An example of IS response. [Bioanalytics]

posted by Ohlbe – France, 2020-05-26 19:27 (1595 d 18:38 ago) – Posting: # 21478
Views: 9,429

Dear Shuanghe,

❝ I would make a suggestion to the CRO for the modification of their SOP to take incidence like this into consideration. One way is like you said, repeat another set of CC at the end. What else can we do to improve it?


They should read the FDA Q&A, but it does not help much here (there are some QC samples with an IS response tracking that of the subject samples).

They should also look at the special issue of Bioanalysis on this topic. It includes papers from both regulators and industry, with nice case studies. The overall conclusion is: acceptance criteria such as mean ± x% are useful, but not in all situations. Plotting the IS response to look for trends and systematic differences is a must.

Also a nice review, just published in the same journal, free access :-).

Regards
Ohlbe

Complete thread:

UA Flag
Activity
 Admin contact
23,249 posts in 4,885 threads, 1,652 registered users;
63 visitors (0 registered, 63 guests [including 8 identified bots]).
Forum time: 14:05 CEST (Europe/Vienna)

Nothing shocks me. I’m a scientist.    Harrison Ford (as Indiana Jones)

The Bioequivalence and Bioavailability Forum is hosted by
BEBAC Ing. Helmut Schütz
HTML5