No random calibrators (and QCs) [Bioanalytics]

posted by Helmut Homepage – Vienna, Austria, 2019-02-02 16:49 (1881 d 23:39 ago) – Posting: # 19846
Views: 2,474

Hi Bert,

❝ Is there any guidance for the location of calibration samples in an analytical run?


Not that I recall. I agree with ElMaestro’s comments.

❝ Should this be at-random?


Not a good idea – even if there is no substantial carry-over (see the EMA’s GL Section 4.2.2). If you decide to inject calibrators at the start and end of the batch do so with increasing concentrations which minimizes the relative impact of carry-over. If you have calibrators within the set of unknown samples inject a blank after. The same is applicable for QC samples as well.

❝ (Like it, called Junior :-) )


Not referring to your age or experience but the number of posts in the forum. :-D
Junior : <25
Regular: 25 < 100
Senior : 100 < 500
Hero   : >500

Dif-tor heh smusma 🖖🏼 Довге життя Україна! [image]
Helmut Schütz
[image]

The quality of responses received is directly proportional to the quality of the question asked. 🚮
Science Quotes

Complete thread:

UA Flag
Activity
 Admin contact
22,957 posts in 4,819 threads, 1,636 registered users;
94 visitors (0 registered, 94 guests [including 9 identified bots]).
Forum time: 16:28 CET (Europe/Vienna)

With four parameters I can fit an elephant,
and with five I can make him wiggle his trunk.    John von Neumann

The Bioequivalence and Bioavailability Forum is hosted by
BEBAC Ing. Helmut Schütz
HTML5