Placement of CC [Bioanalytics]

posted by Helmut Homepage – Vienna, Austria, 2020-05-26 17:35 (1424 d 07:52 ago) – Posting: # 21473
Views: 8,229

Dear Ohlbe & Shuanghe,

I agree with all of your points.

❝ The calibration curve is not repeated at the end of the run, so there is no way to check what's happening at the ULOQ, where you may get linearity issues due to detector saturation or ion suppression of the analyte by the IS.


Yep. Many CROs have the CC in duplicates only at the start of the run. IMHO, that should be avoided. I prefer to have one set of singlets at the start and another at the end.

Dif-tor heh smusma 🖖🏼 Довге життя Україна! [image]
Helmut Schütz
[image]

The quality of responses received is directly proportional to the quality of the question asked. 🚮
Science Quotes

Complete thread:

UA Flag
Activity
 Admin contact
22,988 posts in 4,825 threads, 1,654 registered users;
92 visitors (0 registered, 92 guests [including 2 identified bots]).
Forum time: 01:28 CEST (Europe/Vienna)

The only way to comprehend what mathematicians mean by Infinity
is to contemplate the extent of human stupidity.    Voltaire

The Bioequivalence and Bioavailability Forum is hosted by
BEBAC Ing. Helmut Schütz
HTML5