in-vitro population bioequivalence [Regulatives / Guidelines]

posted by ElMaestro  – Denmark, 2019-02-06 08:55 (753 d 10:54 ago) – Posting: # 19879
Views: 1,115

Hi kumarnaidu,

» We have conducted the Beclomethasone in-vitro population bioequivalence study for spray pattern test. Ten units for three batch of test and reference products were used for beginning life stage only.
»
» In this experiment, three reading were measured each canisters of beginning life stage only (m=1). My question is such case, n: Number of canisters or bottles per batch, for T and R products would be 10 or 30.

n is the number of units per batch; in this case it sounds like you have n=10. This is usually not enough to secure a reasonable chance to show BE for nasal products and inhaled products (regardless of pMDI, DPI or nebs).

Pass or fail!
ElMaestro