Duplicate LLOQ and ULOQ [Bioanalytics]

posted by Ohlbe – France, 2020-05-26 21:43 (1401 d 15:23 ago) – Posting: # 21482
Views: 8,058

Dear Helmut,

❝ ❝ (in the past many labs analysed the LLOQ and ULOQ samples in duplicate, with only one of the results being used following an SOP, but all others were singlets).


❝ I see. How does such an SOP look like? Pick out the best?


Some just picked the one they liked best. I heard they ran into trouble :-D

Most frequent: use the first result only. If, and only if it fails (more than 20% deviation at the LLOQ, 15% at the ULOQ): exclude it and use the second. The aim being of course to avoid having to truncate the curve in case of failure.

I see this less and less often. Not sure why (guess: some FDA inspectors did not like it).

Regards
Ohlbe

Complete thread:

UA Flag
Activity
 Admin contact
22,957 posts in 4,819 threads, 1,639 registered users;
90 visitors (1 registered, 89 guests [including 3 identified bots]).
Forum time: 12:06 CET (Europe/Vienna)

Nothing shows a lack of mathematical education more
than an overly precise calculation.    Carl Friedrich Gauß

The Bioequivalence and Bioavailability Forum is hosted by
BEBAC Ing. Helmut Schütz
HTML5