Dear selwa ts,

❝ (excuse me for my English I'm more French speaking)


So am I ;-)
Few people on this forum are native English speakers anyway.

❝ First i would like to know how to insert suitability test solutions SS in a bach.


The aim of the system suitability test is to make sure that your instrument is working correctly before you start injecting your study samples: peak shape, resolution, sensitivity of the detector, stability of the signal, background noise... These injections will be done at the very beginning of the run, before injecting the actual study samples (calibration samples, QC samples and subject samples).

❝ Secondly I want to know how to prepare them, are those solutions at the same concentration of that of CQ, but rather aqueous, or another level of concentration?


There is nothing specified in any guideline. The usual practice is to have a concentration within your measurement range. You can have it at the same concentration of a QC or standard, or at a different concentration. You can use several levels of concentration.

❝ Do we have to prepare them with an scr or WS, and must we add some internal standard to it?


Yes, you definitely need to have IS in it. Many labs will prepare a system suitability solution which will contain the analyte and IS, and will use it for the duration for which they have established solution stability. Whichever solution you start from is your choice.

Regards
Ohlbe

Complete thread:

UA Flag
Activity
 Admin contact
22,424 posts in 4,694 threads, 1,600 registered users;
21 visitors (2 registered, 19 guests [including 10 identified bots]).
Forum time: 23:51 CET (Europe/Vienna)

The rise of biometry in this 20th century,
like that of geometry in the 3rd century before Christ,
seems to mark out one of the great ages or critical periods
in the advance of the human understanding.    R.A. Fisher

The Bioequivalence and Bioavailability Forum is hosted by
BEBAC Ing. Helmut Schütz
HTML5