Placement of CC [Bioanalytics]

posted by Helmut Homepage – Vienna, Austria, 2020-05-26 15:35 (155 d 14:57 ago) – Posting: # 21473
Views: 4,568

Dear Ohlbe & Shuanghe,

I agree with all of your points.

» The calibration curve is not repeated at the end of the run, so there is no way to check what's happening at the ULOQ, where you may get linearity issues due to detector saturation or ion suppression of the analyte by the IS.

Yep. Many CROs have the CC in duplicates only at the start of the run. IMHO, that should be avoided. I prefer to have one set of singlets at the start and another at the end.

Dif-tor heh smusma 🖖
Helmut Schütz
[image]

The quality of responses received is directly proportional to the quality of the question asked. 🚮
Science Quotes

Complete thread:

Activity
 Admin contact
21,179 posts in 4,414 threads, 1,477 registered users;
online 3 (0 registered, 3 guests [including 2 identified bots]).
Forum time: Thursday 05:33 CET (Europe/Vienna)

With four parameters I can fit an elephant,
and with five I can make him wiggle his trunk.    John von Neumann

The Bioequivalence and Bioavailability Forum is hosted by
BEBAC Ing. Helmut Schütz
HTML5