Good question, next question [Study Assessment]
❝ […] information about the records where Cmax is the first point (as well as records where AUC0-t/AUC0-∞<0.8 and records where pre-dose sample > 5% Cmax). Could the regulators ban the results of the BE study with such data? Did you ever faced with this in your practice? What was the percentage of subjects with such data (for example for 0.8 it is common to suppose more than 20%)? If there is only one or two subjects with Cmax in the first point should we perform additional analysis excluding them in order to show that this doesn't affect the results if it was not stated in the protocol?
OK, let’s have a look at the guideline:
Sampling times
- The sampling schedule should include frequent sampling around predicted tmax to provide a reliable estimate of peak exposure. In particular, the sampling schedule should be planned to avoid Cmax being the first point of a concentration time curve.
- The sampling schedule should also cover the plasma concentration time curve long enough to provide a reliable estimate of the extent of exposure which is achieved if AUC(0-t) covers at least 80% of AUC(0-∞).
Reasons for exclusion
- Subjects with non-zero baseline concentrations > 5% of Cmax. Such data should be excluded from bioequivalence calculation (see carry-over effects below).
The above can, for immediate release formulations, be the result of […] an insufficient wash-out period, […] and should as far as possible be avoided […] by designing the study with a sufficient wash-out period. The samples from subjects excluded from the statistical analysis should still be assayed and the results listed.
If there are any subjects for whom the pre-dose concentration is greater than 5 percent of the Cmax value for the subject in that period, the statistical analysis should be performed with the data from that subject for that period excluded. In a 2-period trial this will result in the subject being removed from the analysis. The trial will no longer be considered acceptable if these exclusions result in fewer than 12 subjects being evaluable.
- AUC(0-t) should cover at least 80% of AUC(0-∞). Subjects should not be excluded from the statistical analysis if AUC(0-t) covers less than 80% of AUC(0-∞), but if the percentage is less than 80% in more than 20% of the observations then the validity of the study may need to be discussed.
- First-point Cmax “should be avoided”. Though not elaborated in the section about exclusion, I would state it in the protocol. See also what Ohlbe wrote above. I would gather that excluding >20% of subjects will lead to troubles (similar to high residual AUC).
Note that the FDA is more generous: If you have data from a pilot study showing that the sampling schedule in the pivotal was adequate, you don’t have to worry (just bad luck, data acceptable).
- Subjects with carry-over can be excluded (hopefully ≥12 left). No discussion needed if procedure stated in the protocol. Do it better next time.
- High residual AUC. Clearly stated in the GL, though I would be happy to discuss “the validity of the study”. For IR products sampling far beyond 2×tmax is scientifically questionable.
Dif-tor heh smusma 🖖🏼 Довге життя Україна!
Helmut Schütz
The quality of responses received is directly proportional to the quality of the question asked. 🚮
Science Quotes
Complete thread:
- BE assessment regarding first time point Cmax Mann 2018-12-09 10:14
- BE assessment regarding first time point Cmax Ohlbe 2018-12-09 23:46
- BE assessment regarding first time point Cmax Mann 2018-12-10 00:28
- BE assessment regarding first time point Cmax Ohlbe 2018-12-11 00:05
- FDA and outliers mittyri 2018-12-11 16:38
- BE assessment regarding first time point Cmax Erkin 2018-12-28 08:29
- BE assessment regarding first time point Cmax ElMaestro 2018-12-28 14:32
- BE assessment regarding first time point Cmax Erkin 2018-12-28 20:57
- regulatory reasons to consider the study inadequate Astea 2019-03-02 18:02
- Good question, next questionHelmut 2019-03-03 19:42
- first point Cmax: trouble waiting to happen? Astea 2019-03-03 21:42
- Good question, next questionHelmut 2019-03-03 19:42
- regulatory reasons to consider the study inadequate Astea 2019-03-02 18:02
- BE assessment regarding first time point Cmax Erkin 2018-12-28 20:57
- BE assessment regarding first time point Cmax ElMaestro 2018-12-28 14:32
- BE assessment regarding first time point Cmax Mann 2018-12-10 00:28
- BE assessment regarding first time point Cmax Ohlbe 2018-12-09 23:46