear-up / logarithmic-down trapezoidal rule. If you use the linear trapezoidal rule, the AUC will be positively biased (i.e., the value ignored and a trapezoid calculated connecting the time point before and after with a straight line). See this article (esp. the example about missing values).
Of note, the lin-up/log-down trapezoidal is recommended in the trainings for Phoenix/WinNonlin (Certara) and PKanalix (LIXOFT). It is the default in the -packages
bear
, PKNCA
, and ncappc.- I use it only for plots since I always give the geometric mean.
- If you want to make assessors, inspectors, and consultants happy: Use more informative codes in the bioanalytical report, e.g.,
- ‘BQL’ for values which where <LLOQ,
- ‘NR’ for not reportable ones (say, problems in sample processing or chromatography),
- ‘missing’ for no sample (vial broken, etc.).
- ‘BQL’ for values which where <LLOQ,
—
Dif-tor heh smusma 🖖🏼 Довге життя Україна!
Helmut Schütz
The quality of responses received is directly proportional to the quality of the question asked. 🚮
Science Quotes
Dif-tor heh smusma 🖖🏼 Довге життя Україна!
Helmut Schütz
The quality of responses received is directly proportional to the quality of the question asked. 🚮
Science Quotes
Complete thread:
- BLQ considerations Loky do 2022-03-28 12:16 [Bioanalytics]
- BLQ considerations dshah 2022-03-28 14:14
- ‘Embedded’ BQLsHelmut 2022-03-28 14:32