Duplicate LLOQ and ULOQ [Bioanalytics]

posted by Helmut Homepage – Vienna, Austria, 2020-05-27 16:30 (428 d 19:41 ago) – Posting: # 21484
Views: 5,543

Dear Ohlbe,

» Most frequent: use the first result only. If, and only if it fails (more than 20% deviation at the LLOQ, 15% at the ULOQ): exclude it and use the second. The aim being of course to avoid having to truncate the curve in case of failure.
»
» I see this less and less often. Not sure why (guess: some FDA inspectors did not like it).

Count me in. If you analyzed duplicates, use them.

Dif-tor heh smusma 🖖
Helmut Schütz
[image]

The quality of responses received is directly proportional to the quality of the question asked. 🚮
Science Quotes

Complete thread:

Activity
 Admin contact
21,592 posts in 4,514 threads, 1,532 registered users;
online 21 (0 registered, 21 guests [including 7 identified bots]).
Forum time: Friday 12:12 CEST (Europe/Vienna)

It is a capital mistake to theorise before one has data.    Arthur Conan Doyle

The Bioequivalence and Bioavailability Forum is hosted by
BEBAC Ing. Helmut Schütz
HTML5