Deviation from the mean response [Bioanalytics]

posted by Helmut Homepage – Vienna, Austria, 2020-08-07 15:28 (1792 d 16:32 ago) – Posting: # 21832
Views: 10,330

Hi dshah,

❝ As per standard SOP - "If the response of internal standard in a sample varies by more than 60% for the isotopically labeled internal standard from the mean ISTD response of CC and QC samples of a particular entire run, then repeat the analysis of the sample under reason Significant variations in response of internal standard"


Let’s try that with Shuanghe’s example:

data          <- read.csv("https://bebac.at/downloads/Post21469.csv",
                          sep = ",", dec = ".")
data          <- data[which(data$type == "C" | data$type == "Q"), c(1:2, 4)]
res           <- cbind(data, mean = mean(data$IS.area))
res$deviation <- sprintf("%+.4f%%", 100*(res$IS.area-res$mean)/res$mean)
print(res, row.names = FALSE)

 injection type  IS.area     mean deviation
         1    C 462564.6 490891.8  -5.7706%
         2    C 427928.5 490891.8 -12.8263%
         3    C 436855.6 490891.8 -11.0078%
         4    C 462800.4 490891.8  -5.7225%
         5    C 467778.6 490891.8  -4.7084%
         6    C 481169.4 490891.8  -1.9806%
         7    C 479012.9 490891.8  -2.4199%
         8    C 435110.2 490891.8 -11.3633%
         9    C 464425.2 490891.8  -5.3915%
        10    C 483497.2 490891.8  -1.5064%
        22    Q 470286.3 490891.8  -4.1976%
        29    Q 479924.4 490891.8  -2.2342%
        36    Q 457085.6 490891.8  -6.8867%
        44    Q 479256.7 490891.8  -2.3702%
        51    Q 462271.3 490891.8  -5.8303%
        58    Q 474794.6 490891.8  -3.2792%
        66    Q 480581.6 490891.8  -2.1003%
        73    Q 463233.0 490891.8  -5.6344%
        80    Q 475543.7 490891.8  -3.1266%
        88    Q 476850.6 490891.8  -2.8604%
        95    Q 470929.6 490891.8  -4.0665%
       102    Q 469649.8 490891.8  -4.3272%
       110    Q 483853.7 490891.8  -1.4337%
       117    Q 543378.7 490891.8 +10.6922%
       124    Q 529264.0 490891.8  +7.8168%
       132    Q 647562.5 490891.8 +31.9155%
       139    Q 602118.0 490891.8 +22.6580%
       146    Q 677244.1 490891.8 +37.9620%

With 60% limited deviation of the quoted SOP there would be no reason to repeat the run.

❝ However - there is no guideline which specifies the limit for variation for IS. Thus, it could be 30/40/50/60 - as per CRO's SOP/requirement/Experience.


Correct. Here only your 30% would “work”.
IMHO, this approach is problematic. It depends on many things: The number of CCs and QCs, where they are placed within the run, etc. In Shuanghe’s example we have 22 before troubles started (according to the broken-stick regression at injection 107) and 6 after. Hence, the overall mean of 490892 is misleading (closer to the 466434 in the 1st limb than to the 580570 in the 2nd). You would have to be very restrictive to set the maximum deviation.

Little bit less restrictive if we look only at the QCs (acc. to the FDA’s guidance and the ICH draft):

 injection type  IS.area     mean deviation
        22    Q 470286.3 507990.4  -7.4222%
        29    Q 479924.4 507990.4  -5.5249%
        36    Q 457085.6 507990.4 -10.0208%
        44    Q 479256.7 507990.4  -5.6563%
        51    Q 462271.3 507990.4  -9.0000%
        58    Q 474794.6 507990.4  -6.5347%
        66    Q 480581.6 507990.4  -5.3956%
        73    Q 463233.0 507990.4  -8.8107%
        80    Q 475543.7 507990.4  -6.3873%
        88    Q 476850.6 507990.4  -6.1300%
        95    Q 470929.6 507990.4  -7.2956%
       102    Q 469649.8 507990.4  -7.5475%
       110    Q 483853.7 507990.4  -4.7514%
       117    Q 543378.7 507990.4  +6.9663%
       124    Q 529264.0 507990.4  +4.1878%
       132    Q 647562.5 507990.4 +27.4753%
       139    Q 602118.0 507990.4 +18.5294%
       146    Q 677244.1 507990.4 +33.3183%


❝ The interesting thing could be - whether the QC's analyte/IS standard response are within acceptable range - i.e. whether the QC's were acceptable?


In Shuanghe’s example the run passed. AFAIK, the CRO didn’t assess the IS response at all.

❝ I am wondering what other things can be directly correlated …


Some possibilities were discussed before.

❝ … and what can be the reason for CRO's acceptability of variation in IS response?


I can only guess (arbitrary order):Shuanghe is on vacation. Maybe he can share some insights once back.

Dif-tor heh smusma 🖖🏼 Довге життя Україна! [image]
Helmut Schütz
[image]

The quality of responses received is directly proportional to the quality of the question asked. 🚮
Science Quotes

Complete thread:

UA Flag
Activity
 Admin contact
23,427 posts in 4,929 threads, 1,684 registered users;
75 visitors (0 registered, 75 guests [including 20 identified bots]).
Forum time: 08:01 CEST (Europe/Vienna)

In the field of observation,
chance favors only the mind that is prepared.    Louis Pasteur

The Bioequivalence and Bioavailability Forum is hosted by
BEBAC Ing. Helmut Schütz
HTML5