recently I saw a deficiency of a European agency:
The Applicant should specify the integration method (automatic or manual) adopted to calculate chromatograms area for all submitted studies. If manual integration was adopted, the Applicant should specify the percentage of the total chromatograms and specify which samples (QC/calibrator or unknown) have been analysed.
Sure, that’s according to the 2011 guideline stating:5.5. Integration
Chromatogram integration and re-integration should be described in a SOP. Any deviation from this SOP should be discussed in the analytical report. Chromatogram integration parameters and in case of re-integration, initial and the final integration data should be documented at the laboratory and should be available upon request.
6.6 MANUAL INTEGRATION
Manual integration is PROHIBITED in any circumstance.
An experienced analyst outperformed the software easily (esp. at low concentrations). Even ‘normal’ ones did so on the average.*
$$\small{CV\;(n=10)}$$$$\small{\begin{array}{lcc}
\text{integration method} & \text{1 ng/mL} & \text{0.1 ng/mL}\\\hline
\text{automated (smoothing 1, bunching 2)} & 6.5\% & 15.1\%\\
\text{manual correction (experienced analyst)} & 6.3\% & 11.1\%\\
\text{manual correction (ten analysts)} &5.2\% & 12.8\%\\
& (3.8\%-6.8\%) & (6.9\%-16.0\%)\\\hline
\end{array}}$$ See also here (slides 26–29) and there (slides 24–31).
- Kirchherr H. Data Evaluation in LC-MS. In: Kuss H-J, Kromidas S, editors. Quantification in LC and GC. Wiley 2009; p 243–59.
Dif-tor heh smusma 🖖🏼 Довге життя Україна!
Helmut Schütz
The quality of responses received is directly proportional to the quality of the question asked. 🚮
Science Quotes
Complete thread:
- Manual integrationHelmut 2021-12-01 13:09 [Bioanalytics]
- Manual integration ElMaestro 2021-12-01 14:40
- Manual integration Helmut 2021-12-01 17:38
- Manual integration ElMaestro 2021-12-01 14:40